Reading OSCAR leak stats ?

Post by ejbpesca » Sun Dec 25, 2022 9:56 am

Does this look like an okay leak stat report? There's that 95% figure again which I still do not understand. Is that an okay % of time over threshold? I do understand that <25 on part of the OSCAR report is an okay leak score. I have leak problems and they seem correlate more with good/bad sleep than the AHI score.

Screenshot 2022-12-25 at 8.02.00 AM.png
Screenshot 2022-12-25 at 8.02.00 AM.png (39.51 KiB) Viewed 2905 times

_________________
MachineMask

User avatar
LSAT
Posts: 13397
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Reading OSCAR leak stats ?

Post by LSAT » Sun Dec 25, 2022 1:13 pm

Hard to tell by looking only at these numbers. Leaks should be 24L or below for the machine to be able to compensate for them. 96% means that 95% of the time you were at this level OR BELOW.
Last edited by LSAT on Thu Dec 29, 2022 4:21 pm, edited 1 time in total.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Reading OSCAR leak stats ?

Post by Okie bipap » Sun Dec 25, 2022 6:02 pm

If the leaks are not bothering you, I wouldn't worry about it. Resmed says anything above 24 liters is excessive. However, the machine can handle at least 34 liters and still adequately treat your apnea.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Re: Reading OSCAR leak stats ?

Post by ozij » Mon Dec 26, 2022 4:34 am

ejbpesca wrote:
Sun Dec 25, 2022 9:56 am
Is that an okay % of time over threshold? I do understand that <25 on part of the OSCAR report is an okay leak score. I have leak problems and they seem correlate more with good/bad sleep than the AHI score.
If the leaks bother you, they're not OK, not matter what the fromal threshold si.

The table you posted shows that you're spending a growing percent of your time above the leak threshold - so clearly something has to be taken care of.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
ejbpesca
Posts: 343
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Reading OSCAR leak stats ?

Post by ejbpesca » Thu Dec 29, 2022 3:11 pm

LSAT wrote:
Sun Dec 25, 2022 1:13 pm
Hard to tell by looking only at these numbers. Leaks should be 24L or below for the machine to be able to compensate for them. 96% means that 95% of the time you were at this level or below.
I get the 95% of the time at that level, it is the or below that makes the level number meaningless to me. If 95% of the time I may have been anywhere below the level indicated, what does that tell me?

_________________
MachineMask

User avatar
Pugsy
Posts: 65283
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Reading OSCAR leak stats ?

Post by Pugsy » Thu Dec 29, 2022 4:06 pm

ejbpesca wrote:
Thu Dec 29, 2022 3:11 pm
If 95% of the time I may have been anywhere below the level indicated, what does that tell me?
95% numbers are just numbers and don't mean much of anything. So it doesn't tell you a lot and you just have to learn to accept the fact that the "or below" part of the definition is just words. 95% numbers...be it pressure or leak...are just numbers. They aren't the holy grail of anything like people want to make them out to be.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Reading OSCAR leak stats ?

Post by robysue1 » Thu Dec 29, 2022 5:47 pm

ejbpesca wrote:
Thu Dec 29, 2022 3:11 pm
LSAT wrote:
Sun Dec 25, 2022 1:13 pm
Hard to tell by looking only at these numbers. Leaks should be 24L or below for the machine to be able to compensate for them. 96% means that 95% of the time you were at this level or below.
I get the 95% of the time at that level, it is the or below that makes the level number meaningless to me. If 95% of the time I may have been anywhere below the level indicated, what does that tell me?
95 percentile, or 95% for short, is a statistics term. It may be easiest to illustrate what this means with a couple of sets of discrete set of data.

The following list contains 100 numbers on it, arranged in increasing order. The 95th number is highlighted.

0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
0, 0, 0, 0, 0, 0, 0, 0, 0, 0,
0, 0, 0, 0, 0, 0, 0, 1, 1, 1,
1, 1, 1, 1, 1, 1, 1, 1, 1, 1,
1, 1, 1, 1, 1, 1, 1, 2, 3, 4,
5, 7, 8, 8, 9, 9, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 11, 13, 15, 18

That 95th number on the list is a 10, and that means that the 95% for this set of numbers is 10. In other words, 95 percent of the numbers on this list are at or below 10. And at most 5 percent of the numbers on the list can be strictly above 10. (In this particular example, only 4 of the numbers on the list are strictly above 10 because the 96th percentile and the 95th percentile are both equal to 10.). You should also notice that 57 of the numbers on the list are actually 0's. So the majority of numbers on this list are significantly less than the 95% number. But that does not always happen. Consider this list of data that also has a 95% equal to 10:

0, 1, 2, 3, 4, 5, 6, 7, 8, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 10, 10, 10, 10, 10,
10, 10, 10, 10, 10, 11, 12, 13, 15, 18

Again, 95 percent of the numbers on this list are at or below 10. And at most 5% of the numbers on the list are strictly above 10. But 86 of the numbers on this list are equal to 10. So in this case the 95 percentile data point equals the vast majority of the data on the list.

In the context of the Resmed leak data, if your 95% leak rate is below 24 L/min (the Resmed definition of "Large Leak"), then you can be sure that the time spent in seriously large leaks is less than 5% of the entire night. From a practical point of view, that means that if your 95% leak rate is below 24 L/min, you only need to worry about leaks if they are waking you up because any genuine large leaks did not last long enough to adversely affect your therapy.

Also in the context of Resmed leak data, if your 95% leak rate is well below 24 L/min---as in your 95% leak rate is in the single digits---that indicates that the total length of time spent in moderate sized leaks (say leaks with a rate between 10 and 24 L/min) is less than 5% of the night.

And finally in the context of Resmed leak data, if your 95% leak rate is at or above 24 L/min, that indicates that it is well worth looking at the leak data for the night in Oscar's daily data to see whether you had really large leaks for a bit more than 5% of the night or whether you were leaking like a sieve most of the night. If your median leak rate is anywhere close to 24 L/min, then you also need to look carefully at the detailed daily data to see just how bad the leaks are.

Does that help answer your questions?
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

User avatar
ejbpesca
Posts: 343
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Reading OSCAR leak stats ?

Post by ejbpesca » Fri Dec 30, 2022 7:22 am

robysue1 wrote:
Thu Dec 29, 2022 5:47 pm
ejbpesca wrote:
Thu Dec 29, 2022 3:11 pm
LSAT wrote:
Sun Dec 25, 2022 1:13 pm

Does that help answer your questions?
Yes. Thank you. I passed stats class with a B, but left more confused than I entered.

I'm sure there is some esoteric, but practical, stats applications of the 95 percentile as shown in example with number groups, but as far as users of CPAP machines go, it would be better to state leaks in a simple quantitative format like events shows in OSCAR. At least it is not shown by a whisker box.

_________________
MachineMask

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Reading OSCAR leak stats ?

Post by robysue1 » Sat Dec 31, 2022 12:19 am

ejbpesca wrote:
Fri Dec 30, 2022 7:22 am
robysue1 wrote:
Thu Dec 29, 2022 5:47 pm
Does that help answer your questions?
Yes. Thank you. I passed stats class with a B, but left more confused than I entered.

I'm sure there is some esoteric, but practical, stats applications of the 95 percentile as shown in example with number groups, but as far as users of CPAP machines go, it would be better to state leaks in a simple quantitative format like events shows in OSCAR. At least it is not shown by a whisker box.
But leaks are often not discrete, short-lived events. When folks have serious leak problems that must be addressed, it is possible that they are leaking at a reasonably high rate all night long. Of course this will be obvious if you look at the leak graph rather than just the statistical numbers in the left side bar or the summary data.

In my own opinion, the 95% leak rate is useful in demonstrating that leaks are not a problem that is adversely affecting the quality of xPAP therapy: If your 95% leak rate is below the definition of "Large Leak", then you only need to worry about leaks if they are waking you up.

On the other hand, if your 95% leak rate is in "Large Leak" territory, more investigation needs to be made. On the Resmeds, the first indication that large leaks are lasting long enough to adversely affect the quality of the xPAP therapy is the dreaded Mr. Red Frowny Face. He shows up if your unintentional leak rate is above 24 L/min for at least 30% of the night. (In stats talk, that means the 70% is at or above 24 L/min.)

The median leak rate shown in Oscar is also useful to look at when trying to figure out whether your leaks are large enough and long enough to adversely affect xPAP therapy: The closer the median leak rate is to 24 L/min, the more likely that your xPAP therapy is being seriously compromised by excessively long large leaks.