Best CPAP backup battery?
- dacotto1984
- Posts: 32
- Joined: Sun Aug 28, 2022 5:32 pm
Best CPAP backup battery?
Hi,
I was wondering, what backup battery would most people recommend?
I was wondering, what backup battery would most people recommend?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15502
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Best CPAP backup battery?
Do you need this for power failures at home only? How many nights do you expect to be out? If you are out for more than one night, can you get to someone during the day who has power to recharge the battery?
Re: Best CPAP backup battery?
I have no idea what machine you have. the portable outlet battery will power most caps for 2 nights on air only. if you need humidity and heated hose, only one night. the best way to go is get a dc converter and plug into a lithium battery.
I use the resmed air sense 11. I use the portable outlet battery. I use 50 percent for 8 hours, air only.
I use the resmed air sense 11. I use the portable outlet battery. I use 50 percent for 8 hours, air only.
- ChicagoGranny
- Posts: 15502
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Best CPAP backup battery?
In a group I’m in on Facebook, I have seen this unit mentioned several times and highly recommended.dacotto1984 wrote: ↑Thu Dec 22, 2022 7:27 pmHi,
I was wondering, what backup battery would most people recommend?
https://www.amazon.com/Jackery-Portable ... 945&sr=8-4
Was said to power a cpap machine for 3 full nights of sleep on a full charge. I don’t know if that is with or without a Humidifier. You would need a step down converter since your ResMed Airsense 10 is a 24v unit and the battery is for 12v. The neat thing about this battery is that they make a solar panel you can attach to the battery, so whether you are camping or in a bad power outage situation, you could still charge the battery during the day. It’s a bit of an investment, but could really see this being a great solution for camping or an emergency situation.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Silicone Nasal Pillow CPAP Mask with Headgear |
APAP mode set to 10.0-20.0/No Ramp/Humidity maxed out
- Dog Slobber
- Posts: 4328
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Best CPAP backup battery?
No, you don't need a step down converter.
ResMed's are 24v, but the DC-DC Converter that ResMed's use draw 12v and does the conversion for you.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Best CPAP backup battery?
Good to know, thanks for clearing that up. I haven’t bought one yet, so I was going off what someone else had stated.No, you don't need a step down converter.
ResMed's are 24v, but the DC-DC Converter that ResMed's use draw 12v and does the conversion for you.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Silicone Nasal Pillow CPAP Mask with Headgear |
APAP mode set to 10.0-20.0/No Ramp/Humidity maxed out
- loggerhead12
- Posts: 585
- Joined: Fri Sep 01, 2017 4:00 pm
Re: Best CPAP backup battery?
There are dozens of battery options presented in dozens of battery threads, all the way from cheap DIY to hyper-expensive turnkey units that will run your CPAP and refrigerator for a couple of nights.
Re: Best CPAP backup battery?
Wasn’t there some sort of cable or adapter that can be used to get more efficient juice?Dog Slobber wrote: ↑Fri Dec 23, 2022 1:30 pmNo, you don't need a step down converter.
ResMed's are 24v, but the DC-DC Converter that ResMed's use draw 12v and does the conversion for you.
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
- Dog Slobber
- Posts: 4328
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Best CPAP backup battery?
There are some batteries designed spicifically for ResMeds that output 24v that are cable only. They don't need the DC-DC converter.lynninnj wrote: ↑Fri Dec 23, 2022 5:17 pmWasn’t there some sort of cable or adapter that can be used to get more efficient juice?Dog Slobber wrote: ↑Fri Dec 23, 2022 1:30 pmNo, you don't need a step down converter.
ResMed's are 24v, but the DC-DC Converter that ResMed's use draw 12v and does the conversion for you.
We see this on some Freedom and the Pilot-24
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Best CPAP backup battery?
Just seems like a reminder to me to call resmed again. Maybe if folks keep asking they will get around to inventing one.Dog Slobber wrote: ↑Fri Dec 23, 2022 5:25 pmThere are some batteries designed spicifically for ResMeds that output 24v that are cable only. They don't need the DC-DC converter.lynninnj wrote: ↑Fri Dec 23, 2022 5:17 pmWasn’t there some sort of cable or adapter that can be used to get more efficient juice?Dog Slobber wrote: ↑Fri Dec 23, 2022 1:30 pmNo, you don't need a step down converter.
ResMed's are 24v, but the DC-DC Converter that ResMed's use draw 12v and does the conversion for you.
We see this on some Freedom and the Pilot-24
When we spoke of this before I thought you said the proper cable would cut down on draw/battery use when used in the DC car jack.
Am I remembering that correctly?
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
- Dog Slobber
- Posts: 4328
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Best CPAP backup battery?
Not quite.lynninnj wrote: ↑Sat Dec 24, 2022 9:05 amJust seems like a reminder to me to call resmed again. Maybe if folks keep asking they will get around to inventing one.Dog Slobber wrote: ↑Fri Dec 23, 2022 5:25 pm
There are some batteries designed spicifically for ResMeds that output 24v that are cable only. They don't need the DC-DC converter.
We see this on some Freedom and the Pilot-24
When we spoke of this before I thought you said the proper cable would cut down on draw/battery use when used in the DC car jack.
Am I remembering that correctly?
The reduced battery consumption is not because of the cable, it's because using a DC-DC converter removes the ac-inversion and back to dc.
A battery uses DC, our ResMeds also use DC. When we use a power-station with an AC outlet;
- The voltage is inverted from 12v-DC to 120v-AC. This results is a lot of power loss.
- The voltage is then converted from 120v-AC to 24v-DC by the ResMed's power brick. This results in more power loss.
When we convert 12v-DC to 24v-DC, some conversion loss will be experienced, but not nearly as much as either of the two steps described above.
ResMed's actually use 2 distinct dc voltages, 24v-DC and 3.5v-DC(*1), the cable has 3 wires. In an ideal world the power-station would output 24v-DC, so minimal conversion is required.
The batteries listed above, output 24v-DC.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Best CPAP backup battery?
Thanks DS!
_________________
| Machine: AirSense 11 Autoset |
| Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
| Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.
-
DaveMunson
- Posts: 266
- Joined: Thu Jun 23, 2005 8:36 pm
Re: Best CPAP backup battery?
I have a nice sine wave inverter, 100 ft extension cord and fused battery clips.
On loss of power I turn off the bipap humidifier heat and go connect the inverter on the car battery and extend extension cord from garage to the machine. If power is restored I plug my NOCO 10A charger on the car's battery. If power is not back, go drive the car for a while to recharge the battery.
Everything is in my closet, wrapped in saran wrap, ready to go. This is the only use for that power cord.
I set this up sometime around 2006 and have used it maybe a dozen times.
On loss of power I turn off the bipap humidifier heat and go connect the inverter on the car battery and extend extension cord from garage to the machine. If power is restored I plug my NOCO 10A charger on the car's battery. If power is not back, go drive the car for a while to recharge the battery.
Everything is in my closet, wrapped in saran wrap, ready to go. This is the only use for that power cord.
I set this up sometime around 2006 and have used it maybe a dozen times.
- dacotto1984
- Posts: 32
- Joined: Sun Aug 28, 2022 5:32 pm
Re: Best CPAP backup battery?
I have had the power go out for a few days. If i could get something that would last 2 or 3 days. then that would be more ideal i would say.ChicagoGranny wrote: ↑Thu Dec 22, 2022 7:49 pmDo you need this for power failures at home only? How many nights do you expect to be out? If you are out for more than one night, can you get to someone during the day who has power to recharge the battery?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |

