pressure support functions philips resperonics bipap

Post by keithy62 » Wed Nov 16, 2022 9:16 pm

Have just purchased a philips respironics bipap sv auto advanced recently and am waiting an appointment to see a sleep specialist which will take sometime but have been using it in the meantime as is much better than my old fisher and paykel cpap. Have set it up with pressures i seem to tolerate reasonably well but have a question regarding epap max or min pressures in relation to the pressure support max and min functions as to what these two functions actually do and how they relate to minimum and maximum epap pressures as raising one of the pressure support functions can seem to alter max or minim pressures. My first experience with a machine with pressure support so dont understand what it's function is

Current settings are: Max pressure 16.0 Min EPAP 6.00 -- Max EPAP 10.5 -- P Support Min 4.0 -- Pres Support Max 6.0 Model Number SN P09887473

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: pressure support functions philips resperonics bipap

Post by palerider » Wed Nov 16, 2022 9:23 pm

keithy62 wrote:
Wed Nov 16, 2022 9:16 pm
have a question regarding epap max or min pressures in relation to the pressure support max and min functions as to what these two functions actually do and how they relate to minimum and maximum epap pressures as raising
So, what's your question?

If it's just "what is PS"? it's the difference in the EPAP (exhale) pressure and IPAP (inhale pressure).
Classically, in bilevel treatment, PS is used to enhance ventilation of patients who're having breathing problems.

For most people on CPAP, it *might* just make it easier for them to breathe with the machine.

Too much PS will increase your chances of over-ventilation and end up increasing your central apnea count (Philips calls them "open airway" apneas, for no apparent good reason.

PS doesn't have any relation to the amount the EPAP increases.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

keithy62
Posts: 13
Joined: Tue Aug 26, 2014 10:44 pm

Re: pressure support functions philips resperonics bipap

Post by keithy62 » Wed Nov 16, 2022 10:04 pm

Thank you for your reply Palerider. Still dont understand function of PS. Thought pressure difference in my case would be the difference Min IPAP min and EPAP max with that being 4.5. Sorry for dumb question,my question prior to seeing my sleep specialist is are my PS settings low enough or not excessively high to cause central apnea events. Once again Thanks for your reply.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: pressure support functions philips resperonics bipap

Post by palerider » Wed Nov 16, 2022 10:57 pm

keithy62 wrote:
Wed Nov 16, 2022 10:04 pm
Thank you for your reply Palerider. Still dont understand function of PS. Thought pressure difference in my case would be the difference Min IPAP min and EPAP max with that being 4.5. Sorry for dumb question,my question prior to seeing my sleep specialist is are my PS settings low enough or not excessively high to cause central apnea events. Once again Thanks for your reply.
You're getting your terms mixed up.

There's no "min ipap" and "epap max" on machines.

There is MinEPAP, and MaxIPAP, and PS on auto adjusting machines. The EPAP varies based on breathing events.

PS is the difference in exhale pressure and inhale pressure on every breath.

If your EPAP on a single breath is 8, and you have a PS of 3, then when you inhale, the pressure will go up to 11, then drop back to 8 when you exhale.

If you want to know more about PS, then watch these videos:

https://www.youtube.com/watch?v=A-SlO2dU8Og
https://www.youtube.com/watch?v=BdeOiDJmbrk

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

keithy62
Posts: 13
Joined: Tue Aug 26, 2014 10:44 pm

Re: pressure support functions philips resperonics bipap

Post by keithy62 » Thu Nov 17, 2022 2:51 am

Apologise for that min ipap max epap mistake.Only saw mistake after i had posted. Thanks for further information pressure support function.functions.

User avatar
Respirator99
Posts: 389
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: pressure support functions philips resperonics bipap

Post by Respirator99 » Thu Nov 17, 2022 4:39 am

keithy62 wrote:
Wed Nov 16, 2022 10:04 pm
... are my PS settings low enough or not excessively high to cause central apnea events. Once again Thanks for your reply.
ASV machines are not like ordinary CPAP or APAPs. They need a high PSmax to enable them to treat central apneas if they occur. For example Resmed recommend PSmax be >=15. Here's an extract from the relevant section of the Resmed ASV manual - the Philips will be similar.

Pressure Support.PNG
Pressure Support.PNG (144.61 KiB) Viewed 451 times
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.