decker12 wrote: ↑Tue Nov 01, 2022 11:41 am
Looking at last night's data it does seem that once I'm asleep I'm not dipping below that 8 which I changed in the machine. However looking at the charts from the past week or two, the lowest my dips go is actually around a 6.
If the min pressure is set at 8cm and the ramp pressure is set at 8cm, there's no way the machine is going to deliver less than 8cm pressure on
inhalation, which is what SleepHQ is showing you.
Your exhalation pressure was dropping to around 5cm whenever your pressure setting was at 8cm during the night: That's because EPR = 3, and that means that every time you exhale, the pressure will be reduced by 3cm since your inhalation pressure is always greater than 7cm.
When you had your minimum pressure set to 5 or 6cm, there were occasional times when your inhalation pressure did drop to around 6. If EPR were set to 3, your inhalation pressure would have been reduced to 4cm, the minimum pressure the machine will deliver under any circumstances.
That seems to indicate that I should change the minimum to a 6 (and probably should have started at a 6 last night after all!)
Does that make sense?
Not necessarily. The lower your minimum pressure is as compared to what you typically need to prevent the worst clusters of your apneas from getting started, the greater the pressure swings will be. And the more likely the machine will find itself in a situation where it has to increase the pressure by 4, 5, 6 cm over a very short period of time (a minute or less). And those sudden pressure increases can cause some people to arouse. They can also cause some people to have problems with swallowing air. And, because the machine is now "chasing" events, rather than preventing them, more apneas and hypopneas might occur while the machine is ramping up the pressure to stabilize your airway.
Now starting the
ramp pressure at 6cm might make some sense: If breathing at 6cm (with the exhalation pressure reduced to 4cm because of EPR = 3 being set) is more comfortable for you than starting at 8cm is and if you get to sleep more quickly with a starting ramp of 6cm than you do at 8cm, then using the AutoRamp feature with a beginning pressure of 6cm and leaving your minimum pressure setting at 8cm would be a reasonable idea to try.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.