SleepHQ Sorry, there's no pressure data for today

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Sun Oct 09, 2022 8:57 am

SleepHQ reports "Sorry, there's no pressure data for today" for every day I've uploaded. All other data is present. The pressure data is shown in Oscar. I have an S9 VPAP Auto. I've tried searching the forums but haven't found anything relating to this. Can anybody steer me to what's wrong?

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: SleepHQ Sorry, there's no pressure data for today

Post by lynninnj » Sun Oct 09, 2022 9:40 am

Citation4444 wrote:
Sun Oct 09, 2022 8:57 am
SleepHQ reports "Sorry, there's no pressure data for today" for every day I've uploaded. All other data is present. The pressure data is shown in Oscar. I have an S9 VPAP Auto. I've tried searching the forums but haven't found anything relating to this. Can anybody steer me to what's wrong?
I have no clue what the S9s track on SD card. I am glad the Oscar at least works for you.

On occasion, if I did not properly insert the SD card, my AS11 doesn't register pressure data. There are other periodic issues from time to time where something is left off the chart and I don't know why. i still prefer Sleephq because I find it fast, easy to use, and portable in that I upload at home and at the office I can skim thru the data without dragging the card all over the place or downloading the program on my roommates computer.

If you select the "trends" page and look at pressure data does it break it down like mine does-where you can hover over a day and see a breakdown of pressures there? Or does it have it in the box near the top? If you hover you can see min/med/95/max in different colors? If you don't have that data it makes it tougher. https://sleephq.com/public/f5bfcbea-655 ... fab6e2ea9c that one didnt have the card in place correctly and missed the days data.Sept 7 and 8 are missing here as well in trends. https://sleephq.com/public/bd3ff895-75d ... e686a517e4

Sorry again as I don't know if your machine tracks pressure on SLeepHQ like Oscar does but maybe you get it on the trend data? hth

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Sun Oct 09, 2022 9:53 am

Does not show in trend data either. If I hover over trend pressure box comes up with all colors at 0.
https://sleephq.com/public/1623a37f-6d6 ... 3c36095628

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: SleepHQ Sorry, there's no pressure data for today

Post by Julie » Sun Oct 09, 2022 10:20 am

Looks like you might have tried to get results after 12 noon - which would give you one hour at best as the machine works from noon to noon... so you need to check results when you wake up and not wait til later.

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Pugsy » Sun Oct 09, 2022 10:24 am

Julie wrote:
Sun Oct 09, 2022 10:20 am
Looks like you might have tried to get results after 12 noon - which would give you one hour at best as the machine works from noon to noon... so you need to check results when you wake up and not wait til later.
The noon thing ONLY pertains to the S9 on the LCD screen data sleep reports.
Has absolutely nothing to do with any software results from OSCAR/SleepyHead/SleepHQ which gets its data from the files added to the machine's SD card.

And just show you know...the noon thing NO LONGER applies to the LCD screen report on the AirSense 10 or 11 models.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Sun Oct 09, 2022 10:26 am

Julie wrote:
Sun Oct 09, 2022 10:20 am
Looks like you might have tried to get results after 12 noon - which would give you one hour at best as the machine works from noon to noon... so you need to check results when you wake up and not wait til later.
Every day reports the same, not just the most current day. For example, here's another day. https://sleephq.com/public/4e357c92-dc9 ... 114136392b

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Pugsy » Sun Oct 09, 2022 10:42 am

My best guess is that there was/is a corrupted file pertaining to pressure. A problem that SleepHQ has a problem with...unknown if OSCAR would have a problem with it but if OSCAR showed pressure line graph then we would assume a SleepHQ problem.

You have the other detailed data and if it was a problem with the machine writing to the SD card you wouldn't have the additional detailed data graphs.

In general when we see one off weird stuff like this that just doesn't make any logical sense we tend to blame "corrupted data file" or my personal favorite....aliens screwing with us again.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Sun Oct 09, 2022 10:49 am

I've tried uploading from my laptop which has a direct SD slot and got the same results. I'll try an SD card from one of my S9 VPAP Adapt machines and see it it behaves the same.
UPDATE: Got exact same results from the SD card of my S9 VPAP Adapt ASV machine. Must be a bug with SleepHQ not reading S9 data correctly. Neither OSCAR, Sleepyhead, nor ResMed programs have this problem.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037
Last edited by Citation4444 on Sun Oct 09, 2022 11:06 am, edited 1 time in total.

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: SleepHQ Sorry, there's no pressure data for today

Post by Julie » Sun Oct 09, 2022 10:58 am

Will keep the noon/noon thing in mind, thx.

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Sun Oct 09, 2022 11:14 am

I have reported this as a bug to the SleepHQ team.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: SleepHQ Sorry, there's no pressure data for today

Post by clownbell » Mon Oct 10, 2022 4:51 pm

Why not just use Oscar? It is a known program.

Why complicate your life with SleepHQ?
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

Citation4444
Posts: 56
Joined: Thu Dec 05, 2013 4:16 pm
Location: North GA Mountains, USA

Re: SleepHQ Sorry, there's no pressure data for today

Post by Citation4444 » Mon Oct 10, 2022 9:52 pm

I’m fighting a duel with aerophagia and need a better way to share data. SleepHQ could help, but I do use Oscar.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Now using S9 VPAP Adapt ASV 36037

User avatar
zonker
Posts: 11349
Joined: Fri Jun 19, 2015 4:36 pm

Re: SleepHQ Sorry, there's no pressure data for today

Post by zonker » Mon Oct 10, 2022 10:04 pm

Citation4444 wrote:
Mon Oct 10, 2022 9:52 pm
I’m fighting a duel with aerophagia and need a better way to share data. SleepHQ could help, but I do use Oscar.
you can post here from oscar. here's a thread to explain it.
viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

also-
viewtopic/t183806/EVERYONE-PLEASE-READ- ... MENTS.html

sorry if you know this already.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg