remstar auto question
remstar auto question
I have a remstar auto. It will come on automatically when I put on my mask, but does not turn off automatically. I think I am the one that turned on auto on, but I can't remember. I didn't see it in the online manual.
Does anyone know how to turn auto off on with the remstar auto? It is an older one without cflex, but it is an auto.
Thanks
Does anyone know how to turn auto off on with the remstar auto? It is an older one without cflex, but it is an auto.
Thanks
This may be what you're looking for:
Patient disconnect setting: 1 = on 0 = off
(It's near the end of the options.)
Den
Patient disconnect setting: 1 = on 0 = off
(It's near the end of the options.)
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
Guest
I think it's the nature of the beast!
When you breath into the mask the air flow should automatically come on.
If you remove the mask (and DO NOT press the on/off dutton) the air flow will go to the low pressure after a few seconds. Then airflow will automatically turn off within 2 minutes.
These are features that ARE, I don't think that there is any provision for turning then off, other than pressing the on/off button before they kick in.
When you breath into the mask the air flow should automatically come on.
If you remove the mask (and DO NOT press the on/off dutton) the air flow will go to the low pressure after a few seconds. Then airflow will automatically turn off within 2 minutes.
These are features that ARE, I don't think that there is any provision for turning then off, other than pressing the on/off button before they kick in.
goofproof, are you saying that if you don't turn the machine off with the buttons, but leave it running and just disconnect it from the mask while, let's say, going to the bathroom, that registers as a large mask leak? I thought the machine could sense whether or not you had the mask off or on, thus the "blower hours" versus "therapy hours."
Forgive me, four week newbie.
Pam
Forgive me, four week newbie.
Pam
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Additional Comments: Oscar Software | APAP: 9-10 |
Re: remstar auto question
[quote="bigcat"]I have a remstar auto. It will come on automatically when I put on my mask, but does not turn off automatically. I think I am the one that turned on auto on, but I can't remember. I didn't see it in the online manual.
Does anyone know how to turn auto off on with the remstar auto? It is an older one without cflex, but it is an auto.
Thanks
Does anyone know how to turn auto off on with the remstar auto? It is an older one without cflex, but it is an auto.
Thanks
someday science will catch up to what I'm saying...
-
Guest
It shows up as a leak and gets avg into the figures. I don't know if it thinks you died, or are trying to play games with it, but that button is real easy to figure out how to use. Some don't play the data game, so for them it doesn't matter.WearyOne wrote:goofproof, are you saying that if you don't turn the machine off with the buttons, but leave it running and just disconnect it from the mask while, let's say, going to the bathroom, that registers as a large mask leak? I thought the machine could sense whether or not you had the mask off or on, thus the "blower hours" versus "therapy hours."
The machine hours can't be user reset, the user hours can be from the chincal screen. I figure as with everything I own, we get a limited numbers hours of use out of it, so I try to use mine for what it's designed for. Jim
The two types of hours are to warn you if someone tries to sell you a unit for new thats used.