Noobie Needs Help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Noobie Needs Help

Post by Pugsy » Sun May 29, 2022 8:46 pm

You might benefit from the additional features of the VAuto. It can go to 25 cm if needed.
And it has the ability to give you higher relief during exhale than the AutoSet if you end up needing higher pressures you might like more relief than EPR can give you.

When people start seeing pressures in the teens being needed then the extra relief often comes in handy.

The AutoSet has EPR 1,2 or 3.
The VAuto has something similar in function called PS pressure support and it will do 1 to 5, or 6 or more.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Mon May 30, 2022 2:05 pm

Pugsy wrote:
Sun May 29, 2022 8:46 pm
You might benefit from the additional features of the VAuto. It can go to 25 cm if needed.
And it has the ability to give you higher relief during exhale than the AutoSet if you end up needing higher pressures you might like more relief than EPR can give you.

When people start seeing pressures in the teens being needed then the extra relief often comes in handy.

The AutoSet has EPR 1,2 or 3.
The VAuto has something similar in function called PS pressure support and it will do 1 to 5, or 6 or more.
Thanks for your unofficial, unmedical opinion.
I will talk to the pulmonologist tomorrow and ask about getting the AC10 VPAP.
I am going to take a printout of the OSCAR screenshots with me, for them to look at.

User avatar
zonker
Posts: 11335
Joined: Fri Jun 19, 2015 4:36 pm

Re: Noobie Needs Help

Post by zonker » Mon May 30, 2022 3:45 pm

Iamstumped wrote:
Mon May 30, 2022 2:05 pm

I will talk to the pulmonologist tomorrow and ask about getting the AC10 VPAP.
I am going to take a printout of the OSCAR screenshots with me, for them to look at.
if it were me, i'd take the sd card as well. not many doctors are aware of the oscar software.

good luck with the vpap.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Fri Jun 03, 2022 4:02 pm

screenshot-20220603-115019.png
This was last night.
I was using an N30i (SW) on a (Sm) frame.
Started wearing a chin strap, but I removed that in the middle of the night, as it became sore to continue to wear it.
I also took 2, 5mg tablets of Melatonin before going to bed.
Fell asleep fairly quickly, did not wake up during the night, that I know of.
Feel somewhat good today.


P.S. the pulmonologist took 8 days of data, and determined that my 30 day average of my 95% pressure was only 10.5 cmh2O
REALLY?
How can you get a 30 day average with only 8 days worth of data? :? :?

clownbell
Posts: 369
Joined: Fri Apr 02, 2021 8:46 pm

Re: Noobie Needs Help

Post by clownbell » Sat Jun 04, 2022 10:22 pm

That last chart was a real keeper. Whatever you did, try and do it more often. I am not an advanced math whiz, but reasonably proficient with plain arithmetic. That said, I can't see how you can take 8 days of data and determine a 30 day average which is LESS than your 8 day average. The good news is that you can change your settings if doc sets to 10.5.
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask

User avatar
zonker
Posts: 11335
Joined: Fri Jun 19, 2015 4:36 pm

Re: Noobie Needs Help

Post by zonker » Sun Jun 05, 2022 11:04 am

Iamstumped wrote:
Fri Jun 03, 2022 4:02 pm

How can you get a 30 day average with only 8 days worth of data? :? :?
ex·tra·po·la·tion
/ikˌstrapəˈlāSH(ə)n/
Learn to pronounce
noun
the action of estimating or concluding something by assuming that existing trends will continue or a current method will remain applicable.
"sizes were estimated by extrapolation"
MATHEMATICS
the extension of a graph, curve, or range of values by inferring unknown values from trends in the known data.
"extrapolation of the logarithmic curve yielded an estimate of 66 species"

sorry for the poor formatting. it's the first thing that pops up when one googles "extrapolation".

https://www.google.com/search?q=extrapo ... e&ie=UTF-8

math is fun!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Tue Jun 07, 2022 2:42 pm

screenshot-20220607-131909.png
Last night I used the Amara FFM (L). Did not have any leaks, that I'm aware of.

Seems like every time that I use a FFM, my AHI numbers go up, and there is an UGLY cluster of UA's, in the beginning, while I am trying to fall asleep, but when I use a nasal mask, my AHI numbers do go down, and there is very few UA's, in the beginning while I am trying to fall asleep.

When I use a FFM, should I up my min pressure? Currently at 14cmh2O. I was looking at the manuals that came with each FFM, and it looks like the maker recommends increasing the pressure due to a pressure drop.

Or maybe I should decrease the EPR from 3 to 2, or even lower it from 3 to 1, or maybe even turn it off completely?

User avatar
Miss Emerita
Posts: 3753
Joined: Sun Nov 04, 2018 8:07 pm

Re: Noobie Needs Help

Post by Miss Emerita » Tue Jun 07, 2022 11:03 pm

The large leak line somehow got nixed from your events graph. It’d be good to restore it. If you look on the left panel, you’ll see you were running large leaks almost the whole time. So check you long hose your short hose, and all your hose connections. If it wasn’t your mask leaking, it’s got to be something else.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Tue Jun 07, 2022 11:58 pm

Miss Emerita wrote:
Tue Jun 07, 2022 11:03 pm
The large leak line somehow got nixed from your events graph. It’d be good to restore it. If you look on the left panel, you’ll see you were running large leaks almost the whole time. So check you long hose your short hose, and all your hose connections. If it wasn’t your mask leaking, it’s got to be something else.
Actually, I unchecked it, under the events tab. I will restore it right away.

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Mon Jun 13, 2022 10:52 am

screenshot-20220613-092131.png
Should I adjust the min pressure settings on my machine?

According to OSCAR, my 95% pressure is just barely above the min pressure setting (12 cmh2O) on my machine.

Question: what is that, that is just below the SD Card slot, on the left side, on the AS10

User avatar
ozij
Posts: 10461
Joined: Fri Mar 18, 2005 11:52 pm

Re: Noobie Needs Help

Post by ozij » Tue Jun 14, 2022 11:34 am

Iamstumped wrote:
Mon Jun 13, 2022 10:52 am
Should I adjust the min pressure settings on my machine?
Why?
According to OSCAR, your pressure goes abov1 12 only 5% of the time.
So what do you mean by "adjust"? Is it bothering you and you'd prefer it to be lower part of the time?

How do you feel?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023


User avatar
zonker
Posts: 11335
Joined: Fri Jun 19, 2015 4:36 pm

Re: Noobie Needs Help

Post by zonker » Fri Jun 17, 2022 6:55 pm

Iamstumped wrote:
Fri Jun 17, 2022 1:57 pm
Let me know if this worked.

https://imgur.com/xhLUNEB

It looks like it did
this is fine for now. next time, would like to see the "events flags" across the top. and we don't need the ahi chart.

how'd you sleep? looks like a decent enough chart all told.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Iamstumped
Posts: 157
Joined: Tue May 17, 2022 2:56 am
Location: Kansas City, Mo.

Re: Noobie Needs Help

Post by Iamstumped » Sat Jun 18, 2022 2:00 pm

zonker wrote:
Fri Jun 17, 2022 6:55 pm
Iamstumped wrote:
Fri Jun 17, 2022 1:57 pm
Let me know if this worked.

https://imgur.com/xhLUNEB

It looks like it did
this is fine for now. next time, would like to see the "events flags" across the top. and we don't need the ahi chart.

how'd you sleep? looks like a decent enough chart all told.
No left side bar, just the actual charts?

Fair, I guess. That was just a test, to see if I could link to imgur, like that instead of doing an attachment to here since it is getting full again.

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Noobie Needs Help

Post by Pugsy » Sat Jun 18, 2022 2:15 pm

Left side bar is always needed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.