Pressure Increases w/o Events

Post by Steptoe » Tue May 03, 2022 8:56 am

I use a DreamStation 2 with APAP pressures at 9.5 min., 11.0 max., with A-Flex at 1.

I have recently noticed numerous pressure increases that do not appear to be in response to or have any relation to apnea, hypopnea, or other events. This often occurs at the beginning of the night before I fall asleep, but it also occurs while I am sleeping at various times throughout the night. I have attached OSCAR screenshots that illustrate this issue. It was my understanding that APAP pressure increases only happen in response to events and do not occur in advance to prevent events.

Any ideas as to why this is happening or if it is something I should be concerned about.

_________________
MachineMask
Additional Comments: Also use a Fisher & Paykel SleepStyle for travel & as a backup.

Tec5
Posts: 173
Joined: Mon Mar 28, 2022 10:32 am

Re: Pressure Increases w/o Events

Post by Tec5 » Tue May 03, 2022 9:23 am

Unfortunately your Oscar charts did not get attached. will you please try again.
I have no experience with Dreamstation 2; however Dreamstation 1 seems to increase the pressure in response to a flow limitation (in absence of any other "event").

See the attachment.

This may apply in your particular case.
Dreamstation 1 Response to FL.jpg
Dreamstation 1 Response to FL.jpg (26.31 KiB) Viewed 992 times
I am neither a physician nor a lawyer, so DO NOT rely on me for professional medical or legal advice.

rick blaine
Posts: 616
Joined: Tue Dec 15, 2015 7:30 am

Re: Pressure Increases w/o Events

Post by rick blaine » Tue May 03, 2022 9:30 am

Hi Steptoe,

I can't speak for the Dreamstation 2, but the algorithms of earlier PR machines – such as my 961 – are such that, even when 'nothing is going on', every so often the machine ups the pressure by 1.5 cm – just to see if that makes any difference. And if it doesn't, the algorithms bring the pressure down again, by 1.5 cm.

If it is that, then it's nothing to be concerned about.

User avatar
SJ-Paddler
Posts: 97
Joined: Wed Feb 09, 2022 2:29 pm
Location: South Jersey

Re: Pressure Increases w/o Events

Post by SJ-Paddler » Tue May 03, 2022 9:35 am

Up front I know nothing about PR machines.

That said the RedMed Airsense manual says it adjust pressure by looking at "Flow Limit, Snore, and Apena Events".

Did you see any Flow Limit increases or Snore increases when the pressure went up?

_________________
MachineMask
Mask: Dreamware Silicone Nasal Pillows
Machine: ResMed Airsense 10 Autoset (For Him)
APAP Pressure: 13 cm - 15 cm / Average 13.3 cm / 3 cm EPR Pressure Relief

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Pressure Increases w/o Events

Post by Rubicon » Tue May 03, 2022 9:42 am

Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Pressure Increases w/o Events

Post by dataq1 » Tue May 03, 2022 10:08 am

Saw your Oscar in the other forum - no need to attach here unless you want the narrative to be complete here.

The clinical manual for your machine does indicate that Respironics algorithm does indeed change pressure to "PROACTIVELY" challenge the airway,
From the photo below, it appears that these challenges are issued to see if the morphology (shape of the breathing curve) can be improved, or conversely a decrease in pressure is warranted.
Dreamstation 1 Proactive search for best flow morphology.jpg
Dreamstation 1 Proactive search for best flow morphology.jpg (59.37 KiB) Viewed 975 times
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Pugsy
Posts: 65267
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure Increases w/o Events

Post by Pugsy » Tue May 03, 2022 10:11 am

pressure probes.PNG
pressure probes.PNG (14.67 KiB) Viewed 973 times
Above taken from your posted images over on Jason's forum. If anyone wants to see what he posted..it's here.
http://freecpapadvice.com/forum/viewtop ... =10&t=6915

Are you talking about these little increases that appear in a saw tooth pattern on parts of the night??
If so, those are normal pressure probes that Respironics auto adjusting machines will do.

Nothing to worry about.
Something that Respironics has done for years and years. See below for a more info.
viewtopic/t80875/Respironics-APAP-press ... se-up.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure Increases w/o Events

Post by palerider » Tue May 03, 2022 9:00 pm

dataq1 wrote:
Tue May 03, 2022 10:08 am
Saw your Oscar in the other forum - no need to attach here unless you want the narrative to be complete here.

The clinical manual for your machine does indicate that Respironics algorithm does indeed change pressure to "PROACTIVELY" challenge the airway,
Way to repost something that's already been posted.

also, you're in error, the clinical manual for the OPs machine *does not* have that in it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Pressure Increases w/o Events

Post by dataq1 » Wed May 04, 2022 8:58 pm

palerider wrote:
Tue May 03, 2022 9:00 pm
Way to repost something that's already been posted.
I just don't understand the animus <sigh>

The graphic that I posted at 11:08 am had NOT been posted in this thread as far as I am aware, so I don't understand what you are talking about regarding "reposting".

Frankly I thought that the graphic (from a Resmed clinicians book) was a rather complete and nicely illustrated explanation for the pressure
" blips" that the original poster was asking about, and that Pugsey referred to 3 minutes after my post as being a technique that Resmed has used for years.

Moreover, it seems to clearly illustrate that the Resmed algorithm does in fact increase pressure (albeit short ~ 2 minute "probes") in advance of an event.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

Steptoe
Posts: 12
Joined: Mon May 24, 2021 3:01 pm

Re: Pressure Increases w/o Events

Post by Steptoe » Wed May 04, 2022 9:37 pm

dataq1

I certainly appreciate the responses I received from you and Pugsy regarding my question. Was confused by the blips on my DS2 OSCAR reports and you resolved my confusion with these detailed answers. I am sorry you had to endure unwarranted abuse from this palerider character. I found palerider’s post outrageous and obnoxious.

Steptoe

_________________
MachineMask
Additional Comments: Also use a Fisher & Paykel SleepStyle for travel & as a backup.
Last edited by Steptoe on Wed May 04, 2022 9:47 pm, edited 2 times in total.

User avatar
Pugsy
Posts: 65267
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure Increases w/o Events

Post by Pugsy » Wed May 04, 2022 9:40 pm

I do believe that the discussion here was about Respironics and not ResMed.
Did you have ResMed on your mind?
I wasn't talking about ResMed at all. Thought I was pretty clear about Respironics pressure probes.

BTW when I started composing my post above you hadn't posted yours yet but it took me a while to get it all done and by then you had posted. I never saw what you posted until way later because I didn't bother to look.


dataq1 wrote:
Wed May 04, 2022 8:58 pm
palerider wrote:
Tue May 03, 2022 9:00 pm
Way to repost something that's already been posted.
I just don't understand the animus <sigh>

The graphic that I posted at 11:08 am had NOT been posted in this thread as far as I am aware, so I don't understand what you are talking about regarding "reposting".

Frankly I thought that the graphic (from a Resmed clinicians book) was a rather complete and nicely illustrated explanation for the pressure
" blips" that the original poster was asking about, and that Pugsey referred to 3 minutes after my post as being a technique that Resmed has used for years.

Moreover, it seems to clearly illustrate that the Resmed algorithm does in fact increase pressure (albeit short ~ 2 minute "probes") in advance of an event.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure Increases w/o Events

Post by palerider » Wed May 04, 2022 10:01 pm

dataq1 wrote:
Wed May 04, 2022 8:58 pm
palerider wrote:
Tue May 03, 2022 9:00 pm
Way to repost something that's already been posted.
I just don't understand the animus <sigh>

The graphic that I posted at 11:08 am had NOT been posted in this thread as far as I am aware, so I don't understand what you are talking about regarding "reposting".
Because you posted a graphic that you got from a post Rubicon had made in the same thread. ergo, a repost.
dataq1 wrote:
Wed May 04, 2022 8:58 pm
Frankly I thought that the graphic (from a Resmed clinicians book) was a rather complete and nicely illustrated explanation for the pressure
" blips" that the original poster was asking about, and that Pugsey referred to 3 minutes after my post as being a technique that Resmed has used for years.

Moreover, it seems to clearly illustrate that the Resmed algorithm does in fact increase pressure (albeit short ~ 2 minute "probes") in advance of an event.
Resmed does not do pressure probes.

Please just stop posting, it makes more work for people that have to correct your noise.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Pressure Increases w/o Events

Post by dataq1 » Thu May 05, 2022 7:44 am

palerider wrote:
Wed May 04, 2022 10:01 pm
Resmed does not do pressure probes.
Yes, of course, Respironics uses these P-optimal searches (aka probes) while Resmed does not. mea cupla
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Pressure Increases w/o Events

Post by dataq1 » Thu May 05, 2022 7:53 am

Pugsy wrote:
Wed May 04, 2022 9:40 pm
I do believe that the discussion here was about Respironics and not ResMed.
Did you have ResMed on your mind?
I wasn't talking about ResMed at all. Thought I was pretty clear about Respironics pressure probes.

BTW when I started composing my post above you hadn't posted yours yet but it took me a while to get it all done and by then you had posted. I never saw what you posted until way later because I didn't bother to look.
Yeah, I had Resmed on my mind, so my mistake!

Regarding composing my reply with the graphic that apparently ____ is upset about, I also fell into the same time warp, in that I had not yet seen Rubicon's link.
Stuff happens.... (and will happen again - we are all just human)