Mold in Heated Respironics Tube.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
clkerl
Posts: 3
Joined: Sat Apr 09, 2022 12:10 pm

Mold in Heated Respironics Tube.

Post by clkerl » Sat Apr 09, 2022 12:58 pm

I currently have a Dreamstation one and I have been using it for at least a couple of years. Before anybody says, I know, I'm a very lazy idiot who hasn't been properly cleaning their very important medial device. I haven't been washing all of the pieces of my headgear and tubing every day, only occasionally. As far as I can tell, the end of the heated tube is the only place I can see any actual mold (pictured in attachment). Trying to clean it off with my fingers has yielded no success so far, it's basically stuck on there. I've already ordered a new tube, but I wanted to know if there might be a way to reliably clean it off so it'd at least be usable again.

I'd also like to know if this is likely to be the cause of my nose feeling very "stuffed up", both when I sleep and periodically throughout the day. It's not actually filled up with snot, it's like I have sinus inflammation or something that makes breathing through my nose difficult, which I fear is effecting my treatment.

Thanks in advance.

Image

User avatar
ChicagoGranny
Posts: 15085
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Mold in Heated Respironics Tube.

Post by ChicagoGranny » Sat Apr 09, 2022 1:30 pm

clkerl wrote:
Sat Apr 09, 2022 12:58 pm
I'm a very lazy idiot who hasn't been properly cleaning their very important medial device. I haven't been washing all of the pieces of my headgear and tubing every day, only occasionally.
Around here, you could get called an idiot if you have been washing all of the pieces every day. If that is indeed mold, it may have grown from your washing the hose and hanging it up to dry. Some of us use the same hose for years without washing it. We never have a problem with mold or dirt of any kind.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Mold in Heated Respironics Tube.

Post by chunkyfrog » Sat Apr 09, 2022 5:39 pm

Agreeing with granny 100%.
I never wash the hose, and only wash the tank when it gets stuff growing in it.
I wipe my nasal pillows with a baby wipe, but not often.
New ultrafine filter every 3 weeks.
I did exactly the same through 4 months of chemo last year; with no ill effects.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
zonker
Posts: 11237
Joined: Fri Jun 19, 2015 4:36 pm

Re: Mold in Heated Respironics Tube.

Post by zonker » Sat Apr 09, 2022 6:33 pm

chunkyfrog wrote:
Sat Apr 09, 2022 5:39 pm

New ultrafine filter every 3 weeks.
oh, somebody mentioned filter, time to check mine.

thanks froggy! it still looks good.
:wink:
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Mold in Heated Respironics Tube.

Post by chunkyfrog » Sat Apr 09, 2022 6:52 pm

You are welcome!
I do stuff old-school--recording my daily essentials-plus data in a notebook.
My notebooks have 21 lines, so the end of the page triggers a new filter, pillows,
and laying down the columns for the next page.
Less for my little amphibian brain to remember.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

clkerl
Posts: 3
Joined: Sat Apr 09, 2022 12:10 pm

Re: Mold in Heated Respironics Tube.

Post by clkerl » Sat Apr 09, 2022 7:13 pm

ChicagoGranny wrote:
Sat Apr 09, 2022 1:30 pm
If that is indeed mold, it may have grown from your washing the hose and hanging it up to dry. Some of us use the same hose for years without washing it. We never have a problem with mold or dirt of any kind.
See, when I do wash it I almost always wash it before bed. I don't leave it up to dry.
Is it really okay not clean every day? To be clear, I am using the humidifier.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Mold in Heated Respironics Tube.

Post by chunkyfrog » Sat Apr 09, 2022 7:51 pm

Totally OK, especially if you are using only distilled water.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
ChicagoGranny
Posts: 15085
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Mold in Heated Respironics Tube.

Post by ChicagoGranny » Sat Apr 09, 2022 8:20 pm

clkerl wrote:
Sat Apr 09, 2022 7:13 pm
See, when I do wash it I almost always wash it before bed. I don't leave it up to dry.
Aha! That may be the source of the mold. The connector still has a bit of water on it when you connect it to the machine and mask. Once firmly connected, the water won't dry - mold potential!

I'm not suggesting you dry the connector. Just quit washing the hose.
clkerl wrote:
Sat Apr 09, 2022 7:13 pm
Is it really okay not clean every day?
You have our experience. Now, it is up to you to choose. If you choose washing, we will remember you when you come again complaining about mold. :lol:

User avatar
Pugsy
Posts: 64936
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Mold in Heated Respironics Tube.

Post by Pugsy » Sat Apr 09, 2022 9:32 pm

Many, many years ago (before even heated hoses) I had one of my hoses develop maybe 3 little specks like yours.
So not nearly as much....but they were there and in just about the same place as yours.

I never washed my hose back then even....so I don't think it is related to not washing of the hose.
I tried everything I could think of to get rid of the black spots but I couldn't.
Now mold (if that is what it is/was) is actually easy to get rid of but I think that maybe what I saw after cleaning was a stain from whatever it was. I know our first thought is mold and it scares us but mold is really easy to kill and remove.
Stains not so much so. Now I don't know for sure but I think what you may be seeing is residual staining that just won't come off. I know I tried Comet cleanser on mine and it didn't touch it.

I tried every cleaning product I owned and couldn't remove the stain.
You can keep trying and maybe you will get lucky but I would just toss it. Hoses are cheap. Peace of mind is priceless.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

southerndoc
Posts: 242
Joined: Sat Apr 23, 2011 9:32 pm
Location: Marietta, GA USA

Re: Mold in Heated Respironics Tube.

Post by southerndoc » Sat Apr 09, 2022 10:49 pm

Did you use regular or distilled water?

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: OSCAR 1.3.1, SleepHQ

Ray4852
Posts: 113
Joined: Tue Nov 16, 2021 7:24 am

Re: Mold in Heated Respironics Tube.

Post by Ray4852 » Sun Apr 10, 2022 6:23 am

dirty hose. get a new one. you got your monies worth out of that hose. The purpose of cleaning a hose is prevent bacteria getting in a hose. if you don't clean it. you are putting yourself at higher risk of getting sick of outside germs. I clean my hose every day. it smells nice and fresh.

User avatar
zonker
Posts: 11237
Joined: Fri Jun 19, 2015 4:36 pm

Re: Mold in Heated Respironics Tube.

Post by zonker » Sun Apr 10, 2022 11:33 am

Ray4852 wrote:
Sun Apr 10, 2022 6:23 am
if you don't clean it. you are putting yourself at higher risk of getting sick of outside germs.
what do you do about the outside germs you breath when not on cpap?
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mold in Heated Respironics Tube.

Post by palerider » Sun Apr 10, 2022 4:13 pm

clkerl wrote:
Sat Apr 09, 2022 12:58 pm
As far as I can tell, the end of the heated tube is the only place I can see any actual mold (pictured in attachment).
IF that is actually mold, it happened because of the way you're "cleaning" things.

If you look the part of your connector is the part that has the discoloration is what is *covered* by the plastic part of your mask connector, there's no way that anything can get to that part of the connector when you've got the mask plugged into it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Mold in Heated Respironics Tube.

Post by palerider » Sun Apr 10, 2022 4:17 pm

Ray4852 wrote:
Sun Apr 10, 2022 6:23 am
dirty hose. get a new one. you got your monies worth out of that hose. The purpose of cleaning a hose is prevent bacteria getting in a hose. if you don't clean it. you are putting yourself at higher risk of getting sick of outside germs. I clean my hose every day. it smells nice and fresh.
The real purpose of 'cleaning' the hose is to degrade it so that it will fail sooner and have to be replace, thus generating more profit for the DMEs and manufacturers.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Grumpy48
Posts: 414
Joined: Tue May 12, 2020 7:57 am

Re: Mold in Heated Respironics Tube.

Post by Grumpy48 » Sun Apr 10, 2022 6:26 pm

I'm thinking the spotting inside the hose end might be a dark colored fabric lint that may have adhered to the mask connector at some point when it was washed and still wet and subsequently stained the rubber in the hose(?)

Mold most usually needs a spore and dampness to start growth. If there are mold spores as dried dust present in the air or surfaces where you washed the hose and mask, it's possible the hose/mask could have become contaminated and growth could have occurred. Is there a remote chance there might be a mold issue in your house? The bathroom in particular if hot showers are taken and the air isn't exhausted well.
https://www.health.ny.gov/publications/7287/
https://www.mayoclinic.org/diseases-con ... c-20351519

If it were me and I wanted to try and save the hose, I might try a 1:1 solution of household bleach and water and use a cotton Q-Tip to swab around the inside of the connection and a cold water rinse to remove any residual bleach. Wash it after as noted below and let it dry well.

I do wash my hose and mask usually once a month if I remember. The mask and headgear get put in a large plastic container that has a screw on lid (large nut container). A couple drops of Dawn dish liquid and hot water to 3/4's full, shake vigorously followed by a couple cold water fills and rinses. The heated hose gets a couple drops of the Dawn and filled with hot water and 'sloshed' back and forth a few times followed by a hot water flush at the tap. Both the hose and mask/headgear are hung outside in the sun to dry. One of the hoses I cleaned this way lasted well over a year and the only reason I replaced it was the DME had supplied me with replacement hoses and not because the hose fell apart or was defective in any way. I have never had and black spotting or other grime what so ever for the two years I've been doing this and the gear smells 'Dawn' fresh after. The water chamber gets a similar cleaning with Dawn and hot water. Daily if I remember the mask gets a wipe down with a baby wipe.

_________________
Machine: DreamStation 2 Auto CPAP Advanced with Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: Resmed AirSense 10 Auto with Humidifier