Did I damage my APAP?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
andy88488
Posts: 234
Joined: Fri Jul 06, 2007 8:09 am
Location: Northern NJ

Did I damage my APAP?

Post by andy88488 » Thu Jan 13, 2022 11:30 am

Hi All,

I have an Airsense 10 APAP. Last night, I knocked it off of the nighttable and onto the floor. (I've never done this before, by the way!)

I picked up the APAP, put it right side up, and tried to go back to sleep. The APAP kept running, but there was a very loud gurgling noise every time I breathed. It turned out that either when the APAP fell, or when I picked it up and righted it, it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.

After I removed the hose and drained it, the machine seemed to function fine for the rest of the night. But I'm worried about what the water might have done to the air pumping mechanism or other innards of the machine.

Can anyone tell me if I did or did not damage the machine?

Thanks,
Andy
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask

User avatar
Pugsy
Posts: 65416
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Did I damage my APAP?

Post by Pugsy » Thu Jan 13, 2022 11:37 am

Generally speaking if the machine seems to run normally now then the water didn't make inside the machine to fry the electronics.
Probably what happened with the overturn was water took the path of least resistance and when from water chamber into the long hose and that's where all the noise came from.
While not totally impossible for water to go into the motor side of things....not nearly as common as it was years ago.
The water chambers or purposely designed to not be so easy for upturned water chamber contents to go into the motor side of things. Instead designed to go to the long hose.

When electronics meet water...you usually know it...either it won't power up at all or there is a god awful noise.
If it is moving air like it should and not making a god awful noise in the motor....it's probably fine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
andy88488
Posts: 234
Joined: Fri Jul 06, 2007 8:09 am
Location: Northern NJ

Re: Did I damage my APAP?

Post by andy88488 » Thu Jan 13, 2022 11:41 am

When electronics meet water...you usually know it
Well put!! Thanks for your help, Pugsy!
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask

User avatar
zonker
Posts: 11504
Joined: Fri Jun 19, 2015 4:36 pm

Re: Did I damage my APAP?

Post by zonker » Thu Jan 13, 2022 11:44 am

Pugsy wrote: ↑
Thu Jan 13, 2022 11:37 am

When electronics meet water...you usually know it...
:lol: :lol: :lol:

this reminds of my stint as an admin at sony service center. at one time, sony was running ads showing happy young people cavorting on the beach capturing their precious memories on video tape. this drove our manager nuts. "don't they realize that water and electronic don't mix?!?!".
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Did I damage my APAP?

Post by Goofproof » Thu Jan 13, 2022 1:16 pm

zonker wrote: ↑
Thu Jan 13, 2022 11:44 am
Pugsy wrote: ↑
Thu Jan 13, 2022 11:37 am

When electronics meet water...you usually know it...
:lol: :lol: :lol:

this reminds of my stint as an admin at sony service center. at one time, sony was running ads showing happy young people cavorting on the beach capturing their precious memories on video tape. this drove our manager nuts. "don't they realize that water and electronic don't mix?!?!".
Actually they do mix, but the result is less than satisfying. With the XPAP running unless the air intake was in the water stream, water would not get in the pump, the air flow would Keep the water out of the machine, and blow it into the hose from HH. Causing the noises you heard. Waterboarding Anyone? :lol:
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Did I damage my APAP?

Post by chunkyfrog » Thu Jan 13, 2022 1:24 pm

I remember the Lucas magic smoke.
Click and Clack--one podcast I miss terribly.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Did I damage my APAP?

Post by palerider » Thu Jan 13, 2022 1:55 pm

andy88488 wrote: ↑
Thu Jan 13, 2022 11:30 am
it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
So, there's the blower unit, THEN the humidifier, THEN the hose, so it's entirely possible, especially if it was running at the time, that no water got into the machine itself, that it just got blown into the hose.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
andy88488
Posts: 234
Joined: Fri Jul 06, 2007 8:09 am
Location: Northern NJ

Re: Did I damage my APAP?

Post by andy88488 » Thu Jan 13, 2022 2:11 pm

palerider wrote: ↑
Thu Jan 13, 2022 1:55 pm
andy88488 wrote: ↑
Thu Jan 13, 2022 11:30 am
it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
So, there's the blower unit, THEN the humidifier, THEN the hose, so it's entirely possible, especially if it was running at the time, that no water got into the machine itself, that it just got blown into the hose.
Ah, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask

scf
Posts: 116
Joined: Sat Sep 18, 2021 9:32 am

Re: Did I damage my APAP?

Post by scf » Thu Jan 13, 2022 2:16 pm

I did the exact same thing about a week ago. And sat it back up on the table upside down, lol. It is working fine and seems to have suffered no damage. I have since put some "museum gel" on the bottom so it won't go anywhere.
ResMed AirSense Autoset FOR HER/F30i

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Did I damage my APAP?

Post by palerider » Thu Jan 13, 2022 4:49 pm

andy88488 wrote: ↑
Thu Jan 13, 2022 2:11 pm
palerider wrote: ↑
Thu Jan 13, 2022 1:55 pm
andy88488 wrote: ↑
Thu Jan 13, 2022 11:30 am
it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
So, there's the blower unit, THEN the humidifier, THEN the hose, so it's entirely possible, especially if it was running at the time, that no water got into the machine itself, that it just got blown into the hose.
Ah, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
Easy enough mistake to make with the way the air circuit appears to go back into the machine, but if you look in the hole with the chamber out, the left opening goes to the blower assembly, and the right opening is an elbow that goes out the back of the machine.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Conrad
Posts: 712
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Did I damage my APAP?

Post by Conrad » Fri Jan 14, 2022 7:44 am

chunkyfrog wrote: ↑
Thu Jan 13, 2022 1:24 pm
I remember the Lucas magic smoke.
Click and Clack--one podcast I miss terribly.
Yep! You never want to lose the magic smoke.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
babydinosnoreless
Posts: 2375
Joined: Fri Nov 02, 2018 2:53 pm

Re: Did I damage my APAP?

Post by babydinosnoreless » Fri Jan 14, 2022 8:19 am

andy88488 wrote: ↑
Thu Jan 13, 2022 11:30 am
Hi All,

I have an Airsense 10 APAP. Last night, I knocked it off of the nighttable and onto the floor. (I've never done this before, by the way!)
I did this with my aircurve. I put it in front of a fan and let it dry all day, then plugged it in the next night and it worked. Thank goodness !

User avatar
andy88488
Posts: 234
Joined: Fri Jul 06, 2007 8:09 am
Location: Northern NJ

Re: Did I damage my APAP?

Post by andy88488 » Fri Jan 14, 2022 8:24 am

palerider wrote: ↑
Thu Jan 13, 2022 4:49 pm
andy88488 wrote: ↑
Thu Jan 13, 2022 2:11 pm
palerider wrote: ↑
Thu Jan 13, 2022 1:55 pm
andy88488 wrote: ↑
Thu Jan 13, 2022 11:30 am
it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
So, there's the blower unit, THEN the humidifier, THEN the hose, so it's entirely possible, especially if it was running at the time, that no water got into the machine itself, that it just got blown into the hose.
Ah, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
Easy enough mistake to make with the way the air circuit appears to go back into the machine, but if you look in the hole with the chamber out, the left opening goes to the blower assembly, and the right opening is an elbow that goes out the back of the machine.
Yeah, I took a look at that last night, and sure enough, you are 100% correct. Thanks, you've set my mind at ease.
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask

User avatar
SleepyCPAP
Posts: 333
Joined: Wed Dec 08, 2010 6:01 am

Re: Did I damage my APAP?

Post by SleepyCPAP » Fri Jan 14, 2022 9:34 am

I took apart my ResMed Airsense 10 AutoSet apart when I first got it (it was well used, and needed to be de-smelled). There is no way for the water to go from the chamber to the electronics board(s), it would just come out the filter if you were adept at pouring water in there.

It is possible to get water from the humidifier into the motor chamber on a fall with water in the humidifier, but you have to be particularly skilled at doing it. I apparently do have that skill, when my machine came off the stand one night. I shook the machine to get drops out of the motor chamber (never opened it up) and let it dry out for three days while I used my backup machine. I don’t think water ever made it into the motor/blower itself. It runs as well as ever.

- SleepyCPAP

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Use OSCAR. Combine AlaxoStent with VAuto for perfect 0.0 AHI at PS 3.6 over 4cm EPAP
-- SleepyCPAP
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI