Fed up with spluttering, what can I do ?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
BlizzardUK
Posts: 77
Joined: Tue Aug 18, 2015 4:20 pm

Fed up with spluttering, what can I do ?

Post by BlizzardUK » Tue Dec 07, 2021 12:31 pm

I use nasal pillows, and for the last 2 months I keep getting woken up with spluttering of water and gurgling noises. I then have to take the mask off, de-attach the pillows, then swing the tube around to remove excess water. Then a couple of hours later I will be woken up again with the same. I have a heated hose, put on number 4 on the Respironics 560P dial, I have the humidity put on only 1. I didn't have this problem for the previous 6 months, seems to have been just when the weather changed, unless that is coincidence. Any idea why this could be happening ? Why does it splutter, spit and gurgle ? I have heard of rainout but I thought a heated hose stops that ?

_________________
Mask

User avatar
Miss Emerita
Posts: 3768
Joined: Sun Nov 04, 2018 8:07 pm

Re: Fed up with spluttering, what can I do ?

Post by Miss Emerita » Tue Dec 07, 2021 12:34 pm

Cooler weather means more condensation. Try upping the temperature a little bit, and invest in a hose cover. I use one made out of soft terrycloth, and I got an extra long one and scrunched it up. This provides a lot of insulation.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Fed up with spluttering, what can I do ?

Post by Julie » Tue Dec 07, 2021 1:04 pm

Also place the machine lower than yourself and the water will run back to it, rather than to you.

User avatar
Pugsy
Posts: 65140
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Fed up with spluttering, what can I do ?

Post by Pugsy » Tue Dec 07, 2021 1:07 pm

What is your bedroom temperature now?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11365
Joined: Fri Jun 19, 2015 4:36 pm

Re: Fed up with spluttering, what can I do ?

Post by zonker » Tue Dec 07, 2021 1:23 pm

.
Last edited by zonker on Tue Dec 07, 2021 6:35 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65140
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Fed up with spluttering, what can I do ?

Post by Pugsy » Tue Dec 07, 2021 1:43 pm

BlizzardUK wrote:
Tue Dec 07, 2021 12:31 pm
? I have heard of rainout but I thought a heated hose stops that ?
It all depends on how cold the bedroom ambient temps are. The heated hose can only do so much warming...and remember there is usually a non heated hose coming after the heated hose that goes to the mask..and when we have water in the nasal pillows...often that heated hose air simply gets cooled off too much by the time the air makes it to the face.

Either increase the hose air temp...and/or add a hose cozy over the short hose from the end of the heated hose to the mask so that the air in the non heated hose can stay warmer.

Or...increase the bedroom air temperature.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

CPAP2017
Posts: 58
Joined: Wed Mar 11, 2020 12:28 pm

Re: Fed up with spluttering, what can I do ?

Post by CPAP2017 » Tue Dec 07, 2021 2:48 pm

I have the same issue going from summer to fall. I manually adjusted the humidity control on my machine down one notch.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Fed up with spluttering, what can I do ?

Post by chunkyfrog » Tue Dec 07, 2021 3:33 pm

Girl's knee socks from the dollar store--cut the toes off, and line up a few on your hose.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

BlizzardUK
Posts: 77
Joined: Tue Aug 18, 2015 4:20 pm

Re: Fed up with spluttering, what can I do ?

Post by BlizzardUK » Tue Dec 07, 2021 6:42 pm

Thanks everyone. My bedroom is usually around 30 percent humidity and 20 degrees centigrade for warmth. I will try a hose cover for the shorter hose part, thanks for the advice, will let you know how I get on.

_________________
Mask

User avatar
Conrad
Posts: 671
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Fed up with spluttering, what can I do ?

Post by Conrad » Wed Dec 08, 2021 7:13 am

BlizzardUK wrote:
Tue Dec 07, 2021 12:31 pm
I use nasal pillows, and for the last 2 months I keep getting woken up with spluttering of water and gurgling noises. I then have to take the mask off, de-attach the pillows, then swing the tube around to remove excess water. Then a couple of hours later I will be woken up again with the same. I have a heated hose, put on number 4 on the Respironics 560P dial, I have the humidity put on only 1. I didn't have this problem for the previous 6 months, seems to have been just when the weather changed, unless that is coincidence. Any idea why this could be happening ? Why does it splutter, spit and gurgle ? I have heard of rainout but I thought a heated hose stops that ?
When was the last time that your heated hose was replaced with a new one?

Keep in mind, although unlikely, it's possible for the heated hose to have stopped working.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.