Large Leak

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
speckledhen
Posts: 2
Joined: Sun Nov 14, 2021 2:28 am

Large Leak

Post by speckledhen » Sun Nov 14, 2021 2:39 am

Last night my ahi was 1.66 which I am very happy with. However sleepyhead shows I had 45% large leak. How is this possible? I would have expected the ahi to be through the roof with such a leak.

User avatar
Dog Slobber
Posts: 4258
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Large Leak

Post by Dog Slobber » Sun Nov 14, 2021 8:13 am

  1. Just because you have a lot of leaks, doesn't mean you aren't being treated, nor does it mean you aren't receiving adequate pressure. The machine compensates for leaks, within reason.
  2. You can't necessarily trust the summary of your machine because of the excessive leaks. As machines experience large leaks, they lose their ability to accurately detect apneas and the type of apnea.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Large Leak

Post by Pugsy » Sun Nov 14, 2021 8:15 am

Need more information please before we can really comment.

First thing is...what machine are you using? Need to know that because different brands report leaks differently.

Second is ....just how bad (how deep into large leak territory did you go) exactly were the leaks? Barely into large leak territory or way deep into large leak territory???

We worry about leaks affecting therapy because we don't know if that nice low AHI is because nothing happened or maybe the leaks were so bad the machine simply couldn't sense the apnea events happening.

Since you posted a 45% of the time in large leak statistic ...I assume using OSCAR so how about posting the detailed report?
This format please
viewtopic/t158560/How-to-post-images-for-review.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

speckledhen
Posts: 2
Joined: Sun Nov 14, 2021 2:28 am

Re: Large Leak

Post by speckledhen » Sun Nov 14, 2021 11:30 am

Thanks for the responses. The machine is a system one and I'me afraid the rest of your reply has gone way over my head.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Large Leak

Post by Pugsy » Sun Nov 14, 2021 11:41 am

speckledhen wrote:
Sun Nov 14, 2021 11:30 am
Thanks for the responses. The machine is a system one and I'me afraid the rest of your reply has gone way over my head.
Respironics machines report large leak differently than the ResMed machines.
OSCAR defaults to the ResMed large leak threshold number for calculations of the percentage of time spent in large leak.
Large leak territory is totally different on a Respironics machine so that 45% statistic may not be an accurate reflection of time spent in large leak.

Look up on the actual EVENTS graph....see the LL line??? Any LL flagged areas up on the Events graph??
That's a machine flagged event and not OSCAR calculation....Machine data point trumps OSCAR generated statistic.

If you would post a screen shot of the detailed report we could tell you whether your large leak statistic of 45% means you really have a problem or not.
Without visually seeing exactly what is going on with the leak it's really impossible to give you more than an educated guess but most likely you aren't having nearly as much time deep in large leak territory for there to be much of a chance for the machine to be clueless and missing flagging events.

If you aren't seeing a lot of LL flagged time on the Events graph...I would either ignore the % of time spent in large leak statistic...or turn it off...or reset the red line threshold to something more accurate for a Respironics machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: Large Leak

Post by zonker » Sun Nov 14, 2021 11:49 am

op states she is using sleepyhead. i know, same-same, but just so they don't get confused.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg