Stuck on Maximum Pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
bgavin
Posts: 13
Joined: Sat Oct 02, 2021 9:50 am

Stuck on Maximum Pressure

Post by bgavin » Sat Oct 02, 2021 10:02 am

REMstar Auto C-Flex CPAP (2006)

I have entered Setup and tested all sorts of configurations to reduce pressure.
Nothing changes.
CPAP, APAP, CFLEX... all stuck at maximum blower pressure.

This is causing me dry mouth from mask leaks and too much pressure.
I did fine for 15 years with nasal mask.
The last six months has been awful dry mouth.
Been through FFM, etc, etc.

I think the machine is broken and stuck on full pressure.
I hate to discard it, if it can be opened and repaired.

Thoughts?
REMstar Auto C-Flex CPAP (2006)

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Stuck on Maximum Pressure

Post by Miss Emerita » Sat Oct 02, 2021 10:43 am

Your machine is pretty old! It seems plausible that it is not working right.

Did you find the settings for maximum pressure (along with the ones you specifically mentioned)?

I'm hoping some other forum members can help with the question about repairs.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
ChicagoGranny
Posts: 15523
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Stuck on Maximum Pressure

Post by ChicagoGranny » Sat Oct 02, 2021 10:47 am

bgavin wrote:
Sat Oct 02, 2021 10:02 am
I have entered Setup and tested all sorts of configurations to reduce pressure.
Nothing changes.
CPAP, APAP, CFLEX... all stuck at maximum blower pressure.
Did you change the minimum pressure to something lower than 20.0? (Yeah, my name is Captain Obvious.)
bgavin wrote:
Sat Oct 02, 2021 10:02 am
REMstar Auto C-Flex CPAP (2006)



I hate to discard it, if it can be opened and repaired.

A 15-year old machine?? You are long overdue a new machine. ResMed AutoSet.

bgavin
Posts: 13
Joined: Sat Oct 02, 2021 9:50 am

Re: Stuck on Maximum Pressure

Post by bgavin » Sat Oct 02, 2021 10:54 am

Yes, the machine is 15 years old.
Yes, I have been all through all the settings.
Yes, I have experimented with CPAP only at minimum pressure, etc, etc, etc, etc, etc, etc,
Yes, I have the chops and tools to work on this sort of equipment.

I cannot find a service manual for this machine (yet).
Before I throw it out, I will crack it open and put a voltage drop across the blower motor to reduce the pressure.

I'm just curious if anybody else has experienced the same problem.
Power outages in my area are chronic, and I suspect something in the processor got fried from a sudden power loss.
REMstar Auto C-Flex CPAP (2006)

User avatar
Pugsy
Posts: 65376
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Stuck on Maximum Pressure

Post by Pugsy » Sat Oct 02, 2021 11:14 am

Send Wulfman a private message. He is familiar with your machine and might have the service manual for it.
He has a lot of manuals saved.

https://www.cpaptalk.com/memberlist.php ... ile&u=1951

Make plans for getting a new machine....just in case because sometimes things can't be repaired or the parts needed to repair something are no longer available.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Hang Fire
Posts: 471
Joined: Mon Nov 04, 2013 8:03 pm
Location: Raleigh, NC

Re: Stuck on Maximum Pressure

Post by Hang Fire » Sat Oct 02, 2021 11:16 am

bgavin wrote:
Sat Oct 02, 2021 10:54 am
I'm just curious if anybody else has experienced the same problem.
Probably, but they are all dead by now.

User avatar
zonker
Posts: 11468
Joined: Fri Jun 19, 2015 4:36 pm

Re: Stuck on Maximum Pressure

Post by zonker » Sat Oct 02, 2021 11:19 am

Hang Fire wrote:
Sat Oct 02, 2021 11:16 am
bgavin wrote:
Sat Oct 02, 2021 10:54 am
I'm just curious if anybody else has experienced the same problem.
Probably, but they are all dead by now.
Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4334
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Stuck on Maximum Pressure

Post by Dog Slobber » Sat Oct 02, 2021 11:27 am

Have you unplugged it for a few minutes, to reset it?

Just throwing it out there, this is obvious to many, but not everyone. And it's not in the list of things you've tried. Even if you do bush-fix it, i'd still be shopping for a replacement.
Battery Backup: EcoFlow Delta 2

bgavin
Posts: 13
Joined: Sat Oct 02, 2021 9:50 am

Re: Stuck on Maximum Pressure

Post by bgavin » Sat Oct 02, 2021 11:43 am

Pugsy, thanks for the referral. PM sent.

The machine has been reset numerous times with power outages over the years.
This bug developed about 6 months back.
I have it on a UPS now, but probably too little and too late for that.

As noted above, I'm old-school, so I'm always willing to try repairs before discarding it.
REMstar Auto C-Flex CPAP (2006)

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Stuck on Maximum Pressure

Post by Wulfman... » Sat Oct 02, 2021 12:17 pm

bgavin wrote:
Sat Oct 02, 2021 11:43 am
Pugsy, thanks for the referral. PM sent.

The machine has been reset numerous times with power outages over the years.
This bug developed about 6 months back.
I have it on a UPS now, but probably too little and too late for that.

As noted above, I'm old-school, so I'm always willing to try repairs before discarding it.
PM received and reply sent (before I saw this post).

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

bgavin
Posts: 13
Joined: Sat Oct 02, 2021 9:50 am

Re: Stuck on Maximum Pressure

Post by bgavin » Sat Oct 02, 2021 12:29 pm

PM sent, to continue through email.
I did find the factory service manual, sent you a copy for just in case.

Thanks for the rapid response.
REMstar Auto C-Flex CPAP (2006)

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Stuck on Maximum Pressure

Post by chunkyfrog » Sat Oct 02, 2021 12:34 pm

Keeping a possible replacement in mind, a good time to consider switching brands.
(Just my totally prejudiced point of view.) 8)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

bgavin
Posts: 13
Joined: Sat Oct 02, 2021 9:50 am

Re: Stuck on Maximum Pressure

Post by bgavin » Sat Oct 02, 2021 12:38 pm

I am clueless about the various brands.
Replacement is a cash purchase, no insurance.

The service manual shows part numbers for all the components.
Unless they are Unobtanium, my unit can be repaired.
REMstar Auto C-Flex CPAP (2006)

User avatar
ChicagoGranny
Posts: 15523
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Stuck on Maximum Pressure

Post by ChicagoGranny » Sat Oct 02, 2021 3:29 pm

bgavin wrote:
Sat Oct 02, 2021 12:38 pm
Unless they are Unobtanium, my unit can be repaired.
I don't know where to get parts, but this would be my first try - https://acbio.com/respiratory-equipment-parts/
bgavin wrote:
Sat Oct 02, 2021 12:38 pm
I am clueless about the various brands.
Replacement is a cash purchase, no insurance.
ResMed AutoSet for sure. When you start wanting to make a purchase, forum members can help you with model selection and where to make cash purchase. Supply throughout the market is restricted due to Philips recall - https://www.usa.philips.com/healthcare/ ... src-update

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Stuck on Maximum Pressure

Post by Wulfman... » Sat Oct 02, 2021 4:38 pm

bgavin wrote:
Sat Oct 02, 2021 12:38 pm
I am clueless about the various brands.
Replacement is a cash purchase, no insurance.

The service manual shows part numbers for all the components.
Unless they are Unobtanium, my unit can be repaired.
The link CG gave you (which I also sent you) is your best possible source for spare parts or service for that machine.
You might also check the firmware version number.......one of the first numbers visible in the LCD when you first power it up.......will be either 2.5, 2.6, 2.7 or 2.8. (I managed to acquire one or more of each version.)

Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow