How are my flow limits?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
danb1996
Posts: 9
Joined: Mon Mar 18, 2019 10:08 pm

How are my flow limits?

Post by danb1996 » Sat Sep 18, 2021 9:32 am

Hi guys,

I would like to know if my flow limits are normal.

I seem to have a lot of them, but I'm not sure if I should worry about them.
Attachments
Screen Shot 2021-09-18 at 11.26.12 PM.png
Screen Shot 2021-09-18 at 11.26.12 PM.png (488.53 KiB) Viewed 732 times

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: How are my flow limits?

Post by jimbud » Sat Sep 18, 2021 9:51 am

Do you have allergies this time of year?

Allergies will cause flow limits.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: How are my flow limits?

Post by Julie » Sat Sep 18, 2021 10:06 am

Your stuff looks pretty good, though I'm curious why you're on a fixed vs auto setting and why it's so low to begin with...?

PS ditch the calendar... not needed.

User avatar
Pugsy
Posts: 65209
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How are my flow limits?

Post by Pugsy » Sat Sep 18, 2021 10:13 am

Elite model....can't do auto adjusting mode

As for the 6 cm fixed pressure....it seems to be doing a very decent job of preventing OA/hypones. Not everyone needs higher pressures. There are people who find that 4 cm pressure can do a fine job holding the airway open and they are quite comfortable with it and there is simply no need to use more.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65209
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How are my flow limits?

Post by Pugsy » Sat Sep 18, 2021 10:20 am

danb1996 wrote:
Sat Sep 18, 2021 9:32 am
I would like to know if my flow limits are normal.

I seem to have a lot of them, but I'm not sure if I should worry about them.
Not exactly "normal" but not horribly horrible either.
As for how worrisome it might be a lot would depend on how you are sleeping and feeling in general.


Now here is a horribly horrible FL graph.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

danb1996
Posts: 9
Joined: Mon Mar 18, 2019 10:08 pm

Re: How are my flow limits?

Post by danb1996 » Sat Sep 18, 2021 10:33 am

Pugsy wrote:
Sat Sep 18, 2021 10:20 am
danb1996 wrote:
Sat Sep 18, 2021 9:32 am
I would like to know if my flow limits are normal.

I seem to have a lot of them, but I'm not sure if I should worry about them.
Not exactly "normal" but not horribly horrible either.
As for how worrisome it might be a lot would depend on how you are sleeping and feeling in general.


Now here is a horribly horrible FL graph.

Image
Would increasing EPR improve the flow limits? The epr was 3.

Here is another night with the EPR set at 1.
Attachments
Screen Shot 2021-09-18 at 11.33.46 PM.png
Screen Shot 2021-09-18 at 11.33.46 PM.png (476.03 KiB) Viewed 695 times

User avatar
Pugsy
Posts: 65209
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How are my flow limits?

Post by Pugsy » Sat Sep 18, 2021 11:00 am

You can't increase EPR above 3 with your machine. You would need a totally different machine.

Sometimes adding or increasing EPR does help reduce FLs but a lot depends on what is causing the FLs.

Do you have much nasal congestion? This is critical to know because increased FL activity can be a by product of nasal congestion and if nasal congestion is the culprit no amount of changes in settings will reduce the FLs. Instead people have to treat the nasal congestion in the normal ways we deal with nasal congestion.

Now you could try increasing your pressure to 7 or 8 because right now even though the EPR setting is 3 you are only getting 2 cm since the machine can't drop below 4 cm.....so with your 6 cm pressure it can only drop to 4 cm.
That experiment might help reduce the FLs from several fronts....let EPR maybe help and also increase IPAP (inhale) which might help reduce the FLs.
Again...assuming not related to nasal congestion.

Easy experiment to try if you wish.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How are my flow limits?

Post by palerider » Sat Sep 18, 2021 4:31 pm

danb1996 wrote:
Sat Sep 18, 2021 10:33 am
Would increasing EPR improve the flow limits? The epr was 3.

Here is another night with the EPR set at 1.
No (as Pugsy said) but raising your minimum might, since it's criminally low.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.