Technical question about Resmed S9 mask settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Benicetoothers
Posts: 11
Joined: Sat Aug 07, 2021 2:12 pm

Technical question about Resmed S9 mask settings

Post by Benicetoothers » Sat Aug 07, 2021 2:29 pm

Hello,

I am new to this board, but not new to CPAP. I have been on CPAP since 2008 and have always used a full face mask because I was a mouth breather. I had a badly deviated nasal septum with enlarged turbinates.

Recently, I had a septoplasty with turbinate reduction surgery. And I was cleared to go back to APAP/CPAP about a week ago. I bought the only nasal mask I am familiar with, a Resmed nasal Activa. Not the Activa LT, the original Mirage Activa. I have been using two Resmed S9 APAPs for years with a Resmed Mirage Quatro full face mask. I know the mask settings in the S9 for that mask. However, I dont know the mask setting in the Resmed S9 APAP for the older Activa nasal mask.

I know its either "Nasal" or "Nasal Ultra."

Do any folks know the mask setting for the original Resmed Mirage Activa in the S9 machines? Does it matter?

Thank you for your help.

Sincerely,

Rick
Try to be nice to others.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Technical question about Resmed S9 mask settings

Post by palerider » Sat Aug 07, 2021 5:16 pm

Benicetoothers wrote:
Sat Aug 07, 2021 2:29 pm
Hello,

I am new to this board, but not new to CPAP. I have been on CPAP since 2008 and have always used a full face mask because I was a mouth breather. I had a badly deviated nasal septum with enlarged turbinates.

Recently, I had a septoplasty with turbinate reduction surgery. And I was cleared to go back to APAP/CPAP about a week ago. I bought the only nasal mask I am familiar with, a Resmed nasal Activa. Not the Activa LT, the original Mirage Activa. I have been using two Resmed S9 APAPs for years with a Resmed Mirage Quatro full face mask. I know the mask settings in the S9 for that mask. However, I dont know the mask setting in the Resmed S9 APAP for the older Activa nasal mask.

I know its either "Nasal" or "Nasal Ultra."

Do any folks know the mask setting for the original Resmed Mirage Activa in the S9 machines? Does it matter?

Thank you for your help.

Sincerely,

Rick
Nasal, not ultra, that was reserved for like one mask.

here's the chart, just search for activa:
https://document.resmed.com/en-us/docum ... st_row.pdf

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11321
Joined: Fri Jun 19, 2015 4:36 pm

Re: Technical question about Resmed S9 mask settings

Post by zonker » Sat Aug 07, 2021 6:57 pm

palerider wrote:
Sat Aug 07, 2021 5:16 pm

Nasal, not ultra, that was reserved for like one mask.

here's the chart, just search for activa:
https://document.resmed.com/en-us/docum ... st_row.pdf
oh good work, sir.

i'm tucking this away for future use.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Technical question about Resmed S9 mask settings

Post by palerider » Sat Aug 07, 2021 7:31 pm

zonker wrote:
Sat Aug 07, 2021 6:57 pm
palerider wrote:
Sat Aug 07, 2021 5:16 pm

Nasal, not ultra, that was reserved for like one mask.

here's the chart, just search for activa:
https://document.resmed.com/en-us/docum ... st_row.pdf
oh good work, sir.

i'm tucking this away for future use.
You're most welcome.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Benicetoothers
Posts: 11
Joined: Sat Aug 07, 2021 2:12 pm

Re: Technical question about Resmed S9 mask settings

Post by Benicetoothers » Sun Aug 08, 2021 2:23 pm

Thank you very much for the official Resmed chart. I thought it was the nasal setting, but was not exactly sure. You were way more helpful than my brick and mortar DME recently. They told me to put it on any setting and it would work and I knew that was not correct.

Sincerely,

Rick


palerider wrote:
Sat Aug 07, 2021 5:16 pm
Benicetoothers wrote:
Sat Aug 07, 2021 2:29 pm
Nasal, not ultra, that was reserved for like one mask.

here's the chart, just search for activa:
https://document.resmed.com/en-us/docum ... st_row.pdf
Try to be nice to others.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Technical question about Resmed S9 mask settings

Post by palerider » Sun Aug 08, 2021 2:50 pm

Benicetoothers wrote:
Sun Aug 08, 2021 2:23 pm
Thank you very much for the official Resmed chart. I thought it was the nasal setting, but was not exactly sure. You were way more helpful than my brick and mortar DME recently. They told me to put it on any setting and it would work and I knew that was not correct.
Well, the differences in the mask settings are subtle, I've measured the pressures with a dual manometer between the machine and mask, and it didn't make much more than 1cm of difference, So they're sort of correct, it doesn't make much difference.

The wrong setting can keep the smartstart from turning the machine off when you take off your mask, though. When I first switched from a FFM to a nasal mask, I thought the machine had broken, because it wouldn't stop running, until I changed the mask setting. :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Dog Slobber
Posts: 4229
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Technical question about Resmed S9 mask settings

Post by Dog Slobber » Sun Aug 08, 2021 3:29 pm

palerider wrote:
Sun Aug 08, 2021 2:50 pm
The wrong setting can keep the smartstart from turning the machine off when you take off your mask, though. When I first switched from a FFM to a nasal mask, I thought the machine had broken, because it wouldn't stop running, until I changed the mask setting. :)
The Mask Setting discussion has come up a few times and we'll often see commentary about it being used for subtle pressure adjustment and possibly leak calculation.

This is the first time I've seen an association between SmartStart not shutting down properly and incorrect Mask Type setting. It makes sense.

You've been holding out again, haven't you.

*EDIT*

Holy Crap, you have posted about it before.
palerider wrote:
Sat Mar 30, 2019 1:51 pm

SmartStart should turn the machine off within a few seconds.

The only time it didn't work for me was when I first switched to nasal pillows, but didn't think to change the 'mask type' to pillow. The added restriction of the pillows tricked the machine into thinking it was just a bad leak, and not a mask off... changing the settings let the machine know to compensate for the added airflow restriction that the pillow mask creates.

Now, if you're using pillows, and the machine is set right, but you've got quite low pressures, the auto-off may not work for you. You could pop the mask off of the main hose and that should do it though.
palerider wrote:
Mon Jul 15, 2019 11:35 am
You change the shut off threshold by setting the mask type, full face is less sensitive and pillows is more sensitive.
palerider wrote:
Thu Oct 11, 2018 8:44 pm
When I first switched to the P10, I was disappointed that the machine didn't turn off when I removed the mask... then I remembered I'd forgotten to change the mask setting... since then, it's worked fine.
palerider wrote:
Tue Jul 17, 2018 1:56 pm
I've got a S9, and when I first switched to the P10, first time I took the mask off, the machine didn't shut off... I was disappointed, because I love the smartstart auto on/off... then I remembered I hadn't changed the mask type... I set it to 'pillows', and since then, it's shut off within a few seconds, every time.
palerider wrote:
Sun Nov 20, 2016 4:08 pm
there are differing opinions on what exactly that function does, but part of what it does is set thresholds for when the machine recognizes your mask is off, and shuts down automatically if smartstart is enabled. I believe that it also affects the pressure compensation algorithm to account for the different types of masks (full face masks have lower flow resistance than pillow masks, etc). the effects are minor.

best to just set the mask type correctly.

User avatar
Pugsy
Posts: 65112
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Technical question about Resmed S9 mask settings

Post by Pugsy » Sun Aug 08, 2021 5:17 pm

Dog Slobber wrote:
Sun Aug 08, 2021 3:29 pm
This is the first time I've seen an association between SmartStart not shutting down properly and incorrect Mask Type setting. It makes sense.
Mask setting isn't always the reason for SmartStart not working properly. Though it is a potential reason in some instances.
My ResMed machines won't auto off properly with the P10 nasal pillow mask (I have tested it at 10 cm pressure and it still won't work right) and I for sure have the mask setting correct.
But they do auto off properly for the Bleep. Go figure that one.
So if someone says SmartStart auto off isn't working like it should...mask setting is first thing to try but it's not an automatic given that it is the culprit.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11321
Joined: Fri Jun 19, 2015 4:36 pm

Re: Technical question about Resmed S9 mask settings

Post by zonker » Sun Aug 08, 2021 5:23 pm

Dog Slobber wrote:
Sun Aug 08, 2021 3:29 pm

You've been holding out again, haven't you.

*EDIT*

Holy Crap, you have posted about it before.

not possible. everyone knows that palerider never helps anyone and is only here to argue!
:lol: :lol: :lol:
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Technical question about Resmed S9 mask settings

Post by palerider » Sun Aug 08, 2021 9:44 pm

Dog Slobber wrote:
Sun Aug 08, 2021 3:29 pm
Holy Crap, you have posted about it before.
Oh sir, you're goooood ;)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.