What off the shelf portable power source for air sense 10?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Jake00
Posts: 22
Joined: Sat Dec 26, 2015 8:26 pm

What off the shelf portable power source for air sense 10?

Post by Jake00 » Sat Aug 07, 2021 6:34 am

Awhile ago I was trying to configure some Milwaukee batteries to run my cpap.
While I still want to make that work, I’m looking for an off the shelf solution 3-5 nights would be ideal.

_________________
MachineMask

User avatar
ChicagoGranny
Posts: 15419
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: What off the shelf portable power source for air sense 10?

Post by ChicagoGranny » Sat Aug 07, 2021 6:48 am

Jake00 wrote:
Sat Aug 07, 2021 6:34 am
3-5 nights would be ideal.
You're going to need a big mobility scooter battery, a ResMed converter and a charger for the battery. No humidifier usage.

Study the ResMed guide on batteries - https://document.resmed.com/en-us/docum ... lo_eng.pdf

Jake00
Posts: 22
Joined: Sat Dec 26, 2015 8:26 pm

Re: What off the shelf portable power source for air sense 10?

Post by Jake00 » Sat Aug 07, 2021 8:34 am


_________________
MachineMask

Tony415
Posts: 13
Joined: Thu Jul 29, 2021 2:28 pm

Re: What off the shelf portable power source for air sense 10?

Post by Tony415 » Sat Aug 07, 2021 1:21 pm

Jake00 wrote:
Sat Aug 07, 2021 6:34 am
Awhile ago I was trying to configure some Milwaukee batteries to run my cpap.
While I still want to make that work, I’m looking for an off the shelf solution 3-5 nights would be ideal.
I would get a 50Ah LiFePO4 Battery Deep cycle lithium iron battery and a 12volt DC-DC converter for your CPAP machine. Those tool battery is too small capacity for 30-50Ah use for 3-5 nights with humidifier on level 4 and a correct DC-DC converter is needed.

User avatar
ChicagoGranny
Posts: 15419
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: What off the shelf portable power source for air sense 10?

Post by ChicagoGranny » Sat Aug 07, 2021 4:14 pm

Jake00 wrote:
Sat Aug 07, 2021 8:34 am
Thinking about something like this https://www.amazon.com/ROCKPALS-Portabl ... merReviews
The specs say it is 500 watt hours. 500 watt hours at 24 volts = 21 amp hours (https://learnmetrics.com/watt-hours-to-amp-hours/)

Check the ResMed guide in the link, and you will see 21 AH won't do what you want it to do.

User avatar
ChicagoGranny
Posts: 15419
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: What off the shelf portable power source for air sense 10?

Post by ChicagoGranny » Sat Aug 07, 2021 4:22 pm

Tony415 wrote:
Sat Aug 07, 2021 1:21 pm
I would get a 50Ah LiFePO4 Battery Deep cycle lithium iron battery
That battery will cost him close to $450.

He can get a 55AH SLA AGM battery for around $170. Much better choice.

He should still calculate his AH needs from the ResMed guide.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What off the shelf portable power source for air sense 10?

Post by palerider » Sat Aug 07, 2021 5:01 pm

Jake00 wrote:
Sat Aug 07, 2021 8:34 am
Thinking about something like this https://www.amazon.com/ROCKPALS-Portabl ... merReviews
That will give you just a *little* more than a 35AH lead acid battery, (which costs about $70 dollars, and will still require the Resmed DC adapter.

So, sure, if you've got a lot of money to throw away, go for it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What off the shelf portable power source for air sense 10?

Post by palerider » Sat Aug 07, 2021 5:06 pm

ChicagoGranny wrote:
Sat Aug 07, 2021 4:14 pm
Jake00 wrote:
Sat Aug 07, 2021 8:34 am
Thinking about something like this https://www.amazon.com/ROCKPALS-Portabl ... merReviews
The specs say it is 500 watt hours. 500 watt hours at 24 volts = 21 amp hours (https://learnmetrics.com/watt-hours-to-amp-hours/)

Check the ResMed guide in the link, and you will see 21 AH won't do what you want it to do.
Well, watthours are watthours. a 35ah lead acid scooter battery is about 420wh (for $70) and that'll last several days.

What you need to do is convert whatever the battery says in AH into WH, getting past the misleading marketing wank that lithium battery packs have, like "50,000mah" of course, the 'm' means 1/1000th, so 50,000mah is 50ah, but that's usually at 3.6volts, so that really impressive 50,000mah works out to a measly 180wh, which *might* get you through two nights. (80-90wh per night).

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Dog Slobber
Posts: 4270
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: What off the shelf portable power source for air sense 10?

Post by Dog Slobber » Sat Aug 07, 2021 5:08 pm

ChicagoGranny wrote:
Sat Aug 07, 2021 4:14 pm
Jake00 wrote:
Sat Aug 07, 2021 8:34 am
Thinking about something like this https://www.amazon.com/ROCKPALS-Portabl ... merReviews
The specs say it is 500 watt hours. 500 watt hours at 24 volts = 21 amp hours (https://learnmetrics.com/watt-hours-to-amp-hours/)

Check the ResMed guide in the link, and you will see 21 AH won't do what you want it to do.
Although the ResMed is a 24V device, that's after the conversion from 12V.

Since the the actual draw from the battery is 12v, would the Amp Hour conversion be 42 AH?
Battery Backup: EcoFlow Delta 2

User avatar
ChicagoGranny
Posts: 15419
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: What off the shelf portable power source for air sense 10?

Post by ChicagoGranny » Sat Aug 07, 2021 5:38 pm

Dog Slobber wrote:
Sat Aug 07, 2021 5:08 pm
Since the the actual draw from the battery is 12v, would the Amp Hour conversion be 42 AH?
Yes, according to this converter - https://learnmetrics.com/watt-hours-to-amp-hours/

Thank you for the correction.
palerider wrote:
Sat Aug 07, 2021 5:06 pm
What you need to do
Agree now?

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: What off the shelf portable power source for air sense 10?

Post by JLROhio » Sat Aug 07, 2021 6:18 pm

What's with all the math? When I signed up here - there was a promise (I think) of No Math!

;-)
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What off the shelf portable power source for air sense 10?

Post by palerider » Sat Aug 07, 2021 6:23 pm

JLROhio wrote:
Sat Aug 07, 2021 6:18 pm
What's with all the math? When I signed up here - there was a promise (I think) of No Math!

;-)
*grabs eraser* show me where you saw that.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: What off the shelf portable power source for air sense 10?

Post by SleepGeek » Sat Aug 07, 2021 6:24 pm

Not only is that thing overpriced they mention the battery is not replaceable. The video in the link shows him using AC and using the humidifier - I would prefer to use the correct DC cable for the cpap. It also mentions the recharge cable that comes with it is an AC charger while it has other options to charge those are extra money.

I would prefer any jump starter to that thing. But have use this one for some time now
https://www.cpaptalk.com/viewtopic.php? ... 50#p434501

Tony415
Posts: 13
Joined: Thu Jul 29, 2021 2:28 pm

Re: What off the shelf portable power source for air sense 10?

Post by Tony415 » Sun Aug 08, 2021 2:05 am

Yeah, limited budget is difficult.
I would go with a car battery. Or even just use it from the battery in the vehicle he drives to the camp. Just start the vehicle every day to charge the battery. He does not need to buy extra battery just for going camping once awhile.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: What off the shelf portable power source for air sense 10?

Post by Goofproof » Sun Aug 08, 2021 8:29 am

You might opt for a car battery, but you would be choosing the poorest option. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire