problems Exhaling

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
rubystandingdeer
Posts: 15
Joined: Fri Aug 06, 2021 8:00 am

problems Exhaling

Post by rubystandingdeer » Fri Aug 06, 2021 8:04 am

:? I have real trouble exhaling and panic. I just got the air fit N20. It leaks, it is painful and not meant for a side sleeper. What else is there that will help me exhale?

User avatar
LSAT
Posts: 13404
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: problems Exhaling

Post by LSAT » Fri Aug 06, 2021 9:11 am

The exhale may have nothing to do with the mask. Your machine has an exhale relief setting that may not be on or may not be set high enough. If you are using a Resmed machine, check to see if the EPR (Exhale Pressure Relief) is on and what it is set at.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: problems Exhaling

Post by Julie » Fri Aug 06, 2021 9:54 am

And Philips machines have something called C- or Aflex.

User avatar
Pugsy
Posts: 65334
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: problems Exhaling

Post by Pugsy » Fri Aug 06, 2021 10:26 am

Impossible to offer anything more than general information without knowing...
What machine (brand and model) you are using????
What pressure you are using?
Are you using the ramp feature?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11453
Joined: Fri Jun 19, 2015 4:36 pm

Re: problems Exhaling

Post by zonker » Fri Aug 06, 2021 10:57 am

rubystandingdeer wrote:
Fri Aug 06, 2021 8:04 am
:? I have real trouble exhaling and panic. I just got the air fit N20. It leaks, it is painful and not meant for a side sleeper. What else is there that will help me exhale?
welcome to the zoo!

we are each different and we don't necessarily use the same brand of cpap and certainly not the same mask. so that's why we need to know which it is you have.

if you have filled out your equipment profile, i'm sorry but it doesn't show up like it once did.

so let us know!

good luck.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

rubystandingdeer
Posts: 15
Joined: Fri Aug 06, 2021 8:00 am

Re: problems Exhaling

Post by rubystandingdeer » Fri Aug 06, 2021 11:07 am

I just actually found this one on line that looks like it might be the solution

https://www.1800cpap.com/resmed-swift-f ... ize=Medium

User avatar
ChicagoGranny
Posts: 15502
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: problems Exhaling

Post by ChicagoGranny » Fri Aug 06, 2021 2:50 pm

rubystandingdeer wrote:
Fri Aug 06, 2021 8:04 am
:? I have real trouble exhaling and panic. I just got the air fit N20. It leaks, it is painful and not meant for a side sleeper. What else is there that will help me exhale?
This is a common problem for newbies. In most cases, the person is concentrating on breathing. If they can relax their mind, divert their attention from breathing, and let their autonomous (automatic nervous) system take control of breathing, the problem goes away. Once they are asleep, they breathe gently and rhythmically. After all, we breathe all day without thinking about it. No need to think about it when falling asleep.

By all means, take advantage of the machine's exhalation relief feature that others have mentioned.

rubystandingdeer
Posts: 15
Joined: Fri Aug 06, 2021 8:00 am

Re: problems Exhaling

Post by rubystandingdeer » Sat Aug 07, 2021 8:43 am

Pugsy wrote:
Fri Aug 06, 2021 10:26 am
Impossible to offer anything more than general information without knowing...
What machine (brand and model) you are using????
What pressure you are using?
Are you using the ramp feature?

Resmed 10 airflow. Just got it
I cannot even see the airpressue and I did try the ramp, but it did not change anything. :(
I appear to locked out.

rubystandingdeer
Posts: 15
Joined: Fri Aug 06, 2021 8:00 am

Re: problems Exhaling

Post by rubystandingdeer » Sat Aug 07, 2021 8:47 am

I could maybe tolerate it better if the mask did not cause to much pain. I need to what mask is better

User avatar
Pugsy
Posts: 65334
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: problems Exhaling

Post by Pugsy » Sat Aug 07, 2021 8:52 am

rubystandingdeer wrote:
Sat Aug 07, 2021 8:47 am
I could maybe tolerate it better if the mask did not cause to much pain. I need to what mask is better
If any mask is causing a lot of pain then it is either fitted incorrectly or the wrong size.
Most common error made with most masks.....too tight.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

metaltech
Posts: 21
Joined: Tue Oct 01, 2019 2:52 pm

Re: problems Exhaling

Post by metaltech » Mon Aug 09, 2021 5:14 pm

When you did your sleep study did they test you for bipap? I can not use a cpap. I can not exhale. The bipap has a lower exhalation pressure setting. I exhale at 1/2 of my inhalation pressure. I felt like I was suffocating. It raised anxiety and panic. After they tried bipap I did better

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Bleep DreamPort CPAP Mask Solution
Humidifier: DreamStation Heated Humidifier
Additional Comments: Dream station bipap 30 s/t . Not listed in equipment. Using Oscar software

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: problems Exhaling

Post by babydinosnoreless » Mon Aug 09, 2021 5:37 pm

I use the N20 and am a lifelong side sleeper. If you have any specific questions on the N20 I would be glad to try and help.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: problems Exhaling

Post by Miss Emerita » Mon Aug 09, 2021 6:27 pm

I hope the Swift FX will work for you. To see your pressure settings, enter the clinical menu by holding don the round button and the home button at the same time. It takes a few seconds for it to appear.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

rubystandingdeer
Posts: 15
Joined: Fri Aug 06, 2021 8:00 am

Re: problems Exhaling

Post by rubystandingdeer » Wed Aug 11, 2021 9:12 am

Pugsy wrote:
Fri Aug 06, 2021 10:26 am
Impossible to offer anything more than general information without knowing...
What machine (brand and model) you are using????
What pressure you are using?
Are you using the ramp feature?
I have a resmed 10
I have the ramp set at 30 mins
the pressure is set at 4-8
Does this help?

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: problems Exhaling

Post by Julie » Wed Aug 11, 2021 10:57 am

Resmed 10 what? Auto?

Raise the min. setting from 4 to 6 or 7 - no one is able to breathe at 4 (the machine's default low).

Stop using the ramp altogether... you''re spending 30 mins 'ramping up' from the bottom (as above) to... a couple of cms higher at best, and for all that time you're not getting adequate therapy.
Most of us don't use it at all beyond the first week.