Apap is hissing and pressure feels too low
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Apap is hissing and pressure feels too low
I noticed it two nights ago when going to bed, my apap made more noise than normal, I can best describe it as a hissing noise coming out of the machine. I also felt like not being able to breathe properly, the pressure felt much lower than it usually is, felt like I was suffocating.
I checked the mask seal, seems to be fine. Changed the hose, no difference. Changed water chamber, that didn't help either.
What could be the reason for this? Two nights of this and I am feeling really desperate and overly tired and exhausted.
I checked the mask seal, seems to be fine. Changed the hose, no difference. Changed water chamber, that didn't help either.
What could be the reason for this? Two nights of this and I am feeling really desperate and overly tired and exhausted.
Airsense 10 For Her
F20 full face mask
F20 full face mask
Re: Apap is hissing and pressure feels too low
you may not have noticed, but forum software doesn't tell us anymore what machine and mask each user has.strawberryfields wrote: ↑Wed Aug 04, 2021 6:38 amI noticed it two nights ago when going to bed, my apap made more noise than normal, I can best describe it as a hissing noise coming out of the machine. I also felt like not being able to breathe properly, the pressure felt much lower than it usually is, felt like I was suffocating.
I checked the mask seal, seems to be fine. Changed the hose, no difference. Changed water chamber, that didn't help either.
What could be the reason for this? Two nights of this and I am feeling really desperate and overly tired and exhausted.
pls let us know so maybe someone will have a suggestion.
good luck!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Apap is hissing and pressure feels too low
Anytime anyone has a problem with a perceived difference in pressure from what they feel they should be getting or the setting or what they got in the past they need to verify the pressure actually being delivered at the mask by using a manometer.
Manometers can be store bought or a DIY homemade manometer.... (sorry don't have the homemade link handy but others do and I suspect will come along shortly and offer it).
Also it would be extremely useful for OP to tell us exactly what brand and model machine is being used along with maybe run hours on the machine.
For long term members here....you may have filled out the equipment profile in the past but it is no longer showing up (broken link that hasn't been fixed in like forever)....so don't assume we know what you are using since that profile is now absent.
Put your equipment in your signature line like I have mine. That is working (at least for now).
Now assuming AirSense/AirCurve model.....
As for hissing...have no idea but if water chamber connection is not the issue or hose or mask...the only thing at the machine level I can think of...and it's a very remote long shot...the air outlet thing might be messed up or seal broken somehow.
This part...which I hear is easily changed.
https://www.cpap.com/productpage/resmed ... air-outlet
Manometers can be store bought or a DIY homemade manometer.... (sorry don't have the homemade link handy but others do and I suspect will come along shortly and offer it).
Also it would be extremely useful for OP to tell us exactly what brand and model machine is being used along with maybe run hours on the machine.
For long term members here....you may have filled out the equipment profile in the past but it is no longer showing up (broken link that hasn't been fixed in like forever)....so don't assume we know what you are using since that profile is now absent.
Put your equipment in your signature line like I have mine. That is working (at least for now).
Now assuming AirSense/AirCurve model.....
As for hissing...have no idea but if water chamber connection is not the issue or hose or mask...the only thing at the machine level I can think of...and it's a very remote long shot...the air outlet thing might be messed up or seal broken somehow.
This part...which I hear is easily changed.
https://www.cpap.com/productpage/resmed ... air-outlet
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
Oh I must have missed this! Thank you.zonker wrote: ↑Wed Aug 04, 2021 8:55 pmyou may not have noticed, but forum software doesn't tell us anymore what machine and mask each user has.strawberryfields wrote: ↑Wed Aug 04, 2021 6:38 amI noticed it two nights ago when going to bed, my apap made more noise than normal, I can best describe it as a hissing noise coming out of the machine. I also felt like not being able to breathe properly, the pressure felt much lower than it usually is, felt like I was suffocating.
I checked the mask seal, seems to be fine. Changed the hose, no difference. Changed water chamber, that didn't help either.
What could be the reason for this? Two nights of this and I am feeling really desperate and overly tired and exhausted.
pls let us know so maybe someone will have a suggestion.
good luck!
I am using the Airsense 10 Autoset For Her machine and a F20 full face mask from resmed.
Airsense 10 For Her
F20 full face mask
F20 full face mask
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
Thank you Pugsy!Pugsy wrote: ↑Wed Aug 04, 2021 9:14 pmAnytime anyone has a problem with a perceived difference in pressure from what they feel they should be getting or the setting or what they got in the past they need to verify the pressure actually being delivered at the mask by using a manometer.
Manometers can be store bought or a DIY homemade manometer.... (sorry don't have the homemade link handy but others do and I suspect will come along shortly and offer it).
Also it would be extremely useful for OP to tell us exactly what brand and model machine is being used along with maybe run hours on the machine.
For long term members here....you may have filled out the equipment profile in the past but it is no longer showing up (broken link that hasn't been fixed in like forever)....so don't assume we know what you are using since that profile is now absent.
Put your equipment in your signature line like I have mine. That is working (at least for now).
Now assuming AirSense/AirCurve model.....
As for hissing...have no idea but if water chamber connection is not the issue or hose or mask...the only thing at the machine level I can think of...and it's a very remote long shot...the air outlet thing might be messed up or seal broken somehow.
This part...which I hear is easily changed.
https://www.cpap.com/productpage/resmed ... air-outlet
I will look into getting a manometer.
I am using the Airsense 10 for Her machine from Resmed, and it has been in usage for 7457 hours.
I will put my machine and mask in the signature, thanks!
Okay, I was wondering about the outlet part as well, I had no idea it was possble to change it out by myself.
Airsense 10 For Her
F20 full face mask
F20 full face mask
- ChicagoGranny
- Posts: 14511
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Apap is hissing and pressure feels too low
What is the machine leak level? You can view this number on the machine display or in OSCAR. Yes, it is also a good idea to check the pressure with a manometer. Your local DME can do this for you, or you can buy your own manometer.strawberryfields wrote: ↑Wed Aug 04, 2021 6:38 amApap is hissing and pressure feels too lowQuote strawberryfields
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
The max leak rate is shown to be 4.80.ChicagoGranny wrote: ↑Thu Aug 05, 2021 8:26 amWhat is the machine leak level? You can view this number on the machine display or in OSCAR. Yes, it is also a good idea to check the pressure with a manometer. Your local DME can do this for you, or you can buy your own manometer.strawberryfields wrote: ↑Wed Aug 04, 2021 6:38 amApap is hissing and pressure feels too lowQuote strawberryfields
Airsense 10 For Her
F20 full face mask
F20 full face mask
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
So something strange happened. The last 2 nights I have been using my backup machine, which is the regular Airsense 10. The machine is quite new, I haven't used it a lot so far (because I can't sleep as well with it as with the For Her model). Anyways, the first night was normal and the second night the same thing happened as with the For Her machine: hissing sound and pressure feels way too low/too weak.
I am really confused. I think the sound mainly comes from the water chamber, but I can't see anything wrong there.
I am really confused. I think the sound mainly comes from the water chamber, but I can't see anything wrong there.
Airsense 10 For Her
F20 full face mask
F20 full face mask
Re: Apap is hissing and pressure feels too low
There could be a 'slit' in your hose somewhere causing the noise - you should look for it (slowly run water through) and duct tape a cut if you find one.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
I will try that, thank you.
Airsense 10 For Her
F20 full face mask
F20 full face mask
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
Update: I tried something that I didn't think would work but somehow the pressure seems more normal now?
I switched the power cord. I tried the Apap two times with power cord 1, and two times with power cord 2, and the pressure problem seems to only occur with power cord 1. I hope this really is the solution because I am desperate for some more normal sleep again.
I switched the power cord. I tried the Apap two times with power cord 1, and two times with power cord 2, and the pressure problem seems to only occur with power cord 1. I hope this really is the solution because I am desperate for some more normal sleep again.
Airsense 10 For Her
F20 full face mask
F20 full face mask
Re: Apap is hissing and pressure feels too low
let us know, please!strawberryfields wrote: ↑Fri Aug 06, 2021 8:26 amUpdate: I tried something that I didn't think would work but somehow the pressure seems more normal now?
I switched the power cord. I tried the Apap two times with power cord 1, and two times with power cord 2, and the pressure problem seems to only occur with power cord 1. I hope this really is the solution because I am desperate for some more normal sleep again.
i don't know nuthin' 'bout electricity so unable to comment.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- strawberryfields
- Posts: 50
- Joined: Thu Mar 07, 2019 7:10 am
Re: Apap is hissing and pressure feels too low
Unfortunately I was wrong I think I just really wanted it to work.zonker wrote: ↑Fri Aug 06, 2021 11:29 amlet us know, please!strawberryfields wrote: ↑Fri Aug 06, 2021 8:26 amUpdate: I tried something that I didn't think would work but somehow the pressure seems more normal now?
I switched the power cord. I tried the Apap two times with power cord 1, and two times with power cord 2, and the pressure problem seems to only occur with power cord 1. I hope this really is the solution because I am desperate for some more normal sleep again.
i don't know nuthin' 'bout electricity so unable to comment.
I also checked the tube, there don't seem to be any holes in there.
If I don't figure it out by Monday, I'm gonna call remsed and send both machines to repair.
Airsense 10 For Her
F20 full face mask
F20 full face mask
- ChicagoGranny
- Posts: 14511
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Apap is hissing and pressure feels too low
Which country? In the US, warranty issues are handled by the equipment suppliers, not machine manufacturers. Nothing to get hung about.strawberryfields wrote: ↑Sat Aug 07, 2021 4:05 amIf I don't figure it out by Monday, I'm gonna call remsed and send both machines to repair.
Re: Apap is hissing and pressure feels too low
ChicagoGranny wrote: ↑Sat Aug 07, 2021 12:37 pmWhich country? In the US, warranty issues are handled by the equipment suppliers, not machine manufacturers. Nothing to get hung about.strawberryfields wrote: ↑Sat Aug 07, 2021 4:05 amIf I don't figure it out by Monday, I'm gonna call remsed and send both machines to repair.
did that come unbidden to your mind or did you have to think about it?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg