U1 AGM battery indoor safety
-
- Posts: 4
- Joined: Fri Jul 16, 2021 5:03 pm
U1 AGM battery indoor safety
I couldn't tell apart a positive and negative terminal. I know very little about batteries. I need a somewhat reliable backup in the event of a power outage , but also something safe to use indoors. I am looking at getting an AGM U1 battery. I wanted to know if it was safe to not only store the battery in my house, but also use it and charge it?
Re: U1 AGM battery indoor safety
Can you PM "Capn Loki', our resident expert on camping, batteries, etc. etc? He might not have seen your note yet.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: U1 AGM battery indoor safety
If it's not marked + or - don't buy it as they are too stupid to sell products. Jim
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
Re: U1 AGM battery indoor safety
One has a + and one has a - often the positive is red.
yesTeamArliss wrote: ↑Fri Jul 16, 2021 5:05 pmI know very little about batteries. I need a somewhat reliable backup in the event of a power outage , but also something safe to use indoors. I am looking at getting an AGM U1 battery. I wanted to know if it was safe to not only store the battery in my house, but also use it and charge it?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: U1 AGM battery indoor safety
This is pretty fool proof. Comes completely assembled. All you need is the DC power cord for your cpap and to charge the battery when you get it, which you should do before using any battery.
It will do everything you are looking to do. Indoor use = no prob. Use while charging = no prob.
viewtopic/t64943/Another-Battery-Option ... ry+option
It will do everything you are looking to do. Indoor use = no prob. Use while charging = no prob.
viewtopic/t64943/Another-Battery-Option ... ry+option
Paraphrasing Ghandi - be the change you wish to see in the world.
Paraphrasing JohnnyGoodman - Beware the person who wishes to legislate the morality and the behavior of his neighbor.
https://cutt.ly/GbE7ltf
Paraphrasing JohnnyGoodman - Beware the person who wishes to legislate the morality and the behavior of his neighbor.
https://cutt.ly/GbE7ltf
- loggerhead12
- Posts: 560
- Joined: Fri Sep 01, 2017 4:00 pm
Re: U1 AGM battery indoor safety
AGM batteries don't release gases while they charge. As long as you don't break them open it's perfectly safe to charge and use them inside.
I put a picture at the end of the "Choosing A Battery" thread. Two bolts on the battery, and the rest is simple plugs. You can do it.
I put a picture at the end of the "Choosing A Battery" thread. Two bolts on the battery, and the rest is simple plugs. You can do it.
-
- Posts: 4
- Joined: Fri Jul 16, 2021 5:03 pm
Re: U1 AGM battery indoor safety
Ok, so the battery I got has two sponge Bob looking metal pieces sticking out the top. Your saying I have to attach bolts to those? Also, should I charge it before I use it, or do they usually come charged?
- loggerhead12
- Posts: 560
- Joined: Fri Sep 01, 2017 4:00 pm
Re: U1 AGM battery indoor safety
If you bought a Battery Tender, attach the ring terminal lead to the battery with the nuts and bolts that came with the battery. The red ring goes to the + Sponge Bob, and the black one goes to the - Sponge Bob. Plug the other end of the lead into the Battery Tender, then the Battery Tender into the wall.TeamArliss wrote: ↑Mon Jul 19, 2021 12:17 pmOk, so the battery I got has two sponge Bob looking metal pieces sticking out the top. Your saying I have to attach bolts to those? Also, should I charge it before I use it, or do they usually come charged?
-
- Posts: 4
- Joined: Fri Jul 16, 2021 5:03 pm
Re: U1 AGM battery indoor safety
I have a Genius 5, but haven't opened it yet. I think it's battery charger comes with clamps
- loggerhead12
- Posts: 560
- Joined: Fri Sep 01, 2017 4:00 pm
Re: U1 AGM battery indoor safety
Same premise. Hopefully the NOCO will have some form of permanent attachment in addition to the clamps.TeamArliss wrote: ↑Mon Jul 19, 2021 3:03 pmI have a Genius 5, but haven't opened it yet. I think it's battery charger comes with clamps
-
- Posts: 4
- Joined: Fri Jul 16, 2021 5:03 pm
Re: U1 AGM battery indoor safety
Thanks to all who responded
Re: U1 AGM battery indoor safety
Realize these posts are a few months old but feel the need to chime in regarding safety concerns not raised. For background I've posted multiple times on this based on my experience in building CPAP battery packs based on half-U1 size agm lead-acid batteries. Added my experience to CapnLoki's extensive set of posts that you can see via the link below. Good detailed explanations across these posts on how to scope and build a battery pack.
https://www.cpaptalk.com/viewtopic/t114 ... ?start=150
While agm batteries are the safest type of lead-acid cell for use they can still be dangerous if not used correctly as they store a lot of power. Drop a wrench across the terminals of one of these (even my half-U1 size 20ah batteries) and hundreds of amps will go across the short circuit welding the wrench to the battery terminals. The keys to a safer design include:
1) Make sure your wiring is of a sufficient gauge to handle the expected current. Also, see details in the posts above, make sure the gauge will be sufficient to prevent voltage drop or you CPAP won't work. (see dialogue on voltage drop issues experienced by one user in the posts above)
2) Ensure that there is no exposed metal in your circuits that could result in an accidental short circuit. This means no use of alligator clip connectors on other than a short term emergency basis. All connections need to be covered by an insulator such as heat shrink tubing or at least a couple layers of electrical tape. Even better use off the shelf connectors, such as Anderson Poles, that also help eliminate the possibility of accidentally connecting a short circuit.
3) Install the correct level of fusing in your circuit. For 12 volt batteries this means your typical auto type fuses. Fuse to protect the circuit/wiring. For CPAPs used remotely without a humidifier the current required is often a lot less than the machine specs that are set to cover humidifier use and the maximum therapy pressure (typically 20cm). You can look these up for many machines or test with a multimeter. My devices only use between .5 and .65 amps at my therapy pressures which is why only half-U1 size batteries were chosen as these easily support two nights of therapy. Based on my wiring, and sometimes review of manufacturer DC cord specs, I've fused my CPAP connectors at just 5 amps with just one nuisance blow in 10 years.
4) Consider a batter box/bag for added safety. My half-U1 size batteries are placed in the bottom part of a plastic ammo box that is then placed in at heavy duty tool bag. For larger batteries the recommendation would be to place the battery and fusing inside the correct size battery box. Provides an added layer of protection to your battery set up. Lots of blogs/videos on how various folks have done this. Have seen one that converted a small igloo cooler.
Hope this helps.
P.S. As I sometimes use my batteries to power other devices via an inverter my design evolved into one with two fuses. A 15 amp main fuse right on the battery with a appropriate lower value fuse on a second fuse on a connector for specific uses. This varies from 3 amps for a 30 watt solar panel, 5 amps for my cpaps and 7.5 amps for my batterytender chargers. Inverter is protected my the main fuse.
https://www.cpaptalk.com/viewtopic/t114 ... ?start=150
While agm batteries are the safest type of lead-acid cell for use they can still be dangerous if not used correctly as they store a lot of power. Drop a wrench across the terminals of one of these (even my half-U1 size 20ah batteries) and hundreds of amps will go across the short circuit welding the wrench to the battery terminals. The keys to a safer design include:
1) Make sure your wiring is of a sufficient gauge to handle the expected current. Also, see details in the posts above, make sure the gauge will be sufficient to prevent voltage drop or you CPAP won't work. (see dialogue on voltage drop issues experienced by one user in the posts above)
2) Ensure that there is no exposed metal in your circuits that could result in an accidental short circuit. This means no use of alligator clip connectors on other than a short term emergency basis. All connections need to be covered by an insulator such as heat shrink tubing or at least a couple layers of electrical tape. Even better use off the shelf connectors, such as Anderson Poles, that also help eliminate the possibility of accidentally connecting a short circuit.
3) Install the correct level of fusing in your circuit. For 12 volt batteries this means your typical auto type fuses. Fuse to protect the circuit/wiring. For CPAPs used remotely without a humidifier the current required is often a lot less than the machine specs that are set to cover humidifier use and the maximum therapy pressure (typically 20cm). You can look these up for many machines or test with a multimeter. My devices only use between .5 and .65 amps at my therapy pressures which is why only half-U1 size batteries were chosen as these easily support two nights of therapy. Based on my wiring, and sometimes review of manufacturer DC cord specs, I've fused my CPAP connectors at just 5 amps with just one nuisance blow in 10 years.
4) Consider a batter box/bag for added safety. My half-U1 size batteries are placed in the bottom part of a plastic ammo box that is then placed in at heavy duty tool bag. For larger batteries the recommendation would be to place the battery and fusing inside the correct size battery box. Provides an added layer of protection to your battery set up. Lots of blogs/videos on how various folks have done this. Have seen one that converted a small igloo cooler.
Hope this helps.
P.S. As I sometimes use my batteries to power other devices via an inverter my design evolved into one with two fuses. A 15 amp main fuse right on the battery with a appropriate lower value fuse on a second fuse on a connector for specific uses. This varies from 3 amps for a 30 watt solar panel, 5 amps for my cpaps and 7.5 amps for my batterytender chargers. Inverter is protected my the main fuse.
- ChicagoGranny
- Posts: 14511
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: U1 AGM battery indoor safety
Let's make it so complicated that people just suffer without CPAP during a power outage.
Re: U1 AGM battery indoor safety
AmenChicagoGranny wrote: ↑Sat Jan 15, 2022 6:04 pmLet's make it so complicated that people just suffer without CPAP during a power outage.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
Re: U1 AGM battery indoor safety
AGM batteries will vent Hydrogen gas if they are overcharged. If you have AGM battery you need to use a modern electronically controlled charger with AGM setting to prevent doing so. It's worth noting this for anyone who was planning to charge one of these with an old school basic charger.loggerhead12 wrote: ↑Sat Jul 17, 2021 11:36 amAGM batteries don't release gases while they charge. As long as you don't break them open it's perfectly safe to charge and use them inside.
I put a picture at the end of the "Choosing A Battery" thread. Two bolts on the battery, and the rest is simple plugs. You can do it.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Oracle HC452 Oral CPAP Mask |