APEX ICH APA machine

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
braddw25
Posts: 25
Joined: Sat Aug 29, 2020 6:47 am

APEX ICH APA machine

Post by braddw25 » Thu Jul 01, 2021 3:41 pm

HI everyone,

I was wondering if anyone here has any experience with an APEC ich auto cpap machine. I bought one to use as a backup while I get my dreamstation recall issues resolved. My problem is that a pressure of 16 mm H20 works best for my therapy. This machine is set at a minimum pressure of 4 and maximum of 20. My wife woke me up last night to tell me that I was snoring VERY loudly and holding me breath. I looked at the machine and it was only providing 13 mmH2O of pressure. Is there any way to change the min pressure setting to 16 so that I can ensure I have the needed pressure level? Thanks for any info that anyone can provide.

Bradd

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: APEX ICH APA machine

Post by Julie » Thu Jul 01, 2021 3:44 pm

Never heard of your machine (where are you?), but know that most clinician manuals can be found on Apneaboard.com's website.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: APEX ICH APA machine

Post by Okie bipap » Thu Jul 01, 2021 4:22 pm

If you don't have the owners manual, you can download it here: https://www.respshop.com/manuals/iCH+AU ... Manual.pdf

Page 8 says you can enter the setting mode by means of the menu button. Once in the setting mode, therapy pressure can be adjusted.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
zonker
Posts: 11430
Joined: Fri Jun 19, 2015 4:36 pm

Re: APEX ICH APA machine

Post by zonker » Thu Jul 01, 2021 4:54 pm

Okie bipap wrote:
Thu Jul 01, 2021 4:22 pm
If you don't have the owners manual, you can download it here: https://www.respshop.com/manuals/iCH+AU ... Manual.pdf

Page 8 says you can enter the setting mode by means of the menu button. Once in the setting mode, therapy pressure can be adjusted.
well done, sir.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

braddw25
Posts: 25
Joined: Sat Aug 29, 2020 6:47 am

Re: APEX ICH APA machine

Post by braddw25 » Thu Jul 01, 2021 6:41 pm

Okie bipap wrote:
Thu Jul 01, 2021 4:22 pm
If you don't have the owners manual, you can download it here: https://www.respshop.com/manuals/iCH+AU ... Manual.pdf

Page 8 says you can enter the setting mode by means of the menu button. Once in the setting mode, therapy pressure can be adjusted.
Thanks very much for the response. I do have a manual and tried to do what it said. However, pushing the up and down arrows do not change the the therapy pressures. The min stays at 4 no matter how many times the up button is pushed. I am wondering if you have to get into the clinical mode or whatever it's called as I had to do with my dreamstation. I just don't know how to do that and was wondering if anyone else knew.

Thanks,

Bradd

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: APEX ICH APA machine

Post by chunkyfrog » Thu Jul 01, 2021 8:30 pm

Yes, pressure settings are commonly done in the "clinical" setting.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Centralpa
Posts: 1
Joined: Sat Jul 03, 2021 9:50 pm

Re: APEX ICH APA machine

Post by Centralpa » Sat Jul 03, 2021 9:53 pm

I have the same machine for the same reason. To change your settings, hold down the up and down buttons together for a few seconds. When the display starts blinking, you can change the value. Hit menu to save.

All in all a good little machine, but I'm not happy with the lack of a dc power supply. Camping with this will be impossible without an electric hookup.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: APEX ICH APA machine

Post by chunkyfrog » Sat Jul 03, 2021 11:01 pm

Centralpa wrote:
Sat Jul 03, 2021 9:53 pm
I have the same machine for the same reason. To change your settings, hold down the up and down buttons together for a few seconds. When the display starts blinking, you can change the value. Hit menu to save.

All in all a good little machine, but I'm not happy with the lack of a dc power supply. Camping with this will be impossible without an electric hookup.

Even so, it beats the crap out of sucking toxic foam all night.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her