New machine causing issues

Post by 1999jacko » Sun Jun 27, 2021 11:34 pm

Hi.
So I've been using a dreamstation for about 5 years now and for the last 4 of those things have been near perfect. By that I mean I sleep 7-8 hours every night with an ahi that is 90 percent of the time below 1. I use a dreamwear nasal mask after teaching myself to stop mouth breathing about 3 and a half years ago so all good.
Of course in the last week or so things have changed with the recall, I didn't mess around and was lucky to be in a position where I could purchase a new resmed airsense 10 as a replacement. It arrived and felt quite premium so I excitedly set it up.......well 4 nights in and things are not going well. My ahi is still very good at below 1 but.... I am in large leak territory for a chunk of the night, around 15% of the time according to Oscar and I'm almost certain it's because I'm mouth breathing. My wife says she can hear me and I'm waking several times in the night to a very dry mouth. I can't see anything particularly standing out on Oscar to explain it and certainly has to be down to be new machine as it would be an extraordinary coincidence otherwise, pressure is set between 11 and 13.5, exactly as on dreamstation. I have the EPR set to off and to be honest can't remember if I had the Philips version of EPR (AFLEX??) set to on or not.
Would appreciate any suggestions/advice as in just 4 nights I can already notice difference between how I feel when I wake each morning and through the day. Thanks for reading my ramble.

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New machine causing issues

Post by Pugsy » Mon Jun 28, 2021 12:28 am

Can you provide a recent detailed report from OSCAR for a DreamStation night and a detailed report from a ResMed night please?
It's impossible to do anything but guess as to what might be going on.
Also...exact details if at all possible...like are you waking often during the night now and you didn't before???
Gotta figure out what might have changed with the machine change.
If it is just more leaks and the leaks are disturbing sleep then we have to try to figure out if the ResMed is maybe going higher with the pressure or something.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

1999jacko
Posts: 29
Joined: Thu Feb 02, 2017 5:48 am

Re: New machine causing issues

Post by 1999jacko » Mon Jun 28, 2021 3:14 am

screenshot-20210628-090401.png
screenshot-20210628-090401.png (254.32 KiB) Viewed 1209 times
screenshot-20210628-090325.png
screenshot-20210628-090325.png (249.5 KiB) Viewed 1209 times

1999jacko
Posts: 29
Joined: Thu Feb 02, 2017 5:48 am

Re: New machine causing issues

Post by 1999jacko » Mon Jun 28, 2021 4:01 am

With regards the other questions, i would normally wake 3 to 4 times a night very briefly and quickly be back asleep. Don't think that's particularly increased these last 4 nights and as mentioned only difference being that in those awakenings now I definitely feel a dry mouth which wasn't happening before. Something is happening as I feel dreadful last few days after feeling brilliant for years. Even on very odd occasion I'd have by my standards a very high ahi (3 for me would be very high) I'd feel brilliant compared to the 4 nights I've had on the Airsense. Thanks.

bossfan
Posts: 42
Joined: Fri Sep 25, 2020 11:49 am

Re: New machine causing issues

Post by bossfan » Mon Jun 28, 2021 6:31 am

I don't want to hijack this thread with my own Oscar charts but I'm interested in this topic also. I've switched from Respironics to Resmed and I am getting the same Large Leak flagging. More leaks flagged with Resmed than Respironics. I know the Oscar reports are more informative than the Resmed Happy Face but except for one night where I got the Sad Face I am getting the Happy Face on the sleep reports.
Resmed Aircurve 10 ASV
Dreamwear Mask With Under Nose Nasal Cushion

User avatar
Pugsy
Posts: 65395
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New machine causing issues

Post by Pugsy » Mon Jun 28, 2021 7:14 am

Turn EPR on to full time and set it to 2 or 3 and see how things go.

The leaks increased for some reason but then so did the pressure variations themselves.
You were using some exhale relief on the Respironics machine and none now....looks like about 3 cm relief so try 2 or 3 cm relief on the ResMed.

While your AHI is low and while the leaks are a bit ugly...not so bad as to really make us think the machine was missing flagging events but very definitely bad enough to cause wake ups or disturb sleep (you may or may not remember it) which then of course affects how we feel.

Let's try stabilizing the pressure a bit by adding EPR to sort of match what you were using and see if leaks also stabilize.
Without exhale relief you are using more pressure all night long...your body is used to exhale relief and it might just be rebelling and missing it. The brain likes predictability and you changed its routine.

Now I will say this for people about leaks....Respironics large leak territory is a lot more forgiving than ResMed's line in the sand at 24 L/min. I happen to think that 24 L/min is a bit ultra conservative in terms of sensing and recording and not necessarily in the toilet at 30 L/min

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

1999jacko
Posts: 29
Joined: Thu Feb 02, 2017 5:48 am

Re: New machine causing issues

Post by 1999jacko » Mon Jun 28, 2021 10:04 am

Thanks a lot for that. I'll try EPR at 2 tonight which still leaves me somewhere to go if there is still a slight issue. Makes sense to me. I ummed and aaahed about turning EPR on and seems like it might have been the wrong call. Will update tomorrow on how I got on. Many thanks.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: New machine causing issues

Post by chunkyfrog » Mon Jun 28, 2021 10:36 am

Pugsy wrote:
Mon Jun 28, 2021 7:14 am

. . . ResMed's line in the sand at 24 L/min. I happen to think that 24 L/min is a bit ultra conservative in terms of sensing and recording and not necessarily in the toilet at 30 L/min

My sleep doc thinks the same way.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

1999jacko
Posts: 29
Joined: Thu Feb 02, 2017 5:48 am

Re: New machine causing issues

Post by 1999jacko » Tue Jun 29, 2021 12:00 am

So this was the latest night with EPR set to 2.....
screenshot-20210629-055221.png
screenshot-20210629-055221.png (252.5 KiB) Viewed 1099 times