Interpreting husband's results correctly?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Interpreting husband's results correctly?

Post by Sunshine50 » Mon May 31, 2021 6:12 pm

I suspect my husband might have some sleep apnea issues due to the following:
1. Seems overly tired some days
2. A little weight gain
3. Does snore a little bit
4. Has admitted he wakes up every few hours and sometimes isn't able to go back to sleep, but attributed this to age (64). He isn't on any medications (besides eye drops for glaucoma).

I put my machine on him for a night, and he actually took to it quite well. I wanted to post his charts and confirm with the experts here that it does look like there are issues with the increases in pressure. I asked and he wasn't aware of being awake for a prolonged period, so I'm not sure what to make of the CA's appearing.

I know this is just one night though.

He wants to avoid doctors, so I've proposed the option of finding a used machine. Thank you for any advice/help!
Screen Shot 2021-05-31 at 6.55.54 PM.jpg
Screen Shot 2021-05-31 at 6.55.54 PM.jpg (879.17 KiB) Viewed 1140 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Interpreting husband's results correctly?

Post by Julie » Mon May 31, 2021 6:43 pm

Don't worry about a handful of CA's... not important at all so far. I would raise his min. pressure to 6 or so and see how that goes for 1-2 nights, otherwise pretty good.

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Mon May 31, 2021 6:58 pm

Julie wrote:
Mon May 31, 2021 6:43 pm
Don't worry about a handful of CA's... not important at all so far. I would raise his min. pressure to 6 or so and see how that goes for 1-2 nights, otherwise pretty good.
Thanks Julie for the recommendation and the reassurance! I have to confess that since I have to go to work tomorrow, I'm using my machine tonight (although I do wonder if he has more need of it than me at this point.)

User avatar
zonker
Posts: 11436
Joined: Fri Jun 19, 2015 4:36 pm

Re: Interpreting husband's results correctly?

Post by zonker » Mon May 31, 2021 7:02 pm

Sunshine50 wrote:
Mon May 31, 2021 6:12 pm
I suspect my husband might have some sleep apnea issues due to the following:
welcome to the zoo! it's very kind of you to look out for your husband this way.
Sunshine50 wrote:
Mon May 31, 2021 6:12 pm
... confirm with the experts here that it does look like there are issues with the increases in pressure.
i don't know what you mean by this. but looking at the pressure graph, we can see that the minimum may want to be raised as it goes up to 6 or so and doesn't come back down really until the end there.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
SleepGeek
Posts: 1058
Joined: Mon Feb 09, 2015 2:56 pm

Re: Interpreting husband's results correctly?

Post by SleepGeek » Mon May 31, 2021 8:21 pm

Sunshine50 wrote:
Mon May 31, 2021 6:12 pm
He wants to avoid doctors, so I've proposed the option of finding a used machine.
Not a bad idea. You're in a good place to get the help to do that.

He's lucky to have you and even luckier you were willing to share your cpap - ha.
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: What exactly do you think you're adding to this thread?

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Tue Jun 01, 2021 10:33 am

Sunshine50 wrote:
Mon May 31, 2021 6:12 pm
... confirm with the experts here that it does look like there are issues with the increases in pressure.
zonker wrote:
Mon May 31, 2021 7:02 pm
i don't know what you mean by this. but looking at the pressure graph, we can see that the minimum may want to be raised as it goes up to 6 or so and doesn't come back down really until the end there.
Sorry for being unclear! I meant that the increases in pressure mean that he does indeed need cpap/apap therapy. In other words, I'm looking for reassurance that we're pursuing the right path by setting him up with a machine, etc. (and it sounds like the answer is likely yes).

User avatar
Pugsy
Posts: 65298
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Interpreting husband's results correctly?

Post by Pugsy » Tue Jun 01, 2021 10:35 am

Sunshine50 wrote:
Tue Jun 01, 2021 10:33 am
Sorry for being unclear! I meant that the increases in pressure mean that he does indeed need cpap/apap therapy. In other words, I'm looking for reassurance that we're pursuing the right path by setting him up with a machine, etc. (and it sounds like the answer is likely yes).
Yes.....correction assumption.
If that report was from my husband....I would be finding him a machine to use.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Tue Jun 01, 2021 10:55 am

SleepGeek wrote:
Mon May 31, 2021 8:21 pm
Sunshine50 wrote:
Mon May 31, 2021 6:12 pm
He wants to avoid doctors, so I've proposed the option of finding a used machine.
Not a bad idea. You're in a good place to get the help to do that.

He's lucky to have you and even luckier you were willing to share your cpap - ha.
My husband has a saying that he doesn't want to encourage doctors or lawyers.

Also, the symptom that led me to investigate what was wrong was that he was uncharacteristically cranky, so it's in my own best interest to get him back to feeling better! :lol:

Now I just need to find a gently used Resmed, preferably Autoset for Her.

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Tue Jun 01, 2021 10:57 am

Pugsy wrote:
Tue Jun 01, 2021 10:35 am
Sunshine50 wrote:
Tue Jun 01, 2021 10:33 am
Sorry for being unclear! I meant that the increases in pressure mean that he does indeed need cpap/apap therapy. In other words, I'm looking for reassurance that we're pursuing the right path by setting him up with a machine, etc. (and it sounds like the answer is likely yes).
Yes.....correction assumption.
If that report was from my husband....I would be finding him a machine to use.
Thanks Pugsy!

User avatar
zonker
Posts: 11436
Joined: Fri Jun 19, 2015 4:36 pm

Re: Interpreting husband's results correctly?

Post by zonker » Tue Jun 01, 2021 11:20 am

Sunshine50 wrote:
Tue Jun 01, 2021 10:55 am


Now I just need to find a gently used Resmed, preferably Autoset for Her.
leave a pm for forum member lsat. he quite often has gently used machines for sale and i've seen the for her model sold by him many times.

while i haven't dealt with him, other forum members have and were really pleased with the results.

good luck!

eta: HUH. he just now put one up for sale!
http://cpaptalk.com/viewtopic/t182169/A ... -SALE.html
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Interpreting husband's results correctly?

Post by Okie bipap » Tue Jun 01, 2021 6:28 pm


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Wed Jun 02, 2021 2:59 am

zonker wrote:
Tue Jun 01, 2021 11:20 am
Sunshine50 wrote:
Tue Jun 01, 2021 10:55 am


Now I just need to find a gently used Resmed, preferably Autoset for Her.
leave a pm for forum member lsat. he quite often has gently used machines for sale and i've seen the for her model sold by him many times.

while i haven't dealt with him, other forum members have and were really pleased with the results.

good luck!

eta: HUH. he just now put one up for sale!
http://cpaptalk.com/viewtopic/t182169/A ... -SALE.html
Yes, thank you zonker and Okie bipap--I nearly danced after work when I saw that posting from LSAT & promptly PM'd him.

User avatar
zonker
Posts: 11436
Joined: Fri Jun 19, 2015 4:36 pm

Re: Interpreting husband's results correctly?

Post by zonker » Wed Jun 02, 2021 10:41 am

Sunshine50 wrote:
Wed Jun 02, 2021 2:59 am

Yes, thank you zonker and Okie bipap--I nearly danced after work when I saw that posting from LSAT & promptly PM'd him.
yay!


Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Sunshine50
Posts: 10
Joined: Mon May 31, 2021 5:45 pm

Re: Interpreting husband's results correctly?

Post by Sunshine50 » Sat Jun 26, 2021 3:14 pm

It's been a few weeks, enabling me to collect more data. I'm so grateful to LSAT for selling me the machine at just the right time--I can't imagine waiting months to go through the insurance hoops to get a sleep study while worrying about the effects of untreated sleep apnea.

My husband is a pretty agreeable, non-complaining fellow. The main problem is that he often doesn't wear his mask long enough, so I think we probably need to experiment with other masks. I've posted some of his charts below. His current pressure is at 6--should I increase it? Any other recommendations? Thank you!
Screen Shot 2021-06-26 at 3.53.30 PM.jpg
Screen Shot 2021-06-26 at 3.53.30 PM.jpg (660.5 KiB) Viewed 961 times
Screen Shot 2021-06-26 at 3.55.15 PM.jpg
Screen Shot 2021-06-26 at 3.55.15 PM.jpg (703.02 KiB) Viewed 961 times
Screen Shot 2021-06-26 at 4.08.59 PM.jpg
Screen Shot 2021-06-26 at 4.08.59 PM.jpg (681.56 KiB) Viewed 961 times

User avatar
Pugsy
Posts: 65298
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Interpreting husband's results correctly?

Post by Pugsy » Sat Jun 26, 2021 3:22 pm

Have him try 8 cm minimum with same EPR and see how he does.
The machine is going up and down a lot but always pretty much averaging at least around 10 ish pressure.

I wonder what his flow limitation graph looks like? Is he complaining of any nasal congestion to speak of?
Flow limitations can drive the pressure upwards and sometimes nasal congestion can cause flow limitations but for nasal congestion flow limitations more pressure isn't the answer. Instead we have to address nasal congestion by other means.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.