resmed F20 question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Lou Cypher
Posts: 347
Joined: Mon Aug 17, 2009 3:51 pm
Location: Pittsburgh

resmed F20 question

Post by Lou Cypher » Wed Apr 28, 2021 8:36 pm

I have recently lost a lot of weight

The medium is a bit too short and before the night is over it is shoved up on my the bottom of my mouth.

The large is a bit too large and it eventually leaks out the right lower corner , if it tighten that , it leaks out the left lower corner.

I have them so tight I have marks in my face most of the day....

I have tried chin straps underneath and that creates a whole new bunch of problems.... ( not to mention I tried taping the mouth too....)

Any help would be appreciated.

I can't find any other full face that comes this close to working for me.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed.

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: resmed F20 question

Post by Sheriff Buford » Thu Apr 29, 2021 3:47 am

Are you filling the cushion with air when you put the mask on? No mask should be real tight. Snug.... yes. If you fill the cushion with air, the mask will feel somewhat better against your face. I will put the mask on but not connect the lower straps. I then start the machine allowing the cushion to fill with air, then attach the bottom straps. A little air may blow out the sides of the mask while connecting the straps.

See if this helps.

Sheriff

User avatar
Lou Cypher
Posts: 347
Joined: Mon Aug 17, 2009 3:51 pm
Location: Pittsburgh

Re: resmed F20 question

Post by Lou Cypher » Thu Apr 29, 2021 5:30 am

I will try this.....

with the little air blowing out the sides ...does this still happen after connecting the bottom straps.... that is ok to leak ???

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed.

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: resmed F20 question

Post by Sheriff Buford » Fri Apr 30, 2021 3:46 am

No... you shouldn't have any air blowing out the sides. The mask should seal the air in. Let me know if the air blows out when you inflate the cushion.

Sheriff

User avatar
LSAT
Posts: 13408
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: resmed F20 question

Post by LSAT » Fri Apr 30, 2021 8:06 am

After trying several FF masks, I have found that masks with the forehead support are the most stable...especially for people that move around at night. Personally I think the F&P Simplus and Vitera fit that description.

User avatar
Lou Cypher
Posts: 347
Joined: Mon Aug 17, 2009 3:51 pm
Location: Pittsburgh

Re: resmed F20 question

Post by Lou Cypher » Fri Apr 30, 2021 12:55 pm

Sheriff Buford wrote:
Fri Apr 30, 2021 3:46 am
No... you shouldn't have any air blowing out the sides. The mask should seal the air in. Let me know if the air blows out when you inflate the cushion.

Sheriff
It did again last night.... but I will say it takes about an hour and then it leaks again... sometimes I just reach over and turn it off and then back on again...kind of like a snooze button...LOL

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed.

jsmit86
Posts: 267
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: resmed F20 question

Post by jsmit86 » Sun May 02, 2021 6:59 pm

Lou Cypher wrote:
Fri Apr 30, 2021 12:55 pm
Sheriff Buford wrote:
Fri Apr 30, 2021 3:46 am
No... you shouldn't have any air blowing out the sides. The mask should seal the air in. Let me know if the air blows out when you inflate the cushion.

Sheriff
It did again last night.... but I will say it takes about an hour and then it leaks again... sometimes I just reach over and turn it off and then back on again...kind of like a snooze button...LOL
Check to make sure that you don't have any hose leaks or humidifier leaks.
I had a time when I kept waking up with air blowing hard, and out of the sides of my mask. It turned out to be due to a leaking hose. The machine would try to compensate by taking the pressure up to 20. Once I replaced the hose, no more issues.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

KingSnore
Posts: 43
Joined: Sat Apr 17, 2021 2:00 pm

Re: resmed F20 question

Post by KingSnore » Sun May 02, 2021 7:12 pm

The way this topic is going I feel like here would be a good place to ask something.

My pressure is so high. Was set at 15 and then recently I've been doing the home titration and when it ends I expect it'll set me at 17-17.5 or so. I wonder about mask tightness too because particularly during any sort of halfway pleasant exhalation I attempt, there will be leakage. Is that okay, exhalation leakage? I don't see too many large leaks in the OSCAR stuff, and I want to avoid over tightening. But this has been an issue for me no matter what mask I've dealt with.

_________________
Mask
PR System One 60 Series since 7/13/15
AirFit F30i

User avatar
Sheriff Buford
Posts: 4111
Joined: Mon Aug 09, 2010 8:01 am
Location: Kingwood, Texas

Re: resmed F20 question

Post by Sheriff Buford » Mon May 03, 2021 3:43 am

Your machine can compensate for small leaks. A resmed will compensate up to 24L. Also, as long as the leak rate doesn't wake you up... you're good.

Sheriff

Grenin
Posts: 7
Joined: Fri Aug 03, 2012 3:09 pm

Re: resmed F20 question

Post by Grenin » Tue May 04, 2021 5:41 pm

Are you using the memory foam cushion type for this mask? I found that really helped with my leak issues(I tend to move quite a bit during the night).

_________________
MachineMask

jsmit86
Posts: 267
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: resmed F20 question

Post by jsmit86 » Wed May 05, 2021 6:57 pm

Grenin wrote:
Tue May 04, 2021 5:41 pm
Are you using the memory foam cushion type for this mask? I found that really helped with my leak issues(I tend to move quite a bit during the night).
Agree! I originally started with the AirFit F20, which has the silicone cushion. It was ok, but I had some leak and fit issues.
I switched to the AirTouch F20 which is a foam mask. It was much easier to get a good fit and seal. The downside is that it will need to be replaced more often.
While they recommend 30 days, I can sometimes get 60-90 days on these. For me, it's worth the extra expense for the comfort and better seal.
They both use exactly the same frame and headgear, so if you want to try foam, you can order just the cushion.
They can be found online for about $25 most of the time.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
Lou Cypher
Posts: 347
Joined: Mon Aug 17, 2009 3:51 pm
Location: Pittsburgh

Re: resmed F20 question

Post by Lou Cypher » Thu May 06, 2021 6:28 am

Grenin wrote:
Tue May 04, 2021 5:41 pm
Are you using the memory foam cushion type for this mask? I found that really helped with my leak issues(I tend to move quite a bit during the night).
yes i am

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed.

User avatar
Lou Cypher
Posts: 347
Joined: Mon Aug 17, 2009 3:51 pm
Location: Pittsburgh

Re: resmed F20 question

Post by Lou Cypher » Thu May 06, 2021 6:30 am

jsmit86 wrote:
Sun May 02, 2021 6:59 pm
Lou Cypher wrote:
Fri Apr 30, 2021 12:55 pm
Sheriff Buford wrote:
Fri Apr 30, 2021 3:46 am
No... you shouldn't have any air blowing out the sides. The mask should seal the air in. Let me know if the air blows out when you inflate the cushion.

Sheriff
It did again last night.... but I will say it takes about an hour and then it leaks again... sometimes I just reach over and turn it off and then back on again...kind of like a snooze button...LOL
Check to make sure that you don't have any hose leaks or humidifier leaks.
I had a time when I kept waking up with air blowing hard, and out of the sides of my mask. It turned out to be due to a leaking hose. The machine would try to compensate by taking the pressure up to 20. Once I replaced the hose, no more issues.
Thanks for this but I changed those and still have the issue.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed.

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: resmed F20 question

Post by Pugsy » Thu May 06, 2021 7:56 am

Lou Cypher wrote:
Thu May 06, 2021 6:28 am
Grenin wrote:
Tue May 04, 2021 5:41 pm
Are you using the memory foam cushion type for this mask? I found that really helped with my leak issues(I tend to move quite a bit during the night).
yes i am
How old is the memory foam cushion? When did you change it for a new cushion last?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.