New APAP user without sleep study looking for advice

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ChicagoGranny
Posts: 15261
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: New APAP user without sleep study looking for advice

Post by ChicagoGranny » Tue May 25, 2021 10:02 am

MaskMan8888 wrote:
Tue May 25, 2021 6:50 am
If you are looking to find a good mask to use either now or in the future, try downloading MaskFit AR on the App Store. IT's a free app that helps suggest the best mask for you from all makes and models. Truly an objective and unbiased system.

https://www.youtube.com/watch?v=t8c-BIT4TjA&t=1s
"... suggests hundreds of masks (that fit the scan) ..."

Wow! That really narrows it down. :lol: :lol:
Last edited by ChicagoGranny on Tue May 25, 2021 10:10 am, edited 1 time in total.

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Tue May 25, 2021 10:05 am

ChicagoGranny wrote:
Tue May 25, 2021 10:02 am
MaskMan8888 wrote:
Tue May 25, 2021 6:50 am
If you are looking to find a good mask to use either now or in the future, try downloading MaskFit AR on the App Store. IT's a free app that helps suggest the best mask for you from all makes and models. Truly an objective and unbiased system.

https://www.youtube.com/watch?v=t8c-BIT4TjA&t=1s
"... suggests hundreds of mask (that fit the scan) ..."

Wow! That really narrows it down. :lol: :lol:
Haha. I actually tried the app right now; doesn't seem too bad. Now I know that I need a large size mask.
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
ChicagoGranny
Posts: 15261
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: New APAP user without sleep study looking for advice

Post by ChicagoGranny » Tue May 25, 2021 10:16 am

gadgetmaniah wrote:
Tue May 25, 2021 10:05 am
Haha. I actually tried the app right now; doesn't seem too bad. Now I know that I need a large size mask.
Most of the masks on cpap.com show a fitting guide provided by the mask manufacturer. I don't know whether the fitting guide or the app is the better tool.

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Tue May 25, 2021 10:23 am

ChicagoGranny wrote:
Tue May 25, 2021 10:16 am
gadgetmaniah wrote:
Tue May 25, 2021 10:05 am
Haha. I actually tried the app right now; doesn't seem too bad. Now I know that I need a large size mask.
Most of the masks on cpap.com show a fitting guide provided by the mask manufacturer. I don't know whether the fitting guide or the app is the better tool.
Oh right, I was just taking a look at Resmed's fitting guide on their website. Seems simpler.
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
ChicagoGranny
Posts: 15261
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: New APAP user without sleep study looking for advice

Post by ChicagoGranny » Tue May 25, 2021 10:25 am

gadgetmaniah wrote:
Tue May 25, 2021 10:23 am
Seems simpler.
Narrows it down to dozens? :lol: :lol:

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Tue May 25, 2021 10:41 am

ChicagoGranny wrote:
Tue May 25, 2021 10:25 am
gadgetmaniah wrote:
Tue May 25, 2021 10:23 am
Seems simpler.
Narrows it down to dozens? :lol: :lol:
Lol!
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
Pugsy
Posts: 65114
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New APAP user without sleep study looking for advice

Post by Pugsy » Tue May 25, 2021 11:27 am

ChicagoGranny wrote:
Tue May 25, 2021 10:02 am
MaskMan8888 wrote:
Tue May 25, 2021 6:50 am
If you are looking to find a good mask to use either now or in the future, try downloading MaskFit AR on the App Store. IT's a free app that helps suggest the best mask for you from all makes and models. Truly an objective and unbiased system.

https://www.youtube.com/watch?v=t8c-BIT4TjA&t=1s
"... suggests hundreds of masks (that fit the scan) ..."

Wow! That really narrows it down. :lol: :lol:

I bet somewhere out there there is an app for how to effectively pick your nose. :lol: :lol: or wipe your butt....and people will think it is the greatest thing since sliced bread.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11331
Joined: Fri Jun 19, 2015 4:36 pm

Re: New APAP user without sleep study looking for advice

Post by zonker » Tue May 25, 2021 11:31 am

Pugsy wrote:
Tue May 25, 2021 11:27 am


I bet somewhere out there there is an app for how to effectively pick your nose. :lol: :lol: or wipe your butt....and people will think it is the greatest thing since sliced bread.
where can i get me that sliced bread app?
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Tue May 25, 2021 4:46 pm

I have a question. Even with my replacement machine (Resmed Auto), I'm experiencing large leaks throughout my sleep. Could this have something to do with having washed my mask? I washed it for the first time before using it with the Resmed machine that I ordered. In addition, I'm experiencing some rough/fluttering type of noise when inhaling.

Does this sound like an issue with the mask or the machine? I doubt that it's with the machine as I'm experiencing the issue on two different machines that I've tried that I received from the supplier.

Also, my mask is a cheap mask by an obscure company called Galemed, so that's why I haven't mentioned its name.
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
Pugsy
Posts: 65114
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New APAP user without sleep study looking for advice

Post by Pugsy » Tue May 25, 2021 5:07 pm

Mask movement during the night as we change position or facial oils/sweat accumulate and mess with the seal is very common.
When the mask moves from whatever reason....leaks happen.

I once woke up with one of my nasal pillows blowing air into my ear and the other one was blowing air into my eyeball. :lol: :lol:
Had a big leak show up on the reports when that happened.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Wed May 26, 2021 9:20 am

Pugsy wrote:
Tue May 25, 2021 5:07 pm
Mask movement during the night as we change position or facial oils/sweat accumulate and mess with the seal is very common.
When the mask moves from whatever reason....leaks happen.

I once woke up with one of my nasal pillows blowing air into my ear and the other one was blowing air into my eyeball. :lol: :lol:
Had a big leak show up on the reports when that happened.
Right, thank you!

Also, I wanted to ask, why is my humidifer status showing as 'On' on OSCAR even though I'm not using a humidifier? Is this normal?
Attachments
Humidifier status.png
Humidifier status.png (12.5 KiB) Viewed 1057 times
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
Pugsy
Posts: 65114
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New APAP user without sleep study looking for advice

Post by Pugsy » Wed May 26, 2021 9:44 am

gadgetmaniah wrote:
Wed May 26, 2021 9:20 am
why is my humidifer status showing as 'On' on OSCAR even though I'm not using a humidifier? Is this normal?
It's a bug in the OSCAR software. Thought they had fixed that little problem.
Ignore it. Nothing you can do about it. Sometimes the actual settings OSCAR doesn't get right due to the little bug in the software.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Wed May 26, 2021 9:49 am

Pugsy wrote:
Wed May 26, 2021 9:44 am
gadgetmaniah wrote:
Wed May 26, 2021 9:20 am
why is my humidifer status showing as 'On' on OSCAR even though I'm not using a humidifier? Is this normal?
It's a bug in the OSCAR software. Thought they had fixed that little problem.
Ignore it. Nothing you can do about it. Sometimes the actual settings OSCAR doesn't get right due to the little bug in the software.
Oh, I see. Thank you.
Resmed Airsense 10 Autoset and BMC F5 Mask.

gadgetmaniah
Posts: 129
Joined: Mon Apr 26, 2021 6:12 am
Location: Pakistan

Re: New APAP user without sleep study looking for advice

Post by gadgetmaniah » Fri May 28, 2021 1:20 pm

I have a question about my machine settings. I've heard that an EPR of 3 is optimal. So, if I normally have my minimum at 6 cm, would I need to increase the minimum pressure if I set EPR to 3? If so, what should I increase pressure to? As a guess I've set it at 7 cm right now.
Resmed Airsense 10 Autoset and BMC F5 Mask.

User avatar
ChicagoGranny
Posts: 15261
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: New APAP user without sleep study looking for advice

Post by ChicagoGranny » Fri May 28, 2021 3:35 pm

gadgetmaniah wrote:
Fri May 28, 2021 1:20 pm
I've heard that an EPR of 3 is optimal. So, if I normally have my minimum at 6 cm, would I need to increase the minimum pressure if I set EPR to 3?
Each patient is an individual case. There is no optimal EPR for the patient population, and there is no formula for assigning an EPR. The optimal EPR/Min/Max setting is the one that is effective for the user and doesn't interrupt the user's sleep. For example, some like very much an EPR of 3, and others, like me, don't like the constant up and down cycling of the pressure. I don't use EPR. Others won't do without it.

To find optimal settings, the user can experiment with different settings, make subjective judgments about his sleep quality and energy level the next day, and check the daily data in OSCAR. Trial and error.

As far as turning EPR on and increasing the minimum pressure, some have found they have to do this.

Here is a quote from ResMed.
The ResMed AirSense 10 has built-in sensors that detect a patient’s breathing. As an individual exhales, the CPAP machine responds by reducing the air pressure delivered by a specific amount. When an inhale is detected again, the machine raises the pressure back to the prescribed setting. This is known as EPR, or expiratory pressure relief.

The ResMed AirSense 10 has four EPR settings: 0-3. At 0, pressure relief is turned off. At 3, air pressure is reduced by 3 cm H2O. For example, if a patient’s CPAP is set to 10 cm H2O with an EPR of 3, the air pressure during exhale is lowered to 7 cm H2O. Once an inhale is detected, air pressure increases back to 10.
Note if EPR is set at 3 and minimum pressure is set at 6.0, the machine will not go below 4.0. So, the EPR setting is effectively only 2.