I've been doing well since starting CPAP almost one year ago but I still get some fluctuation generally between 2.0 and 6.0 AHI from day to day and I can't figure out why. I've included two OSCAR reports as examples.

Also a question about Leak Rate. I see two graph lines that pretty much follow each other, what do they represent?
screenshot-20210215-092132.png
screenshot-20210215-092132.png (169.86 KiB) Viewed 1103 times
screenshot-20210215-091251.png
screenshot-20210215-091251.png (167.67 KiB) Viewed 1103 times

_________________
MachineMask

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Fluctuation

Post by ChicagoGranny » Mon Feb 15, 2021 1:20 pm

Your pressure minimum is set to 9.0. You are spending most of the night from 12.0 to 20.0.

I would set the minimum at 12.0 and check each day how things go.

User avatar
LSAT
Posts: 13374
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI Fluctuation

Post by LSAT » Mon Feb 15, 2021 2:46 pm

I agree with Granny

User avatar
gwfl
Posts: 42
Joined: Wed Mar 04, 2020 4:04 pm

Re: AHI Fluctuation

Post by gwfl » Tue Feb 16, 2021 2:22 pm

What effect would that have (raising it to say 12.0) if my need doesn't drop to 9.0 anyway?

_________________
MachineMask

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AHI Fluctuation

Post by jimbud » Tue Feb 16, 2021 2:30 pm

gwfl wrote:
Tue Feb 16, 2021 2:22 pm
What effect would that have (raising it to say 12.0) if my need doesn't drop to 9.0 anyway?
9.0 is not your need. It is your machine set minimum.

Your need is higher. (To prevent/knock down obstructive events.)

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
zonker
Posts: 11398
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI Fluctuation

Post by zonker » Tue Feb 16, 2021 2:48 pm

gwfl wrote:
Tue Feb 16, 2021 2:22 pm
What effect would that have (raising it to say 12.0) if my need doesn't drop to 9.0 anyway?
it will be able to react more quickly to your events and kill them off.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
gwfl
Posts: 42
Joined: Wed Mar 04, 2020 4:04 pm

Re: AHI Fluctuation

Post by gwfl » Tue Feb 16, 2021 5:05 pm

Thank you. I'll give it a try.

_________________
MachineMask

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AHI Fluctuation

Post by jimbud » Tue Feb 16, 2021 5:15 pm

gwfl wrote:
Tue Feb 16, 2021 5:05 pm
Thank you. I'll give it a try.
A jump in pressure from 9 to 12 is difficult for most. You may have to raise pressure in smaller increments.
How much one can tolerate may take some experimentation.

Good luck.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Fluctuation

Post by ChicagoGranny » Wed Feb 17, 2021 8:54 am

jimbud wrote:
Tue Feb 16, 2021 5:15 pm
gwfl wrote:
Tue Feb 16, 2021 5:05 pm
Thank you. I'll give it a try.
A jump in pressure from 9 to 12 is difficult for most. You may have to raise pressure in smaller increments.
How much one can tolerate may take some experimentation.

Good luck.

JPB
He's already running near 20.0 at times.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AHI Fluctuation

Post by jimbud » Wed Feb 17, 2021 9:43 am

ChicagoGranny wrote:
Wed Feb 17, 2021 8:54 am
He's already running near 20.0 at times.
Yes I know.

His starting pressure is at 9. A jump from 9 to 12 starting pressure can be difficult for some. At least it was for me. Maybe I am the only one?
I think some of you old hands forget how it was when you started this trip. Or maybe it was just easy for you. Lucky draw. :D

Anyway, I do not think it was Bad advice. :wink:

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Fluctuation

Post by Pugsy » Wed Feb 17, 2021 9:53 am

jimbud wrote:
Wed Feb 17, 2021 9:43 am
ChicagoGranny wrote:
Wed Feb 17, 2021 8:54 am
He's already running near 20.0 at times.
Yes I know.

His starting pressure is at 9. A jump from 9 to 12 starting pressure can be difficult for some. At least it was for me. Maybe I am the only one?
I think some of you old hands forget how it was when you started this trip. Or maybe it was just easy for you. Lucky draw. :D

Anyway, I do not think it was Bad advice. :wink:

JPB

It wasn't bad advice. That big of a jump would indeed very likely to be difficult for someone to make in a big jump.

Yes, he is using near 20 most of the night but we assume that is because he is asleep when the machine is going there.
The machine is essentially setting a new minimum pressure though...so I am not thinking that it is so critical that the minimum of even 12 is going to make that big of a difference in the results. There's just not much evidence of the machine doing a lot of pressure changes once it goes up.

It did go back down at around 4:30 and stayed down...makes me wonder why? Maybe a wake up and he didn't go back to sleep????

The report shown screams poor sleep quality to me....now would a higher minimum help??? Dunno but it is worth trying but I agree that small steps might be the best way to go. It is also possible that the poor sleep quality isn't from the pressures not being optimal but instead being from something else so I would also investigate the other usual causes of crappy sleep.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Fluctuation

Post by ChicagoGranny » Wed Feb 17, 2021 11:25 am

jimbud wrote:
Wed Feb 17, 2021 9:43 am
Anyway, I do not think it was Bad advice.
Pugsy wrote:
Wed Feb 17, 2021 9:53 am
It wasn't bad advice.
Who said it was bad advice?

I would change the minimum to 12 if those were my charts. Or, better yet, wear a foam cervical collar and see if a lower pressure is effective.

He could start with a one-inch collar for a week, then a two-inch collar for a week, ...

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: AHI Fluctuation

Post by jimbud » Wed Feb 17, 2021 11:43 am

ChicagoGranny wrote:
Wed Feb 17, 2021 11:25 am
He could start with a one-inch collar for a week, then a two-inch collar for a week, ...
Or he could just jump right in and start with a 12" collar.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
zonker
Posts: 11398
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI Fluctuation

Post by zonker » Wed Feb 17, 2021 11:51 am

ChicagoGranny wrote:
Wed Feb 17, 2021 11:25 am

He could start with a one-inch collar for a week, then a two-inch collar for a week, ...
then wear the 1" and the 2" at the same time, then wear two 2" collars at the same time....
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65196
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Fluctuation

Post by Pugsy » Wed Feb 17, 2021 11:51 am

ChicagoGranny wrote:
Wed Feb 17, 2021 11:25 am
Who said it was bad advice?
No one directly.
I was just agreeing with jimbud.
His advice wasn't bad but I guess someone made him think that someone thought it was bad advice.

If that were my report...I would gradually increase the minimum 1 cm per week and see what happens.
I personally don't want any more crap on my face, head, neck or whatever so adding anything is not where I would want to go but I understand some people are okay with adding more stuff to wear and if they are good with it...more power to them.

Nice thing though....there are multiple options available for the OP to consider and try and decide for himself how he wants to proceed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.