AirSense 10, Dry and Running out of Water
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
AirSense 10, Dry and Running out of Water
Hey friends,
Things feel really dry in my mask throughout the night, so I set the humidity at the highest level for my machine (8) and I have the heated tube at 81 percent. The problem is that it still feels dry and the reservoir runs out of water after about five hours. I wake up and have to turn the machine because it's just so dry. Does anyone have any suggestions? I tried to search cpaptalk to see if others had this problem and I don't see a similar thread.
Thanks,
Mike
Things feel really dry in my mask throughout the night, so I set the humidity at the highest level for my machine (8) and I have the heated tube at 81 percent. The problem is that it still feels dry and the reservoir runs out of water after about five hours. I wake up and have to turn the machine because it's just so dry. Does anyone have any suggestions? I tried to search cpaptalk to see if others had this problem and I don't see a similar thread.
Thanks,
Mike
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16
Re: AirSense 10, Dry and Running out of Water
What is the ambient humidity in your bedroom?
What is your pressure setting?
What are your leak numbers?
What is your pressure setting?
What are your leak numbers?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4258
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AirSense 10, Dry and Running out of Water
You moved from whatever the setting was, to the highest setting available.
Instead of doing that you should have just increased the humidity by 1. That may have been enough. And then see how that works for you, then readjust again if neccesary.
When you run out of water, of course it will be dry.
Instead of doing that you should have just increased the humidity by 1. That may have been enough. And then see how that works for you, then readjust again if neccesary.
When you run out of water, of course it will be dry.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
Re: AirSense 10, Dry and Running out of Water
Thanks.
It's dry in this house. I do run a humidifier during the night. My pressure setting is 16. I haven't checked the specific number, but my Cpap machine does not show a leak warning.
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16
Re: AirSense 10, Dry and Running out of Water
Are you a mouth breather?michaeldwilson wrote: ↑Fri Jan 29, 2021 9:34 amThanks.
It's dry in this house. I do run a humidifier during the night. My pressure setting is 16. I haven't checked the specific number, but my Cpap machine does not show a leak warning.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
Re: AirSense 10, Dry and Running out of Water
LSAT wrote: ↑Fri Jan 29, 2021 9:37 amAre you a mouth breather?michaeldwilson wrote: ↑Fri Jan 29, 2021 9:34 amThanks.
It's dry in this house. I do run a humidifier during the night. My pressure setting is 16. I haven't checked the specific number, but my Cpap machine does not show a leak warning.
Thanks, yes.
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16
Re: AirSense 10, Dry and Running out of Water
So test the ambient humidity and see what it is with the humidifier added.
3 things affect water consumption
Ambient humidity
Pressure used
Leaks...meaning excessive.
Mouth feeling dry isn't always fixable with maximum humidity which you have found out and also found out that you run out of water during the night.
As to why the mouth feels dry...now and I assume it didn't in the past....any new meds or change in meds? Any new nasal congestion that might cause more mouth breathing?
You may have 2 problems here...dry mouth that even a full face mask at maximum humidity doesn't fix and the fact that the machine is running out of water trying to deliver that maximum humidity.
3 things affect water consumption
Ambient humidity
Pressure used
Leaks...meaning excessive.
Mouth feeling dry isn't always fixable with maximum humidity which you have found out and also found out that you run out of water during the night.
As to why the mouth feels dry...now and I assume it didn't in the past....any new meds or change in meds? Any new nasal congestion that might cause more mouth breathing?
You may have 2 problems here...dry mouth that even a full face mask at maximum humidity doesn't fix and the fact that the machine is running out of water trying to deliver that maximum humidity.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
Re: AirSense 10, Dry and Running out of Water
Thanks, Pugsy! I checked SleepyHead just now and it looks like I had a leak rate of 10 for an hour and then 6 for an hour. But hours before that it was 0 (zero). I'll work on making sure it leaks less during the night and turn up the humidifier in the room.Pugsy wrote: ↑Fri Jan 29, 2021 9:47 amSo test the ambient humidity and see what it is with the humidifier added.
3 things affect water consumption
Ambient humidity
Pressure used
Leaks...meaning excessive.
Mouth feeling dry isn't always fixable with maximum humidity which you have found out and also found out that you run out of water during the night.
As to why the mouth feels dry...now and I assume it didn't in the past....any new meds or change in meds? Any new nasal congestion that might cause more mouth breathing?
You may have 2 problems here...dry mouth that even a full face mask at maximum humidity doesn't fix and the fact that the machine is running out of water trying to deliver that maximum humidity.
I think you are right that maximum humidity is not fixing the dry feeling. I'll work with it to see what's tolerable.
Do you know if raising or lowering the temperature of the heated tube will affect water usage?
Again, much appreciated!
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16
Re: AirSense 10, Dry and Running out of Water
Those leak numbers are small and unlikely to have caused so much water consumption.
Leaks need to be really pretty large (like over 30 L/min and prolonged) to make much of an impact on the amount of water used.
Most likely your culprit is going to be the ambient humidity whatever it is.
I know that I use a setting of 6 and my ambient humidity runs usually 45 to 60% and with 8 hours of use I will use up almost all my water.
But it might cause you to have water condense in your mask or hose which is highly annoying.
The main job of the heated air....prevent the air from cooling and releasing the moisture in the air coming from your machine.
The second job of the heated air...simply comfort...whatever you personally like better. I am not real fond of really warm air coming from my machine but with my preferred humidity setting and my preferred rather cold bedroom temps I find that condensation in my mask or hose is pretty much guaranteed if I don't use almost maximum hose air temp.
So while I am not fond of the warmer air...I am even less fond of the condensation in the mask or hose so I use the warmer air.
The dry mouth (even with maximum humidity being delivered and not running out of water) is a separate issue you will need to investigate. It's a common complaint. We first look for medication side effects being a potential culprit and then we look for changes in other health issues....like something causing a person to breathe more through their mouth than they used to be doing.
You may need to add some sort of oral hydrating agent to your bedtime routine at least short term until you can maybe figure out if there is something else going on.
And of course...extra hydration during the day. I am not so much for loading up on water at bedtime because that may or may not help the mouth dryness but it sure will likely guarantee a wake up to go pee during the night.
It's better I think to hydrate optimally throughout the day.
Leaks need to be really pretty large (like over 30 L/min and prolonged) to make much of an impact on the amount of water used.
Most likely your culprit is going to be the ambient humidity whatever it is.
I know that I use a setting of 6 and my ambient humidity runs usually 45 to 60% and with 8 hours of use I will use up almost all my water.
It will NOT affect the water consumption.michaeldwilson wrote: ↑Fri Jan 29, 2021 10:07 amDo you know if raising or lowering the temperature of the heated tube will affect water usage?
But it might cause you to have water condense in your mask or hose which is highly annoying.
The main job of the heated air....prevent the air from cooling and releasing the moisture in the air coming from your machine.
The second job of the heated air...simply comfort...whatever you personally like better. I am not real fond of really warm air coming from my machine but with my preferred humidity setting and my preferred rather cold bedroom temps I find that condensation in my mask or hose is pretty much guaranteed if I don't use almost maximum hose air temp.
So while I am not fond of the warmer air...I am even less fond of the condensation in the mask or hose so I use the warmer air.
The dry mouth (even with maximum humidity being delivered and not running out of water) is a separate issue you will need to investigate. It's a common complaint. We first look for medication side effects being a potential culprit and then we look for changes in other health issues....like something causing a person to breathe more through their mouth than they used to be doing.
You may need to add some sort of oral hydrating agent to your bedtime routine at least short term until you can maybe figure out if there is something else going on.
And of course...extra hydration during the day. I am not so much for loading up on water at bedtime because that may or may not help the mouth dryness but it sure will likely guarantee a wake up to go pee during the night.
It's better I think to hydrate optimally throughout the day.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
Re: AirSense 10, Dry and Running out of Water
I appreciate all the good information and advice. I can't thank you enough.Pugsy wrote: ↑Fri Jan 29, 2021 10:28 amThose leak numbers are small and unlikely to have caused so much water consumption.
Leaks need to be really pretty large (like over 30 L/min and prolonged) to make much of an impact on the amount of water used.
Most likely your culprit is going to be the ambient humidity whatever it is.
I know that I use a setting of 6 and my ambient humidity runs usually 45 to 60% and with 8 hours of use I will use up almost all my water.
It will NOT affect the water consumption.michaeldwilson wrote: ↑Fri Jan 29, 2021 10:07 amDo you know if raising or lowering the temperature of the heated tube will affect water usage?
But it might cause you to have water condense in your mask or hose which is highly annoying.
The main job of the heated air....prevent the air from cooling and releasing the moisture in the air coming from your machine.
The second job of the heated air...simply comfort...whatever you personally like better. I am not real fond of really warm air coming from my machine but with my preferred humidity setting and my preferred rather cold bedroom temps I find that condensation in my mask or hose is pretty much guaranteed if I don't use almost maximum hose air temp.
So while I am not fond of the warmer air...I am even less fond of the condensation in the mask or hose so I use the warmer air.
The dry mouth (even with maximum humidity being delivered and not running out of water) is a separate issue you will need to investigate. It's a common complaint. We first look for medication side effects being a potential culprit and then we look for changes in other health issues....like something causing a person to breathe more through their mouth than they used to be doing.
You may need to add some sort of oral hydrating agent to your bedtime routine at least short term until you can maybe figure out if there is something else going on.
And of course...extra hydration during the day. I am not so much for loading up on water at bedtime because that may or may not help the mouth dryness but it sure will likely guarantee a wake up to go pee during the night.
It's better I think to hydrate optimally throughout the day.
Mike
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16
- Miss Emerita
- Posts: 3758
- Joined: Sun Nov 04, 2018 8:07 pm
Re: AirSense 10, Dry and Running out of Water
Hi Mike, for dryness in your mouth, some people get help from Xylimelts, which are small discs that attach to your teeth or gums and stimulate the production of saliva throughout the night.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
michaeldwilson
- Posts: 43
- Joined: Wed Nov 05, 2014 8:31 pm
Re: AirSense 10, Dry and Running out of Water
Thanks for the suggestion. If this water drinking I'm doing today doesn't help, I will give it a try!Miss Emerita wrote: ↑Fri Jan 29, 2021 1:43 pmHi Mike, for dryness in your mouth, some people get help from Xylimelts, which are small discs that attach to your teeth or gums and stimulate the production of saliva throughout the night.
Machine: AirSense™ 10 AutoSet For Her CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit F30i
Additional Comments: Pressure 16
Mask: AirFit F30i
Additional Comments: Pressure 16