New Machine Settings Help Required

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
vbrik
Posts: 2
Joined: Fri Jan 22, 2021 6:28 am
Location: United Kingdom

New Machine Settings Help Required

Post by vbrik » Fri Jan 22, 2021 6:58 am

I was diagnosed with severe apnea end of December and provided with a Dreamstation Auto CPAP plus Resmed F30 full face mask.

Tend to find as a side sleeper that the mask often leaks around the top of the cushion around the nose area when pressure increases, no matter what I do with the fit. Have asked if I can switch to an F&P Simplus

My AHI was 35 prior to commencing with the CPAP treatment and now down to 5.5 but i'm looking to improve on this as get it down to 1 if possible and not sure how to fully interpret the data in Oscar yet, still learning. Do feel the min pressure is pretty low and wonder if still getting higher AHI due to this or not.

Could i have some advice or feedback please on the OSCAR output below and what I could potentially do to improve the AHI, plus any general observations on the stats/figures provided?

Thanks
Mike
screenshot-20210122-123705.png
screenshot-20210122-123705.png (75.17 KiB) Viewed 755 times

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Machine Settings Help Required

Post by Pugsy » Fri Jan 22, 2021 12:47 pm

Increase your minimum pressure from 5 to 6.5 cm and see how that goes for a few nights...then if AHI is not reduced to your satisfaction...try 7 cm.

Please note you can only expect a reduction in the AHI in OAs and hyponeas. Increasing the pressure won't help with CAs/centrals.

Also please note I have been having some major problems accessing the forum...long story...and I don't know how long this window will stay open for me so if you respond and I don't respond back it is because I can't get back onto the forum.

Known problem affecting not just me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: New Machine Settings Help Required

Post by Okie bipap » Fri Jan 22, 2021 12:50 pm

It appears you an use a little more pressure on the low end. I recommend raising your minimum pressure to 6. We recommend you only make one change at a time, then give it a few nights to see what effects it has. As with a lot of things in llife, we need time to adjust to them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

vbrik
Posts: 2
Joined: Fri Jan 22, 2021 6:28 am
Location: United Kingdom

Re: New Machine Settings Help Required

Post by vbrik » Mon Feb 08, 2021 4:29 pm

Sorry for late reply didn’t realise the site redirect errors would last so long.

Ok thanks will increase minimum pressure to 6 and see how that goes for a few nights.

Many thanks
Mike

User avatar
upforsleep
Posts: 51
Joined: Mon Sep 28, 2009 6:13 am
Location: Texas-South East

Re: New Machine Settings Help Required

Post by upforsleep » Tue Feb 09, 2021 7:43 am

If you are looking at the Simplus the F&P viterus is a newer upgrade to condsider.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Cpap Pillow, HoseBuddy, soft cervical collar, mask liner