How do my machine readings look?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

How do my machine readings look?

Post by tbeb » Fri Jan 01, 2021 10:07 am

My sleep test AHI was 28.5. My machine is a Dreamstation Auto with humidifier. Hose is heated. Last night was my 51st night using cpap. Here are my machine readings, shown as
AHI/90% pressure, for the last 14 nights: 4.5/11.0, 2.4/11.0, 3.8/10.0, 3.0/9.0, 4.0/10.5, 3.0/10.5, 4.5/9.0, 2.4/8.0, 4.7/9.0, 3.4/10.5, 2.2/10.0, 2.7/11.5, 3.6/11.5, and 4.8/11.0.
My sleep coach is very polite but honestly not very helpful so that's why I'm asking how my readings look. Thanks.
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: How do my machine readings look?

Post by Miss Emerita » Fri Jan 01, 2021 11:56 am

Hello, tbeb. Your AHI is certainly better than it was, but I wonder whether it could be better still. As a very general rule of thumb, an AHI between 1 and 2 is good and under 1 is excellent. Sleep technicians tend to think anything under 5 is fine.

To recommend possible tweaks to your settings, people would need to see the data displayed via the Oscar software (see below). If you have a laptop or desktop, plus an SD card in your machine and an SD card reader (in your computer or as a plug-in), you are ready to go. We'd be happy to help if you have any problems getting Oscar to work.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: How do my machine readings look?

Post by zonker » Fri Jan 01, 2021 12:05 pm

tbeb, to follow up on what miss e is saying, the ahi is a combination of individual "events". these events can be different for each of us. and their combination can be different. there are "obstructive" events. there are "hypopnea" events. there are "central" events. and so and and so forth.

what the oscar software can tell us is which events make up your particular ahi. once that is seen, the experts can give you tips on how to reduce the events and therefor improve your total ahi.

if you can, please post a chart here, in this thread, so folk can see what is going on.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65250
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How do my machine readings look?

Post by Pugsy » Fri Jan 01, 2021 12:12 pm

It all depends on what that AHI is composed of (the 3 event categories that make up the AHI) and how you feel during the day and how you are sleeping.

Also depends on when those events got flagged...like maybe at the beginning and ending of the sleep session and maybe you were awake and if that's the case they simply don't count.

So can't really tell much about numbers only.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Fri Jan 01, 2021 2:17 pm

zonker wrote:
Fri Jan 01, 2021 12:05 pm
tbeb, to follow up on what miss e is saying, the ahi is a combination of individual "events". these events can be different for each of us. and their combination can be different. there are "obstructive" events. there are "hypopnea" events. there are "central" events. and so and and so forth.

what the oscar software can tell us is which events make up your particular ahi. once that is seen, the experts can give you tips on how to reduce the events and therefor improve your total ahi.

if you can, please post a chart here, in this thread, so folk can see what is going on.

good luck!
Should I post a chart of Usage Statistics or Overview or Daily?
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: How do my machine readings look?

Post by zonker » Fri Jan 01, 2021 6:24 pm

tbeb wrote:
Fri Jan 01, 2021 2:17 pm

Should I post a chart of Usage Statistics or Overview or Daily?
daily please. pugsy tells us this-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

click on the above link, then scroll down to #4, which gets to the heart of the matter.

basically, you want your charts to be organized like this-

"Suggested Support Order for your charts
Event Flags
Flow Rate
Pressure
Leak Rate
Flow Limit (Only for Resmed machines, not on Respironics)"

as miss e says, any problems with this, please don't hesitate to ask.

OH! and you post 3 or more charts(that is, a chart for each of three nights) or more. the experts are looking for trends, not just one crappy night's sleep.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Fri Jan 01, 2021 9:15 pm

screenshot-20210101-210234.png
screenshot-20210101-210234.png (62.54 KiB) Viewed 1208 times
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Fri Jan 01, 2021 10:00 pm

Does my 1-day chart look right? The leak rate chart isn't visible and I don't know why. If it looks alright then I'll post a chart for a few more days.

I fall asleep in 5 minutes or less. Wife wakes me once pretty much every night when she hears the mask leaking (and I don't). I've experienced vertigo every night when I move around in bed. This definitely began with cpap! Long story short, doctor hastily put me on bp medicine. After 2 weeks I got off that and I'm on 2nd day of using a prescription motion sickness type pill. I sleep long enough but don't feel rested when I wake up. Adjusting to cpap and dealing with the vertigo has been exhausting. If motion sickness pill stops the room from spinning then I think I'll be okay.
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: How do my machine readings look?

Post by Julie » Fri Jan 01, 2021 10:12 pm

Look up side effects of that pill - see if anything might interfere with sleep (and don't count on your MD to have done that).

User avatar
Pugsy
Posts: 65250
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How do my machine readings look?

Post by Pugsy » Fri Jan 01, 2021 11:09 pm

Don't really need to see the leak graph at this time. You aren't having any large leaks and we know that because there aren't any large leaks flagged up on the Events graph. The LL line...that's large leak.

Are you waking often during the night?

Is the vertigo you are experiencing during the night lasting throughout the day as well?

The bulk of what you see flagged on the Events graph (except the CA/central stuff) points to the airway trying to collapse a bit still yet but I hesitate to offer the usual fix of "more minimum pressure" because of the vertigo. If the vertigo is related to pressure then more pressure might make the vertigo worse. If the vertigo is indeed for sure related to cpap...you probably need to see an ENT doctor and have the eustachian tubes evaluated. The tubes shouldn't be allowing the pressure into the inner ear but I have heard that sometimes it happens.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Fri Jan 01, 2021 11:59 pm

Pugsy wrote:
Fri Jan 01, 2021 11:09 pm
Don't really need to see the leak graph at this time. You aren't having any large leaks and we know that because there aren't any large leaks flagged up on the Events graph. The LL line...that's large leak.

Are you waking often during the night?

Is the vertigo you are experiencing during the night lasting throughout the day as well?

The bulk of what you see flagged on the Events graph (except the CA/central stuff) points to the airway trying to collapse a bit still yet but I hesitate to offer the usual fix of "more minimum pressure" because of the vertigo. If the vertigo is related to pressure then more pressure might make the vertigo worse. If the vertigo is indeed for sure related to cpap...you probably need to see an ENT doctor and have the eustachian tubes evaluated. The tubes shouldn't be allowing the pressure into the inner ear but I have heard that sometimes it happens.
Thanks for the info. I'm not waking often during the night. During the day more like lightheadedness than vertigo. Will post a few more charts tomorrow.
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Sat Jan 02, 2021 7:14 pm

screenshot-20210102-172715.png
screenshot-20210102-172715.png (63.21 KiB) Viewed 1157 times
screenshot-20210102-161257.png
screenshot-20210102-161257.png (65.29 KiB) Viewed 1157 times
Attachments
screenshot-20210102-172043.png
screenshot-20210102-172043.png (66.52 KiB) Viewed 1157 times
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: How do my machine readings look?

Post by Muse-Inc » Sat Jan 02, 2021 7:47 pm

Make sure the vertigo is not a reflection of dehydration. I get that if I am just slightly dehydrated. I keep a large, handled mug by my side to remember to drink a swallow or 2 often.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.

tbeb
Posts: 54
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: How do my machine readings look?

Post by tbeb » Mon Jan 04, 2021 12:27 am

Muse-Inc wrote:
Sat Jan 02, 2021 7:47 pm
Make sure the vertigo is not a reflection of dehydration. I get that if I am just slightly dehydrated. I keep a large, handled mug by my side to remember to drink a swallow or 2 often.
Thanks. I don't think it's due to dehydration but I did start drinking more water a couple weeks ago.
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: How do my machine readings look?

Post by zonker » Mon Jan 04, 2021 12:18 pm

tbeb wrote:
Mon Jan 04, 2021 12:27 am
Muse-Inc wrote:
Sat Jan 02, 2021 7:47 pm
Make sure the vertigo is not a reflection of dehydration. I get that if I am just slightly dehydrated. I keep a large, handled mug by my side to remember to drink a swallow or 2 often.
Thanks. I don't think it's due to dehydration but I did start drinking more water a couple weeks ago.
did the motion sickness pill work?

sorry if you've said this and i missed it.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win