New Member

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mh5591
Posts: 2
Joined: Sat Oct 17, 2020 8:20 am

New Member

Post by mh5591 » Sat Oct 17, 2020 10:21 am

Thank you for approval for website. I'm having a problem with my resmed 10 pulsating. I recently replaced cushion, tubing etc. No leaks, but when either I inhale or exhale air pulsates high then low. I am a long time cpap user, 20+ years and never had this problem with my resmed 10.

User avatar
Pugsy
Posts: 65390
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Member

Post by Pugsy » Sat Oct 17, 2020 10:35 am

How many run hours on the machine?

Check to make sure the hose diameter selection choice in the setup menu matches the hose diameter that you are using.

If using a heated hose it should automatically detect it and it knows the diameter but if using a non heated hose the diameter needs to match.

https://www.respshop.com/manuals/ResMed ... %20her.pdf

Try unplugging the machine from electric source and then plug back in....sometimes the forced reboot fixes problems.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mh5591
Posts: 2
Joined: Sat Oct 17, 2020 8:20 am

Re: New Member

Post by mh5591 » Sat Oct 17, 2020 1:11 pm

Thank you, I will try unplug/ plug reboot tonight!! I have a heated hose and this machine was acquired July 2018

tim7388
Posts: 1
Joined: Fri Jun 04, 2021 9:10 pm

Re: New Member

Post by tim7388 » Fri Jun 04, 2021 9:16 pm

mh5591 wrote:
Sat Oct 17, 2020 10:21 am
Thank you for approval for website. I'm having a problem with my resmed 10 pulsating. I recently replaced cushion, tubing etc. No leaks, but when either I inhale or exhale air pulsates high then low. I am a long time cpap user, 20+ years and never had this problem with my resmed 10.
I’m having the same problem with AirSense10 pulsating. The noise wakes mg wife up. My machine was replaced but it didn’t solve the issue. I’ve been using the CPAP for seven years and never had a problem before. Lincare, who services my supplies and machine, are baffled. I have an appointment for next Tuesday to try to solve the problem again. I will let you know what turns out.