Struggling with Seal

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
IRSIV
Posts: 7
Joined: Fri Sep 18, 2020 11:55 am

Struggling with Seal

Post by IRSIV » Fri Sep 18, 2020 11:57 am

Hi -

I’ve been on CPAP for a couple of months, and am struggling with mask seal. I’m a solidly built 6 foot 200# man. Don’t smoke, rare drinker, and take an antidepressant in the morning.I’m a side sleeper and have a deviated septum (mouth breather) and a beard and mustache. I fall asleep quickly with or without the mask.

I started with an AirSense 10 machine and AirTouch F20 with memory foam which would seal great maybe 1 day out of 7.

I went to a local CPAP store and tried a dozen masks there, all of which were leaky according to the assistant, especially around the bridge of my nose and apparently deep eye sockets.

I ordered a Respironics FitLife Total Face mask to avoid that problem area, and it seals well if I lie still only on my back, but unseats as soon as I end up on my side. And even the XL size tries to ride up my chin over my mouth.

In general CPAP has been ruining my sleep, waking me up relentlessly for two to four hours before I chuck it off, and I feel more tired than I ever did before, even though I have what is severe apnea with 47 events per hour in my at home sleep test. The CPAP does get that down under 9 when it is working (or maybe because I'm NOT sleeping), so I do want to try to push through. The machine averages 9 and peaks at 14 psi according to the monitor, and I’m using the ramp, exhale relief, and heated humidity.

I feel that without an "in-person" sleep doctor I'm really just shooting in the dark. Shoulc I push for that? Any ideas on what to try next?

Thanks.

_________________
MachineMask
Last edited by IRSIV on Fri Sep 18, 2020 12:37 pm, edited 1 time in total.

User avatar
LSAT
Posts: 13349
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Struggling with Seal

Post by LSAT » Fri Sep 18, 2020 12:32 pm


User avatar
IRSIV
Posts: 7
Joined: Fri Sep 18, 2020 11:55 am

Re: Struggling with Seal

Post by IRSIV » Fri Sep 18, 2020 12:39 pm

LSAT wrote:
Fri Sep 18, 2020 12:32 pm
Take a good look at this....https://www.cpap.com/productpage/bleep- ... -cpap-mask
Thanks - but I can really only breathe through one nostril.

_________________
MachineMask

User avatar
ChicagoGranny
Posts: 15357
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Struggling with Seal

Post by ChicagoGranny » Fri Sep 18, 2020 1:31 pm

IRSIV wrote:
Fri Sep 18, 2020 11:57 am
struggling with mask seal
For the masks you own, have you watched three or four fitting and adjusting videos on youtube?
IRSIV wrote:
Fri Sep 18, 2020 11:57 am
I feel that without an "in-person" sleep doctor I'm really just shooting in the dark.
The best sleep doctors shoot in the dark when it comes to masks. The others don't bother to load the gun. Finding, fitting and adjusting the right mask is pretty much a DIY process. Go to cpap.com, find the full face mask section, and look at some masks and read the reviews. Stick with the most popular masks.
IRSIV wrote:
Fri Sep 18, 2020 12:39 pm
I can really only breathe through one nostril.
Have a consultation with an ENT and get this corrected. It will make using CPAP easier, and you also will breathe healthier during the day.
IRSIV wrote:
Fri Sep 18, 2020 11:57 am
Struggling with Seal
~.JPG
~.JPG (22.98 KiB) Viewed 2122 times

User avatar
poppi2
Posts: 590
Joined: Thu Nov 18, 2010 2:54 pm
Location: Houston, near JSC

Re: Struggling with Seal

Post by poppi2 » Sat Sep 19, 2020 11:22 am

ChicagoGranny wrote:
Fri Sep 18, 2020 1:31 pm
IRSIV wrote:
Fri Sep 18, 2020 11:57 am
Struggling with Seal
~.JPG
Exactly my initial thought.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 12.0 - 14.0 cm

User avatar
ChicagoGranny
Posts: 15357
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Struggling with Seal

Post by ChicagoGranny » Sun Sep 20, 2020 5:03 am

poppi2 wrote:
Sat Sep 19, 2020 11:22 am
ChicagoGranny wrote:
Fri Sep 18, 2020 1:31 pm
IRSIV wrote:
Fri Sep 18, 2020 11:57 am
Struggling with Seal
~.JPG
Exactly my initial thought.
Was this your second thought?
~.JPG
~.JPG (11.01 KiB) Viewed 2056 times

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Struggling with Seal

Post by chunkyfrog » Sun Sep 20, 2020 10:53 am

Heidi Klum knows karate?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Miss Emerita
Posts: 3758
Joined: Sun Nov 04, 2018 8:07 pm

Re: Struggling with Seal

Post by Miss Emerita » Sun Sep 20, 2020 11:25 am

IRSIV, is it possible you'd do better with a hybrid-type mask, like the Mirage Liberty? It's expensive, but if you buy it from cpap.com, they'll give you a refund if it doesn't work and you send it back to them.

You may already know this, but it's healthier to breathe through the nose, because the nose filters out lots of stuff before it delivers the air to your lungs. The nasal sinuses also produce nitric oxide, which we inhale when we breathe through the nose. The NO in turn helps us fight bacterial infections, improves arterial oxygenation, and decreases blood pressure in the lungs. Some more information here:

https://pubmed.ncbi.nlm.nih.gov/8971255/

So these would all be additional reasons to see whether an ENT could correct your deviated septum and make it possible for you to breathe through your nose.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Struggling with Seal

Post by klm49 » Sun Sep 20, 2020 11:38 am

IRSIV wrote:
Fri Sep 18, 2020 11:57 am
Hi -

I’ve been on CPAP for a couple of months, and am struggling with mask seal. I’m a solidly built 6 foot 200# man. Don’t smoke, rare drinker, and take an antidepressant in the morning.I’m a side sleeper and have a deviated septum (mouth breather) and a beard and mustache. I fall asleep quickly with or without the mask.

I started with an AirSense 10 machine and AirTouch F20 with memory foam which would seal great maybe 1 day out of 7.

I went to a local CPAP store and tried a dozen masks there, all of which were leaky according to the assistant, especially around the bridge of my nose and apparently deep eye sockets.

I ordered a Respironics FitLife Total Face mask to avoid that problem area, and it seals well if I lie still only on my back, but unseats as soon as I end up on my side. And even the XL size tries to ride up my chin over my mouth.

In general CPAP has been ruining my sleep, waking me up relentlessly for two to four hours before I chuck it off, and I feel more tired than I ever did before, even though I have what is severe apnea with 47 events per hour in my at home sleep test. The CPAP does get that down under 9 when it is working (or maybe because I'm NOT sleeping), so I do want to try to push through. The machine averages 9 and peaks at 14 psi according to the monitor, and I’m using the ramp, exhale relief, and heated humidity.

I feel that without an "in-person" sleep doctor I'm really just shooting in the dark. Shoulc I push for that? Any ideas on what to try next?

Thanks.


Have you considered shaving? If I don't shave for 3 or 4 days I start having seal problems. I don't see how anyone with a beard could get a seal, even though some claim to.






You'll have to ignore the clown posts, they can't help it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: Struggling with Seal

Post by zonker » Sun Sep 20, 2020 11:45 am

klm49 wrote:
Sun Sep 20, 2020 11:38 am


You'll have to ignore the clown posts, they can't help it.
and we'll ignore your concern troll comments as well.

you can't help it.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Struggling with Seal

Post by Pugsy » Sun Sep 20, 2020 12:55 pm

Is that you in your avatar?

If it is....you might want to rethink nasal pillow mask of some sort. If you can move enough air to be comfortable through one nostril...they still will work. You are going to have major issues with any mask that has to come in contact with that much facial hair.
Now it can be done but it's going to require a lot of work.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Struggling with Seal

Post by Pugsy » Sun Sep 20, 2020 12:58 pm

Enough with the snarky comments.

Stay on topic please. This is a newbie and a helping thread....please don't derail it any further.

Thank you in advance for your help in this matter.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Struggling with Seal

Post by chunkyfrog » Sun Sep 20, 2020 1:15 pm

Back on topic: the Respironics FitLife Total Face mask would fit a whole lot better
with an inflatable cushion or even a memory foam gasket.
I have considered cutting the silicone off an old one and gluing in something better--or custom.
Let's face it--the factory seals leave much to be desired.
Of course, this from the ultimate diy frogs.
My pandemic PPE mask of choice is LEATHER--kind of a road warrior vibe.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Struggling with Seal

Post by Okie bipap » Sun Sep 20, 2020 1:16 pm

Due to a mistake on the part of my first doctor, I had to have four sleep tests. During the first two I had a full beard and they had a very hard time getting a mask to seal. They tried a mask liner and that helped enough to finish the tessts. Prior to my last two tests, the sleep clinic suggested I shave my beard prior to testing. That was over five years ago and I have been clean shaven ever since. After having a beard for over thirty years, it was hard to give it up, but I have been clean shaven ever since. Apparently, some people are able to get a good seal with a beard. I think it depends on how long the beard is and how thick it is.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: Struggling with Seal

Post by zonker » Sun Sep 20, 2020 1:33 pm

Pugsy wrote:
Sun Sep 20, 2020 12:58 pm
Enough with the snarky comments.

Stay on topic please. This is a newbie and a helping thread....please don't derail it any further.

Thank you in advance for your help in this matter.
yes, m'am.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win