New Here - Questions About Pillows and Problems Exhaling

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 12:05 pm

Hi all, I'm new here. I received my CPAP machine last week and started trying to use it and am having problems. Seems when I go to bed, I am sleepy. After donning the mask, which are pillows, I become wide awake because of the breathing. Inhaling isn't too bad. Exhaling is not fun at all. Exhaling against forced air is difficult. I find myself inhaling deeply to compensate. I cannot adjust the flow of air either. Is there something I can do, or do I just need to get used to it by using it periodically during the day.

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: New Here - Questions About Pillows and Problems Exhaling

Post by chunkyfrog » Thu Sep 03, 2020 12:12 pm

EPR (exhalation pressure relief) can help with that,
but most of us have the same experience at first.
Your muscles will soon adapt, and it will feel normal.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

Re: New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 12:16 pm

How do you adjust the EPR?

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Here - Questions About Pillows and Problems Exhaling

Post by Pugsy » Thu Sep 03, 2020 12:40 pm

You may have to turn on the access to EPR.
The manual explains how.
https://www.respshop.com/manuals/ResMed ... %20her.pdf

Now bear in mind that EPR only works if the pressure setting is above 4 cm because the machine can't go below 4 cm.
So I don't know what your current settings are but using say 6 cm with EPR of 2...is going to give you 6 inhale and 4 exhale and it may sound weird but that's easier to do than just 4 for inhale and 4 for exhale without EPR.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

Re: New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 1:04 pm

I don't think that I can adjust the EPR on mine. I've gone through all the settings and it's nowhere. Maybe it has been blocked or something?

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Here - Questions About Pillows and Problems Exhaling

Post by Pugsy » Thu Sep 03, 2020 1:08 pm

You may have to go into the clinical setup menu area and the manual explains how to do that.
It's just a combination of button pushing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

Re: New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 1:18 pm

I was able to go into the Clinical settings mode. Min pressure was set to 6.0. Max pressure was set to 14.0. I was able to change the settings so that both min and max pressures were 6.0. Then I went out of the settings and it looked like it didn't keep my settings. Went back to 6.0 and 14.0. Just realized that min and max pressure were not what you were talking about. I found the EPR settings. I could only set them from 1,2, or 3. Couldn't go to 6.

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.

User avatar
Dog Slobber
Posts: 4274
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: New Here - Questions About Pillows and Problems Exhaling

Post by Dog Slobber » Thu Sep 03, 2020 1:39 pm

Nobody told you to set EPR to 6. We are all aware that EPR is limited to off, 1, 2 or 3.

Pugsy was explaining to you that EPR will never drop below 4, and she gave an example of using a pressure of 6 with EPR of 2.
Battery Backup: EcoFlow Delta 2

MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

Re: New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 1:41 pm

Dog Slobber wrote:
Thu Sep 03, 2020 1:39 pm
Nobody told you to set EPR to 6. We are all aware that EPR is limited to off, 1, 2 or 3.

Pugsy was explaining to you that EPR will never drop below 4, and she gave an example of using a pressure of 6 with EPR of 2.
Guess I misunderstood. My apologies.

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.

User avatar
Pugsy
Posts: 65147
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Here - Questions About Pillows and Problems Exhaling

Post by Pugsy » Thu Sep 03, 2020 2:25 pm

The only change I wanted you to make to anything was turn EPR on to full time and set it to 2.
Leave everything the same.
We may have a problem though if the supplier wrote the settings through ResScan and they set EPR to be off and the software directions are on the SD card itself. It can be done with ResMed's software.

So if you change EPR to on and set it to 2 and it doesn't hold we may have to erase the SD card to eliminate the SD card settings overriding the manual setting change of turning EPR on and the setting change. The reason I mention this is you say you changed the min and max to 6 but it didn't stick. That tells me the SD card may contain some settings via software that make it more difficult for you to change things. We can fix that though if that is the case. Just erase the SD card and start fresh. The important summary data that the DME might need will just get put back on the SD card when it is put back in the machine. They are probably using MyAir anyway.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

MDSnyder3
Posts: 12
Joined: Thu Sep 03, 2020 4:41 am

Re: New Here - Questions About Pillows and Problems Exhaling

Post by MDSnyder3 » Thu Sep 03, 2020 2:45 pm

Pugsy wrote:
Thu Sep 03, 2020 2:25 pm
The only change I wanted you to make to anything was turn EPR on to full time and set it to 2.
Leave everything the same.
We may have a problem though if the supplier wrote the settings through ResScan and they set EPR to be off and the software directions are on the SD card itself. It can be done with ResMed's software.

So if you change EPR to on and set it to 2 and it doesn't hold we may have to erase the SD card to eliminate the SD card settings overriding the manual setting change of turning EPR on and the setting change. The reason I mention this is you say you changed the min and max to 6 but it didn't stick. That tells me the SD card may contain some settings via software that make it more difficult for you to change things. We can fix that though if that is the case. Just erase the SD card and start fresh. The important summary data that the DME might need will just get put back on the SD card when it is put back in the machine. They are probably using MyAir anyway.
Understood, thanks.

_________________
MachineMask
Additional Comments: My humidifier is a HumidAir Heated humidifier. That product is not listed in the Humidifier choices.