I have a Dreamstation CPAP. I'd like to homebrew a battery backup. The power supply cable has three wires, black, white, blue. Does anyone know how to wire this to a 2 post battery? I was planning to use power poles to battery clamps, but I am unsure how to move from the 3 wires to 2.
Thanks
Dreamstation DIY Battery backup
- Dog Slobber
- Posts: 4325
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Dreamstation DIY Battery backup
Wire a cigarette adapter on to the battery, then use a DC cord specifically designed for the DreamStation.
$30 bucks at CPAP.COM
https://www.cpap.com/productpage/pr-dre ... ed-dc-cord
You can also get cheaper off brand DC cords.
$30 bucks at CPAP.COM
https://www.cpap.com/productpage/pr-dre ... ed-dc-cord
You can also get cheaper off brand DC cords.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Dreamstation DIY Battery backup
Thanks for the advice. I have both ordered.
I'd still like to know how to take a regular Dreamstation power supply, cut the cable that connects to the Dreamstation at the power supp and wire that to battery clamps.
Again, thank you for the input.
I'd still like to know how to take a regular Dreamstation power supply, cut the cable that connects to the Dreamstation at the power supp and wire that to battery clamps.
Again, thank you for the input.
- Dog Slobber
- Posts: 4325
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Dreamstation DIY Battery backup
http://www.apneaboard.com/forums/Thread ... ery-Cables
3 conductor cable:
Outer Barrel: Common
Inner Barrel: +12v
Centre Pin: 20K ohm resister +12v . In the topic there are ambiguities regarding resister. Originally claim to be 100K, then 20K, then 24.3K.
The connector is similar to that found on many laptop chargers.
3 conductor cable:
Outer Barrel: Common
Inner Barrel: +12v
Centre Pin: 20K ohm resister +12v . In the topic there are ambiguities regarding resister. Originally claim to be 100K, then 20K, then 24.3K.
The connector is similar to that found on many laptop chargers.