OSCAR: "CPAP Data Located" NOT

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Karen Avery
Posts: 6
Joined: Sat Jun 20, 2020 2:25 pm

OSCAR: "CPAP Data Located" NOT

Post by Karen Avery » Thu Jun 25, 2020 9:26 am

New CPAP user since June 16, 2020.

I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR - the process at http://www.apneaboard.com/wiki/index.ph ... FIRST_time stops, and I never see the following "CPAP Data Located" message.
CPAP_data_located.png
CPAP_data_located.png (13.18 KiB) Viewed 4458 times

The http://www.apneaboard.com/wiki/index.ph ... leshooting troubleshooting page has a blank "Data won't load correctly or at all" section http://www.apneaboard.com/wiki/index.ph ... _at_all.29.

Thoughts, suggestions, advice?
Thank you! Karen

_________________
MachineMask
"Karen"

User avatar
khauser
Posts: 1308
Joined: Fri Feb 01, 2013 9:31 am
Location: Nashua, NH

Re: OSCAR: "CPAP Data Located" NOT

Post by khauser » Thu Jun 25, 2020 9:41 am

Did you click the Import button?

In Linux I don't believe it would show as a drive letter. It'd probably be something like /mnt/SD. though I believe the name is up to you. I am also not sure Linux auto-mounts SD cards as Windows does ... you may have to manually mount it with the mount command.

Once you're sure it is mounted, you should be able to click import and navigate to where the root of the SD card is mounted.

-Kevin

_________________
MachineMask
Sleep loss is a terrible thing. People get grumpy, short-tempered, etc. That happens here even among the generally friendly. Try not to take it personally.

User avatar
zonker
Posts: 11496
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR: "CPAP Data Located" NOT

Post by zonker » Thu Jun 25, 2020 11:55 am

Karen Avery wrote: ↑
Thu Jun 25, 2020 9:26 am
New CPAP user since June 16, 2020.
welcome to the zoo!

as far as i can tell, you may be the first person on the forum to try to run oscar on a chromebook, though maybe not the first person to run it on linux.

you may get better results posting in this thread-

viewtopic/t179460/OSCAR-v111-is-Released.html

the authors are on here regularly and may "see" your post there first.

good luck!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Karen Avery
Posts: 6
Joined: Sat Jun 20, 2020 2:25 pm

Re: OSCAR: "CPAP Data Located" NOT

Post by Karen Avery » Thu Jun 25, 2020 12:14 pm

zonker wrote:
Karen Avery wrote: ↑
Thu Jun 25, 2020 9:26 am
New CPAP user since June 16, 2020.
welcome to the zoo!

as far as i can tell, you may be the first person on the forum to try to run oscar on a chromebook, though maybe not the first person to run it on linux.

you may get better results posting in this thread-

viewtopic/t179460/OSCAR-v111-is-Released.html

the authors are on here regularly and may "see" your post there first.

good luck!!
I was hoping you'd chime in, Zonker. Thank you, Arlen

Sent from my MHA-L29 using Tapatalk


_________________
MachineMask
"Karen"

User avatar
BlueDragon
Posts: 571
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR: "CPAP Data Located" NOT

Post by BlueDragon » Thu Jun 25, 2020 1:01 pm

Karen Avery wrote: ↑
Thu Jun 25, 2020 9:26 am
I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR ....
Which version of OSCAR did you download? Chromebooks are not yet on our list of supported computers although support for them is high on the wish list. I'll direct one of the developers familiar with linux and chromebook to your question.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
zonker
Posts: 11496
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR:

Post by zonker » Thu Jun 25, 2020 5:27 pm

Karen Avery wrote: ↑
Thu Jun 25, 2020 12:14 pm
I was hoping you'd chime in, Zonker. Thank you, Arlen

Sent from my MHA-L29 using Tapatalk
or just wait right here and blue dragon will answer!
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
BlueDragon
Posts: 571
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR: "CPAP Data Located" NOT

Post by BlueDragon » Thu Jun 25, 2020 6:30 pm

I've sent you a PM (private message) about helping test a Chromebook version of OSCAR.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
zonker
Posts: 11496
Joined: Fri Jun 19, 2015 4:36 pm

Re: OSCAR: "CPAP Data Located" NOT

Post by zonker » Thu Jun 25, 2020 6:50 pm

BlueDragon wrote: ↑
Thu Jun 25, 2020 6:30 pm
I've sent you a PM (private message) about helping test a Chromebook version of OSCAR.
kewel! if you can get a version going for chromebook(in the far future, natch) i'd be interested.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
bonjour
Posts: 322
Joined: Thu May 14, 2009 5:39 pm

Re: OSCAR: "CPAP Data Located" NOT

Post by bonjour » Thu Jun 25, 2020 7:23 pm

Karen,
one of the projects I'm working on is cleaning up the Wiki for installing OSCAR on Linux, on a Chromebook. We know it runs on Chromebook, no question, but we need solid instructions on how-to that can be followed by novices. Linux Pros don't seem too major problems. (I'm sure there are small things they can easily solve). This needs to be solid before we announce it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: VAuto 11-25cmH2O PS 3 --- OSCAR software, Many masks - Amara View FFM to P10 Pillows Several Nasal and FFM

Karen Avery
Posts: 6
Joined: Sat Jun 20, 2020 2:25 pm

Re: OSCAR: "CPAP Data Located" NOT

Post by Karen Avery » Fri Jun 26, 2020 8:35 am

I would love to help you betatest anything you're working on. BlueDragon PM'ed me about the Apnea Board Forum discussion about OSCAR and Chromebooks, too, so I've signedup there as well.
"Karen"
bonjour wrote: ↑
Thu Jun 25, 2020 7:23 pm
Karen,
one of the projects I'm working on is cleaning up the Wiki for installing OSCAR on Linux, on a Chromebook. We know it runs on Chromebook, no question, but we need solid instructions on how-to that can be followed by novices. Linux Pros don't seem too major problems. (I'm sure there are small things they can easily solve). This needs to be solid before we announce it.

_________________
MachineMask
"Karen"

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: OSCAR: "CPAP Data Located" NOT

Post by palerider » Thu Jul 02, 2020 12:20 pm

khauser wrote: ↑
Thu Jun 25, 2020 9:41 am
Did you click the Import button?

In Linux I don't believe it would show as a drive letter. It'd probably be something like /mnt/SD. though I believe the name is up to you. I am also not sure Linux auto-mounts SD cards as Windows does ... you may have to manually mount it with the mount command.

Once you're sure it is mounted, you should be able to click import and navigate to where the root of the SD card is mounted.

-Kevin
plus, Linux doesn't use backslashes

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mjphyi
Posts: 1
Joined: Sun Jan 12, 2020 8:44 am

Re: OSCAR: "CPAP Data Located" NOT

Post by mjphyi » Tue Jul 07, 2020 1:23 pm

Karen Avery wrote: ↑
Thu Jun 25, 2020 9:26 am
New CPAP user since June 16, 2020.

I've downloaded OSCAR (onto a Chromebook), gave LINUX permission to the locked SD card, filled out the profile, etc. The data, however, is not being recognized by OSCAR - the process at http://www.apneaboard.com/wiki/index.ph ... FIRST_time stops, and I never see the following "CPAP Data Located" message. CPAP_data_located.png
The http://www.apneaboard.com/wiki/index.ph ... leshooting troubleshooting page has a blank "Data won't load correctly or at all" section http://www.apneaboard.com/wiki/index.ph ... _at_all.29.

Thoughts, suggestions, advice?
Thank you! Karen
It's good that you gave Linux permission to access the SD Card and indeed managed to install on a Chromebook. Have you seen this wiki page before?
http://www.apneaboard.com/wiki/index.ph ... _Chrome_OS

It contains a header that reads There are 5 topics we need to deal with here:
and you seem to have managed the first 4 steps, but stumbled at the 5th.

That section has the header
Importing SD Card Data

OSCAR is not currently capable of finding your SD Card for you, you have to tell it where it is.

Try working your way through that section, and it should all work. Post back here if you get stuck.

Michael