First night not looking so good

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
corvous256
Posts: 7
Joined: Sun May 10, 2020 11:14 am

First night not looking so good

Post by corvous256 » Sat Jun 06, 2020 11:46 pm

First night back on therapy after a 5 year break. Originally prescribed a pressure of 8 cmH20. Now I have a ResMed S10 auto set to 6 - 12. I noticed that my time seems to be off by about 12 hours. Where can I adjust the time to make it correct?

I attached a screenshot from Oscar. Can someone please help me to understand why the centrals are so high and what changes I might need to make to my machine to improve this AHI.

Thanks to everyone in advance.
screenshot-20200607-123304.png
screenshot-20200607-123304.png (127.91 KiB) Viewed 1653 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: First night not looking so good

Post by Julie » Sun Jun 07, 2020 9:04 am

Fill out Profile with equipment (not just e.g. "Resmed") so we have what to consider. Any reason your max pressure's only at 12?

User avatar
Pugsy
Posts: 65279
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: First night not looking so good

Post by Pugsy » Sun Jun 07, 2020 9:08 am

There is a clock and calendar on your machine that can be set manually.
Explained in the combo manual.
https://www.respshop.com/manuals/ResMed ... %20her.pdf

Turn off EPR or use it in Ramp only mode and make use of the ramp feature (explained in the manual).

You also probably need to increase the minimum and maximum a bit but I suggest one change at a time and the first change being turning EPR off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

corvous256
Posts: 7
Joined: Sun May 10, 2020 11:14 am

Re: First night not looking so good

Post by corvous256 » Sun Jun 07, 2020 9:41 am

Julie wrote:
Sun Jun 07, 2020 9:04 am
Fill out Profile with equipment (not just e.g. "Resmed") so we have what to consider. Any reason your max pressure's only at 12?
Changed my profile. Thanks for the pointer. The only reason the max pressure was set at 12 is because I never needed above 11 before on my last machine but obviously things have changed over the years.

corvous256
Posts: 7
Joined: Sun May 10, 2020 11:14 am

Re: First night not looking so good

Post by corvous256 » Sun Jun 07, 2020 9:43 am

Pugsy wrote:
Sun Jun 07, 2020 9:08 am
There is a clock and calendar on your machine that can be set manually.
Explained in the combo manual.
https://www.respshop.com/manuals/ResMed ... %20her.pdf

Turn off EPR or use it in Ramp only mode and make use of the ramp feature (explained in the manual).

You also probably need to increase the minimum and maximum a bit but I suggest one change at a time and the first change being turning EPR off.
Thanks for the advice. I will turn EPR off and up the maximum pressure to 15 while I am in the settings.

MFRep
Posts: 5
Joined: Sun Jun 07, 2020 11:09 am

Re: First night not looking so good

Post by MFRep » Sun Jun 07, 2020 11:32 am

Im trying to recall, but centrals can be caused by an increase in C02, as caused by........too much or too little
variation pressure???

_________________
MachineMask

User avatar
Pugsy
Posts: 65279
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: First night not looking so good

Post by Pugsy » Sun Jun 07, 2020 11:56 am

MFRep wrote:
Sun Jun 07, 2020 11:32 am
Im trying to recall, but centrals can be caused by an increase in C02, as caused by........too much or too little
variation pressure???
Actually centrals are caused by too little carbon dioxide in the blood stream the majority of the time.
Now what can cause too little carbon dioxide levels in the blood can vary a bit.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

corvous256
Posts: 7
Joined: Sun May 10, 2020 11:14 am

Re: First night not looking so good

Post by corvous256 » Sat Jun 13, 2020 5:01 pm

Sorry for to delay in posing new charts but I had a few nights where I didn't sleep but a couple of hours and I wasn't sure it the data would be useful. Here are my charts on nights that I slept for more than 2 hours. What do you all think?
Attachments
screenshot-20200614-055008.png
screenshot-20200614-055008.png (115.72 KiB) Viewed 1538 times
screenshot-20200614-054951.png
screenshot-20200614-054951.png (119.01 KiB) Viewed 1538 times
screenshot-20200614-054924.png
screenshot-20200614-054924.png (115.02 KiB) Viewed 1538 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: First night not looking so good

Post by Julie » Sat Jun 13, 2020 5:37 pm

What are the big gaps in the last one - on purpose, or ??

corvous256
Posts: 7
Joined: Sun May 10, 2020 11:14 am

Re: First night not looking so good

Post by corvous256 » Sun Jun 14, 2020 12:11 am

Julie wrote:
Sat Jun 13, 2020 5:37 pm
What are the big gaps in the last one - on purpose, or ??
Just woke up and had trouble going back to sleep so I decided to get up for a little while. I have fairly severe insomnia which has been a little more pronounced since starting CPAP.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: First night not looking so good

Post by Miss Emerita » Sun Jun 14, 2020 11:06 am

You've responded well to the changes in your settings. If I were you, I'd leave the settings where they are for a week or two. That may allow your body to adjust and bring down your CA numbers further. Later, it may be worth seeing whether some very gradual increases in your minimum would reduce the obstructive events. So, my take is "steady as you go."
Oscar software is available at https://www.sleepfiles.com/OSCAR/