ASV DATA in OSCAR

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SAM_S
Posts: 89
Joined: Thu Apr 13, 2017 2:36 pm
Location: NY

ASV DATA in OSCAR

Post by SAM_S » Mon May 18, 2020 10:26 am

Guys I need some help.I can not upload my ASV DATA to OSCAR.Trying for last 2 days with no luck.

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV DATA in OSCAR

Post by Pugsy » Mon May 18, 2020 10:52 am

A bit of history that I am aware of....S9 Adapt (ASV model 36037). That's all I know.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Mon May 18, 2020 10:53 am

Pugsy wrote:
Mon May 18, 2020 10:52 am
A bit of history that I am aware of....S9 Adapt (ASV model 36037). That's all I know.
???
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65366
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV DATA in OSCAR

Post by Pugsy » Mon May 18, 2020 10:56 am

zonker wrote:
Mon May 18, 2020 10:53 am
Pugsy wrote:
Mon May 18, 2020 10:52 am
A bit of history that I am aware of....S9 Adapt (ASV model 36037). That's all I know.
???
He's using the S9 Adapt and it won't upload or work with OSCAR and it should. Beyond that I don't know anything to trouble shoot and maybe help figure out what is going on. He just got the Adapt and wants to try it out.
Prior to this OSCAR worked just fine with the Respironics APAP machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Mon May 18, 2020 10:59 am

Pugsy wrote:
Mon May 18, 2020 10:56 am
zonker wrote:
Mon May 18, 2020 10:53 am
Pugsy wrote:
Mon May 18, 2020 10:52 am
A bit of history that I am aware of....S9 Adapt (ASV model 36037). That's all I know.
???
He's using the S9 Adapt and it won't upload or work with OSCAR and it should. Beyond that I don't know anything to trouble shoot and maybe help figure out what is going on. He just got the Adapt and wants to try it out.
Prior to this OSCAR worked just fine with the Respironics APAP machine.
thank you. i just couldn't read your response in any sort of context.

maybe i need to go back to bed?
8)
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Mon May 18, 2020 11:01 am

SAM_S wrote:
Mon May 18, 2020 10:26 am
Guys I need some help.I can not upload my ASV DATA to OSCAR.Trying for last 2 days with no luck.
can you explain exactly what the problem is? at what point is it failing?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Respiration99
Posts: 37
Joined: Fri Apr 24, 2020 4:55 pm

Re: ASV DATA in OSCAR

Post by Respiration99 » Tue May 19, 2020 4:47 am

From the limited information available, I suggest you create a new profile in Oscar for the S9 and keep the data separate from your Respironics data. Ie separate profiles for each machine.

The S9 VPAP Adapt works fine with Oscar - mine does, anyhow.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 VPAP Adapt

SAM_S
Posts: 89
Joined: Thu Apr 13, 2017 2:36 pm
Location: NY

Re: ASV DATA in OSCAR

Post by SAM_S » Thu May 21, 2020 1:35 pm

Thing is whenever I try to start OSCAR it got hanged. Before using the ASV S9 memory card It was working pretty well.I opened a second account for asv. Becoz the oscar doesnot taking the data from s9 ASv.Now i can not even open it.It hanged everytime I open it.I use windows 10.Plz help.

SAM_S
Posts: 89
Joined: Thu Apr 13, 2017 2:36 pm
Location: NY

Re: ASV DATA in OSCAR

Post by SAM_S » Fri May 22, 2020 10:06 am

guys plz help me.I can not put my ASV s9 Data to OSCAR

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Fri May 22, 2020 10:52 am

SAM_S wrote:
Fri May 22, 2020 10:06 am
guys plz help me.I can not put my ASV s9 Data to OSCAR
? respiration99 has given you an answer.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

SAM_S
Posts: 89
Joined: Thu Apr 13, 2017 2:36 pm
Location: NY

Re: ASV DATA in OSCAR

Post by SAM_S » Fri May 22, 2020 11:49 am

I create a new profile.But still when ever I open OSCAR it got hanged

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Fri May 22, 2020 1:38 pm

SAM_S wrote:
Fri May 22, 2020 11:49 am
I create a new profile.But still when ever I open OSCAR it got hanged
try re-installing oscar.

meanwhile, let's see if i can shout for help here-
Dog Slobber wrote:
Thu May 21, 2020 7:41 pm
hey, dog slobber? can you help this fellow out? thanks!
if i work this right, dog slobber will see this notification when he next logs in

<fingers crossed>
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4332
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ASV DATA in OSCAR

Post by Dog Slobber » Fri May 22, 2020 2:18 pm

zonker wrote:
Fri May 22, 2020 1:38 pm
meanwhile, let's see if i can shout for help here-
Dog Slobber wrote:
Thu May 21, 2020 7:41 pm
hey, dog slobber? can you help this fellow out? thanks!
if i work this right, dog slobber will see this notification when he next logs in

<fingers crossed>
LOL.

If OSCAR is hanging since trying to import the new data, but you can't remove the profile containing the corrupt data (because OSCAR hangs), then you'll need to remove the profile from outside of OSCAR.

In Windows explorer browse to Documents/OSCAR_Data/Profiles you will then see a directory for every profile. You should you can then delete the directory of the offending profile.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11465
Joined: Fri Jun 19, 2015 4:36 pm

Re: ASV DATA in OSCAR

Post by zonker » Fri May 22, 2020 4:07 pm

Dog Slobber wrote:
Fri May 22, 2020 2:18 pm

LOL.

If OSCAR is hanging since trying to import the new data, but you can't remove the profile containing the corrupt data (because OSCAR hangs), then you'll need to remove the profile from outside of OSCAR.

In Windows explorer browse to Documents/OSCAR_Data/Profiles you will then see a directory for every profile. You should you can then delete the directory of the offending profile.
that's JUST what i was gonna say!
:lol: :lol: :lol:

thanks for the helping hand.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

SAM_S
Posts: 89
Joined: Thu Apr 13, 2017 2:36 pm
Location: NY

Re: ASV DATA in OSCAR

Post by SAM_S » Fri May 22, 2020 6:02 pm

thank you guys.I am going to do that.Thanks