New CPAP user with OSCAR data---High AHI

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Sat Apr 25, 2020 9:12 am

Hi There I am a new CPAP user.

I dunno how come my AHI is always high, Could any one tell me unless leak reason (I will change L size to M size tomorrow)

any other reason cause so high AHI? and flow rate and mask pressure is normal? other opinion for me to improve AHI? thanx a lot


AHI as below:
Attachments
圖片 44.jpg
圖片 44.jpg (199.94 KiB) Viewed 17285 times
圖片 43.jpg
圖片 43.jpg (199.9 KiB) Viewed 17285 times

User avatar
LSAT
Posts: 13406
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: New CPAP user with OSCAR data---High AHI

Post by LSAT » Sat Apr 25, 2020 9:25 am

Raise your minimum pressure to 15 and maximum to 20...Your AHI should come down

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New CPAP user with OSCAR data---High AHI

Post by Pugsy » Sat Apr 25, 2020 9:36 am

Please review this thread for image posting examples
viewtopic/t158560/How-to-post-images-for-review.html

The zoomed in view isn't particularly useful without the original for reference.

Why are you limiting the max to 17? Any particular reason?
It wants to go higher and can't...so it can't prevent the airway from collapsing.

Without seeing the overall detailed report I don't know if you need more minimum but you for sure need more maximum.
Your current minimum might be enough to get the job done if the maximum was higher.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

Re: New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Sat Apr 25, 2020 10:57 am

I try to Organised my OSCAR chart like this, please comment on this, thanx

圖片 46.jpg
圖片 46.jpg (202.41 KiB) Viewed 17266 times

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: New CPAP user with OSCAR data---High AHI

Post by Miss Emerita » Sat Apr 25, 2020 11:06 am

I will leave it to the experts to advise you about your pressure settings, just noting that your EPR seems to be set for ramp only.

You have several big clusters of events. Clusters could be caused by REM sleep, sleep position (e.g., on your back), or tucking your chin down toward your chest, which can restrict your airway and make it easier for it to close. Given that the second cluster follows pretty closely on the first, REM seems less likely as a culprit.

What position(s) do you sleep in? And do you happen to know whether you tuck your chin?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New CPAP user with OSCAR data---High AHI

Post by Pugsy » Sat Apr 25, 2020 11:08 am

Your machine desperately wants to go higher than 17 cm....so much so it essentially has set a new minimum itself as it never really goes below 14 or 15 once it goes up.
Pugsy wrote:
Sat Apr 25, 2020 9:36 am
Why are you limiting the max to 17? Any particular reason?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: New CPAP user with OSCAR data---High AHI

Post by Miss Emerita » Sat Apr 25, 2020 11:09 am

P.S. Could you fill out your profile with your machine and mask information? That way it'll be right there in case other people need to know it.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

Re: New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Sun Apr 26, 2020 10:44 am

Miss Emerita wrote:
Sat Apr 25, 2020 11:09 am
P.S. Could you fill out your profile with your machine and mask information? That way it'll be right there in case other people need to know it.
ok, thank you all to answer my question. i will do something to lower my AHI.

pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

Re: New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Sat May 30, 2020 7:21 pm

Miss Emerita wrote:
Sat Apr 25, 2020 11:06 am
I will leave it to the experts to advise you about your pressure settings, just noting that your EPR seems to be set for ramp only.

You have several big clusters of events. Clusters could be caused by REM sleep, sleep position (e.g., on your back), or tucking your chin down toward your chest, which can restrict your airway and make it easier for it to close. Given that the second cluster follows pretty closely on the first, REM seems less likely as a culprit.

What position(s) do you sleep in? And do you happen to know whether you tuck your chin?
Dear all:

I use cervical collar's cover already, but still need very high pressure value, is it normal? or I need to change to Vauto due to vauto can raise pressure to 25,

Please advise , thanx a lot
REM (2).jpg
REM (2).jpg (258.8 KiB) Viewed 17129 times

User avatar
zonker
Posts: 11463
Joined: Fri Jun 19, 2015 4:36 pm

Re: New CPAP user with OSCAR data---High AHI

Post by zonker » Sat May 30, 2020 7:38 pm

pcdvd.buyer wrote:
Sat May 30, 2020 7:21 pm


I use cervical collar's cover already, but still need very high pressure value, is it normal? or I need to change to Vauto due to vauto can raise pressure to 25,

Please advise , thanx a lot

REM (2).jpg
it's the minimum pressure right now that we are trying to get you to raise. whether your pressure is "very high" is relative.

my pressure, on your same machine, is set at 17.8 as minimum and 20 as the maximum. i don't go near 20. and you are only there part of the time as your graph shows.

raise the minimum to 17 as pugsy has mentioned. then leave it there for a few nights and see how it goes.

notice that your ahi is going down, at least in the graphs you've chosen to share with us. this process can take time, so please stick with it.

you can do this.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

Re: New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Tue Jun 02, 2020 7:05 am

Miss Emerita wrote:
Sat Apr 25, 2020 11:06 am
I will leave it to the experts to advise you about your pressure settings, just noting that your EPR seems to be set for ramp only.

You have several big clusters of events. Clusters could be caused by REM sleep, sleep position (e.g., on your back), or tucking your chin down toward your chest, which can restrict your airway and make it easier for it to close. Given that the second cluster follows pretty closely on the first, REM seems less likely as a culprit.

What position(s) do you sleep in? And do you happen to know whether you tuck your chin?
For my understanding is if I setting ramp only then my EPAP can set higher to deal with OA events,

like I have wrong concept. You mean enable EPR is better for AHI, right? thanks

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New CPAP user with OSCAR data---High AHI

Post by Pugsy » Tue Jun 02, 2020 7:24 am

When EPR is used the pressure on exhale drops by whatever setting is used. For some people that drop during exhale can mean the pressure drops below a level where the exhale pressure can continue to stent the airway open. The job of the pressure....hold that airway open and prevent collapse which ends up with some sort of apnea event occurring.

So does using EPR always mean that the drop equals a higher AHI because it allows apnea events to happen....no not always...but it can depending on how close a person is to that pressure setting where a person finds that below that line in the sand the airway cannot stay stented open.

When you use EPR set to ramp only then all it means is your pressure doesn't drop during exhale after ramp has ended. Your pressure still has to be high enough to hold/stent the airway open with or without any exhale relief.

Your pressure needs to hold your airway open are at time extremely high...at other times not quite so high.
When we see wide changes in pressure needs we wonder why there is a difference and there are a few usual reasons.
Supine sleeping...big common reason because it is well known to worsen OSA.
REM stage sleep...again very common because it is also well known to worsen OSA. I have this myself. Now REM normally comes in cycles and is fairly predictable....yours doesn't follow a typical REM pattern so that leaves sleeping position as being a likely culprit.

There are times when you need pressures above what your current machine can deliver...that's when you see the pressure maxed out at 20 cm. There are times when the machine doesn't have to go so high and does a decent job holding the airway open.

In your situation adding in that drop in exhale pressure by using EPR all night instead of ramp only would likely result in a higher AHI because I believe you are already very near the line where any drop during exhale will result in the airway collapsing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

pcdvd.buyer
Posts: 24
Joined: Sat Oct 12, 2019 5:59 am

Re: New CPAP user with OSCAR data---High AHI

Post by pcdvd.buyer » Fri Jun 12, 2020 4:57 pm

Pugsy wrote:
Tue Jun 02, 2020 7:24 am
When EPR is used the pressure on exhale drops by whatever setting is used. For some people that drop during exhale can mean the pressure drops below a level where the exhale pressure can continue to stent the airway open. The job of the pressure....hold that airway open and prevent collapse which ends up with some sort of apnea event occurring.

So does using EPR always mean that the drop equals a higher AHI because it allows apnea events to happen....no not always...but it can depending on how close a person is to that pressure setting where a person finds that below that line in the sand the airway cannot stay stented open.

When you use EPR set to ramp only then all it means is your pressure doesn't drop during exhale after ramp has ended. Your pressure still has to be high enough to hold/stent the airway open with or without any exhale relief.

Your pressure needs to hold your airway open are at time extremely high...at other times not quite so high.
When we see wide changes in pressure needs we wonder why there is a difference and there are a few usual reasons.
Supine sleeping...big common reason because it is well known to worsen OSA.
REM stage sleep...again very common because it is also well known to worsen OSA. I have this myself. Now REM normally comes in cycles and is fairly predictable....yours doesn't follow a typical REM pattern so that leaves sleeping position as being a likely culprit.

There are times when you need pressures above what your current machine can deliver...that's when you see the pressure maxed out at 20 cm. There are times when the machine doesn't have to go so high and does a decent job holding the airway open.

In your situation adding in that drop in exhale pressure by using EPR all night instead of ramp only would likely result in a higher AHI because I believe you are already very near the line where any drop during exhale will result in the airway collapsing.
yes, I enable EPR and Using an EPR of 3 to overcome the flow limitations, also I put on cervical collar's cover already, still high AHI.

How to improve it? What I can do? please advise, thanks
chang.jpg
chang.jpg (257.32 KiB) Viewed 17015 times

User avatar
Pugsy
Posts: 65360
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New CPAP user with OSCAR data---High AHI

Post by Pugsy » Fri Jun 12, 2020 5:11 pm

More minimum pressure to hold the airway open better and prevent the collapse in the first place.
Unfortunately you are fast coming to the point that your machine can't go any higher. You may need a bilevel machine that can go higher than 20 cm.
pcdvd.buyer wrote:
Fri Jun 12, 2020 4:57 pm
yes, I enable EPR and Using an EPR of 3 to overcome the flow limitations, also I put on cervical collar's cover already, still high AHI.

How to improve it? What I can do? please advise, thanks

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.