Suggestions for reducing AHI’s

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Suggestions for reducing AHI’s

Post by CorgiWalker » Mon Apr 20, 2020 8:36 am

I started CPAP in 2018. I averaged about 3.5 AHI’s for the first year on the machine and started at 80. I was morbidly obese. My repetitive breathing was increasing. So, I turned off the EPR which helped. My AHI was still creeping up, so I contacted my CPR. She increased my pressures from 7-15 to 10-15. Starting in Dec, I have lost 31 pounds. So,I am still obese. But, my AHI’s have continued to rise. I am now averaging around 7. I am changing my cushions regularly. My charts in Oscar are so busy that I cannot determine anything obvious. Any suggestions?

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suggestions for reducing AHI’s

Post by Pugsy » Mon Apr 20, 2020 8:48 am

Impossible to offer ideas on reducing AHI numbers without knowing what the 3 event categories that compose the AHI are.

Post a typical detail report per the guidelines in this thread.
viewtopic/t158560/How-to-post-images-for-review.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Suggestions for reducing AHI’s

Post by DreamDiver » Mon Apr 20, 2020 9:06 am

CorgiWalker wrote:
Mon Apr 20, 2020 8:36 am
I started CPAP in 2018. I averaged about 3.5 AHI’s for the first year on the machine and started at 80. I was morbidly obese. My repetitive breathing was increasing. So, I turned off the EPR which helped. My AHI was still creeping up, so I contacted my CPR. She increased my pressures from 7-15 to 10-15. Starting in Dec, I have lost 31 pounds. So,I am still obese. But, my AHI’s have continued to rise. I am now averaging around 7. I am changing my cushions regularly. My charts in Oscar are so busy that I cannot determine anything obvious. Any suggestions?
Why not post the last few days of daily charts? Hiding snore events usually helps declutter the flow rate chart.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Suggestions for reducing AHI’s

Post by CorgiWalker » Mon Apr 20, 2020 10:15 am

Here is the imgur link to 4/19 and 4/20 Oscar screenshots.
https://imgur.com/a/dkIQhug

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suggestions for reducing AHI’s

Post by Pugsy » Mon Apr 20, 2020 10:40 am

You have April 19 twice from what I can see.

I suspect you need more minimum pressure as well as more maximum.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

CorgiWalker
Posts: 20
Joined: Sat Jun 30, 2018 12:33 pm
Location: Alabama

Re: Suggestions for reducing AHI’s

Post by CorgiWalker » Mon Apr 20, 2020 11:15 am

Wow, it didn’t take the 2 different files. They aren’t very different, though. Thank you for the feedback, Pugsy! I will try upping and see what it does. Then, I can talk to the pulmonologist’s again.

User avatar
zonker
Posts: 11462
Joined: Fri Jun 19, 2015 4:36 pm

Re: Suggestions for reducing AHI’s

Post by zonker » Mon Apr 20, 2020 11:24 am

CorgiWalker wrote:
Mon Apr 20, 2020 11:15 am
Wow, it didn’t take the 2 different files. They aren’t very different, though. Thank you for the feedback, Pugsy! I will try upping and see what it does. Then, I can talk to the pulmonologist’s again.
if you'd like, you can post the pics right here, using the "attachments" command below.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg