Weird Stats - Is this normal?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Weird Stats - Is this normal?

Post by DreamDiver » Mon Apr 13, 2020 7:45 pm

Not sure any of these numbers circled in red should be negative. What's up with that?

Image

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Weird Stats - Is this normal?

Post by Pugsy » Mon Apr 13, 2020 7:53 pm

Software bug.

Might try creating a new profile and importing again.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Weird Stats - Is this normal?

Post by Pugsy » Mon Apr 13, 2020 8:04 pm

Statistic time frame is also totally messed up
December 1969 to April 2020...
Is this OSCAR or SleepyHead?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Weird Stats - Is this normal?

Post by Jas_williams » Mon Apr 13, 2020 11:42 pm

December 1969 was an import bug.in Oscar .. it will be fixed in the next release which is dew in the next few weeks

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

multicast
Posts: 30
Joined: Sun Nov 11, 2018 11:48 am

Re: Weird Stats - Is this normal?

Post by multicast » Tue Apr 14, 2020 2:53 am

Jas_williams wrote:
Mon Apr 13, 2020 11:42 pm
December 1969 was an import bug.in Oscar .. it will be fixed in the next release which is dew in the next few weeks
To fix that bug go to the date in Dezember 1969 and simply delete that entry. Works for me.

Mike

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Weird Stats - Is this normal?

Post by DreamDiver » Tue Apr 14, 2020 5:14 am

Thanks everyone for the suggestions. I'm going to start with Pugsy's suggestion to reimport data into a new profile.
Jas_williams wrote:
Mon Apr 13, 2020 11:42 pm
December 1969 was an import bug.in Oscar .. it will be fixed in the next release which is dew in the next few weeks
Thanks Jas, I thought it might be a bug. There's no data for any of
Pugsy wrote:
Mon Apr 13, 2020 8:04 pm
Statistic time frame is also totally messed up
December 1969 to April 2020...
Is this OSCAR or SleepyHead?
I imported data from Sleepyhead and from old manual card backups.
Scrolling through the overview, I can see that I've got respironics data from my old M-series Remstar CPAP that shouldn't exist in 2011, since I started that way back in 2007. Data are missing for anything earlier than 2011 and for a portion of 2016. So yeah... It's absolutely messed up. I should reimport. I think I have good data through 2017, so I'll just stop there.
multicast wrote:
Tue Apr 14, 2020 2:53 am
Jas_williams wrote:
Mon Apr 13, 2020 11:42 pm
December 1969 was an import bug.in Oscar .. it will be fixed in the next release which is dew in the next few weeks
To fix that bug go to the date in Dezember 1969 and simply delete that entry. Works for me.

Mike
I'd like to try deleting all data from 1969 through until 2017 for this profile, but I don't see how to do it. Is there a link you can share that shows the process? I couldn't find it in the wiki, nor in a search on CPAPtalk.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Weird Stats - Is this normal?

Post by DreamDiver » Tue Apr 14, 2020 7:16 am

Okay, I tried a new profile with the most recent S9 data and AS10 data and I still get the same wonkiness, though I'm not seeing the 1969 thing. It's only a year's worth of data, plus the data from the AS10. I suspect the problem may be how OSCAR is interpreting the data from the S9's SD card.

I tried making a profile with just the AS10 data, and the problem goes away for that profile. It looks like the problem is specific to S9.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

multicast
Posts: 30
Joined: Sun Nov 11, 2018 11:48 am

Re: Weird Stats - Is this normal?

Post by multicast » Fri Apr 17, 2020 3:34 am

DreamDiver wrote:
Tue Apr 14, 2020 5:14 am
I'd like to try deleting all data from 1969 through until 2017 for this profile, but I don't see how to do it. Is there a link you can share that shows the process? I couldn't find it in the wiki, nor in a search on CPAPtalk.
Go that day in December 69 in "Daily View". The Choose Data->Advanced->Clean actual day.

Mike

User avatar
ChicagoGranny
Posts: 15501
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Weird Stats - Is this normal?

Post by ChicagoGranny » Fri Apr 17, 2020 10:58 am

Jas_williams wrote:
Mon Apr 13, 2020 11:42 pm
the next release which is dew in the next few weeks
I'll tell the lawn service to wait until the grass dries.

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Weird Stats - Is this normal?

Post by DreamDiver » Sun Apr 19, 2020 8:49 am

multicast wrote:
Fri Apr 17, 2020 3:34 am
DreamDiver wrote:
Tue Apr 14, 2020 5:14 am
I'd like to try deleting all data from 1969 through until 2017 for this profile, but I don't see how to do it. Is there a link you can share that shows the process? I couldn't find it in the wiki, nor in a search on CPAPtalk.
Go that day in December 69 in "Daily View". The Choose Data->Advanced->Clean actual day.

Mike
Thanks, Mike! That actually worked. Now the first profile shows the correct number of days, starting with May 5, 2011.

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

User avatar
DreamDiver
Posts: 3082
Joined: Thu Oct 04, 2007 11:19 am

Re: Weird Stats - Is this normal?

Post by DreamDiver » Tue Apr 21, 2020 3:08 pm

Just upgraded to the latest version 1.1.0 and all the data transferred flawlessly. All the weird stats are now showing what look like good data.

Thanks OSCAR devs (@bonjour) !

Chris

_________________
Mask: ResMed AirFit™ F20 Mask with Headgear + 2 Replacement Cushions
Additional Comments: Pressure: APAP 10.4 | 11.8 | Also Quattro FX FF, Simplus FF
Image
Most members of this forum are wonderful.
However, if you are the target of bullying on this forum, please consider these excellent alternative forums:
Apnea Board
Sleep Apnea Talk Forum
Free CPAP Advice

Be well,
Chris

User avatar
zonker
Posts: 11449
Joined: Fri Jun 19, 2015 4:36 pm

Re: Weird Stats - Is this normal?

Post by zonker » Tue Apr 21, 2020 7:00 pm

DreamDiver wrote:
Tue Apr 21, 2020 3:08 pm
Just upgraded to the latest version 1.1.0 and all the data transferred flawlessly. All the weird stats are now showing what look like good data.

Thanks OSCAR devs (@bonjour) !

Chris
cheers!.gif
ain't it great that we have these hard working people in our corner?

and happy to hear your problem is solved.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

tirediceman
Posts: 4
Joined: Sun Dec 13, 2015 11:18 am

Re: Weird Stats - Is this normal?

Post by tirediceman » Thu Apr 23, 2020 10:17 am

MyAir website. Lately, my sleep time(usage hours) changes,(previously logged day after), sometimes logs usage on day not used. usually minimal 30+ mins or so. Any ideas? It does seem off about 30 mins a nite short in time. Not overlapping in timeframes.
Tks

_________________
MachineMask

User avatar
zonker
Posts: 11449
Joined: Fri Jun 19, 2015 4:36 pm

Re: Weird Stats - Is this normal?

Post by zonker » Thu Apr 23, 2020 10:45 am

tirediceman wrote:
Thu Apr 23, 2020 10:17 am
MyAir website. Lately, my sleep time(usage hours) changes,(previously logged day after), sometimes logs usage on day not used. usually minimal 30+ mins or so. Any ideas? It does seem off about 30 mins a nite short in time. Not overlapping in timeframes.
Tks
welcome to the zoo! please feel free to start your own thread. it's considered impolite to bomb someone else's.

please see the links at the bottom of my post for more info.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15501
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Weird Stats - Is this normal?

Post by ChicagoGranny » Thu Apr 23, 2020 11:00 am

tirediceman wrote:
Thu Apr 23, 2020 10:17 am
MyAir website. Lately, my sleep time(usage hours) changes,(previously logged day after), sometimes logs usage on day not used. usually minimal 30+ mins or so. Any ideas? It does seem off about 30 mins a nite short in time. Not overlapping in timeframes.
Tks
1. Is the time set correctly on your machine?

2. Do you sleep past noon? ResMed machines start a new reporting day at noon each day.

(I assume, with the latest release of OSCAR, Dream Diver is through with this thread.)