Questions about filters for DreamStation Auto CPAP

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Sun Apr 12, 2020 10:18 pm

I have four questions regarding my DreamStation Auto CPAP:

1. Is it possible/advisable to clean the ultra-fine pollen disposable filter instead of buying a new one twice a month?

2. Both filters take forever to air dry, especially in the winter, like more than a day. Is there any way to dry them faster? I am tired to using a hair dryer on them for 15 minutes to dry them befre I go to bed and buring my hand.

3. Does following the replacement schedule of every 6 months for the non-disposable foam filter and twice a month for the ultra-fine disposable filter really extend the life of the machine? Could I possibly double the usage times to save money without suffering adverse consequences?

4. What is your opinion on using non-genuine filters, like those bulk kits where you insert the filter into the frame yourself?

Thank you.

Carlitosbaby
Posts: 13
Joined: Thu May 16, 2019 7:00 pm
Location: New York

Re: Questions about filters for DreamStation Auto CPAP

Post by Carlitosbaby » Mon Apr 13, 2020 4:37 am

1. They arent washable
2.i wash main filter once a month when i wake and it's always dry by bedtime. How cold is your house?
3.Not sure but i change the micro filter on 1st of every month and main filter ive never changed in over a year. Just clean it on 1st of month and it still looks brand new 14 months later (wow 14 months already)
4. Never used them i buy 6 packs of original ones on ebay or Amazon they are cheap

_________________
MaskHumidifier

User avatar
LSAT
Posts: 13317
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Questions about filters for DreamStation Auto CPAP

Post by LSAT » Mon Apr 13, 2020 5:04 am

Who told you to change the filter twice a month? Change it when it's dirty.

fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Re: Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Mon Apr 13, 2020 2:53 pm

Carlitosbaby wrote:
Mon Apr 13, 2020 4:37 am
1. They arent washable
2.i wash main filter once a month when i wake and it's always dry by bedtime. How cold is your house?
3.Not sure but i change the micro filter on 1st of every month and main filter ive never changed in over a year. Just clean it on 1st of month and it still looks brand new 14 months later (wow 14 months already)
4. Never used them i buy 6 packs of original ones on ebay or Amazon they are cheap
Thanks for your help! I like the idea of changing the disposable ones on the first of each month.

As for drying them, my house is normal temperature. I do wash them in the morning, but they are still a little wet when I get ready for bed.

As for the original filters, you cannot buy them on Amazon. I checked. But I found the original Philips Respironics disposable ones for $1.29 each elsewhere, and the original non-disposable ones for $3.99 each, with no tax and free shipping, plus a 15% discount and a price match guarantee.

fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Re: Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Mon Apr 13, 2020 2:54 pm

LSAT wrote:
Mon Apr 13, 2020 5:04 am
Who told you to change the filter twice a month? Change it when it's dirty.
The Medicare replacement schedule and HHS and sleepapnea.org all suggest replacing the non-disposable filters twice a month. Otherwise, your warranty will be voided.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Questions about filters for DreamStation Auto CPAP

Post by Goofproof » Mon Apr 13, 2020 3:10 pm

When I wash my washable filter (It's black foam) after the very good rinse, I put it in a cotton towel and squeeze it dry as good as I can get it more than once. Then let it air dry, a fan will speed drying up if needed. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
raeraeiam
Posts: 1
Joined: Thu Apr 09, 2020 4:21 pm

Re: Questions about filters for DreamStation Auto CPAP

Post by raeraeiam » Mon Apr 13, 2020 3:24 pm

My instructions for cleaning the filters said to dab them with a dry paper towel after rinsing. I did that and it worked fine.
I also rinsed out the micro filter... Although I've only used my machine for 2 weeks so it wasn't very dirty yet. But it definitely looked cleaner after rinsing.

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Questions about filters for DreamStation Auto CPAP

Post by Okie bipap » Mon Apr 13, 2020 3:28 pm

fuchsia wrote:
Mon Apr 13, 2020 2:54 pm
The Medicare replacement schedule and HHS and sleepapnea.org all suggest replacing the non-disposable filters twice a month. Otherwise, your warranty will be voided.
This sounds like something a DME said in order to sell you more filters. Your users manual has directions for filter maintenance on page 24 which states the fine filter should be replaced after 30 nights of use or sooner if it gets dirty. Just because Medicare allows two filters a month does not mean you must change them that often. A little common sense goes a long way when dealing with DMEs.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Questions about filters for DreamStation Auto CPAP

Post by Goofproof » Mon Apr 13, 2020 3:45 pm

raeraeiam wrote:
Mon Apr 13, 2020 3:24 pm
My instructions for cleaning the filters said to dab them with a dry paper towel after rinsing. I did that and it worked fine.
I also rinsed out the micro filter... Although I've only used my machine for 2 weeks so it wasn't very dirty yet. But it definitely looked cleaner after rinsing.
When washing a filter that isn't made to be washed is a poor idea, many filters change their spec flow rate after washing, and no longer keep the machines working at their peak. :roll: Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Questions about filters for DreamStation Auto CPAP

Post by palerider » Mon Apr 13, 2020 4:37 pm

LSAT wrote:
Mon Apr 13, 2020 5:04 am
Who told you to change the filter twice a month? Change it when it's dirty.
they're tiny filters, they get dirty faster than the Resmed filters.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Re: Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Mon Apr 13, 2020 5:00 pm

Goofproof wrote:
Mon Apr 13, 2020 3:10 pm
When I wash my washable filter (It's black foam) after the very good rinse, I put it in a cotton towel and squeeze it dry as good as I can get it more than once. Then let it air dry, a fan will speed drying up if needed. Jim
Good advice! A fan is a good idea, but not for the winter. I will try a paper towel in my filter, althrough there really insn't enough space to squeeze, as you can't really fit your fingers in the plastic frame:

Image

fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Re: Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Mon Apr 13, 2020 5:03 pm

Okie bipap wrote:
Mon Apr 13, 2020 3:28 pm
fuchsia wrote:
Mon Apr 13, 2020 2:54 pm
The Medicare replacement schedule and HHS and sleepapnea.org all suggest replacing the non-disposable filters twice a month. Otherwise, your warranty will be voided.
This sounds like something a DME said in order to sell you more filters. Your users manual has directions for filter maintenance on page 24 which states the fine filter should be replaced after 30 nights of use or sooner if it gets dirty. Just because Medicare allows two filters a month does not mean you must change them that often. A little common sense goes a long way when dealing with DMEs.
Thanks for pointing out the page number. You're right. I just read my manual, and it says replace the reusable filter every 6 months and once a month for the ultra fine one.

I also learned that a dirty filter can cause the machine to operate hotter, which is bad. And not to use anything but genuine Philips Rspironics filters.

It says to just rinse the filter upside down. But I have been soaking mine for an hour and then rinsing. Maybe it will leave soap residue, which will affect the airflow and damage my unit. Have I been doing it wrong?

fuchsia
Posts: 8
Joined: Thu Feb 27, 2020 10:34 am

Re: Questions about filters for DreamStation Auto CPAP

Post by fuchsia » Mon Apr 13, 2020 5:05 pm

Goofproof wrote:
Mon Apr 13, 2020 3:45 pm
raeraeiam wrote:
Mon Apr 13, 2020 3:24 pm
My instructions for cleaning the filters said to dab them with a dry paper towel after rinsing. I did that and it worked fine.
I also rinsed out the micro filter... Although I've only used my machine for 2 weeks so it wasn't very dirty yet. But it definitely looked cleaner after rinsing.
When washing a filter that isn't made to be washed is a poor idea, many filters change their spec flow rate after washing, and no longer keep the machines working at their peak. :roll: Jim
I never throught of this. I will take your advice and just change them once a month, per the manufacturer.

Carlitosbaby
Posts: 13
Joined: Thu May 16, 2019 7:00 pm
Location: New York

Re: Questions about filters for DreamStation Auto CPAP

Post by Carlitosbaby » Wed Apr 15, 2020 6:45 am

fuchsia wrote:
Mon Apr 13, 2020 2:53 pm
Carlitosbaby wrote:
Mon Apr 13, 2020 4:37 am
1. They arent washable
2.i wash main filter once a month when i wake and it's always dry by bedtime. How cold is your house?
3.Not sure but i change the micro filter on 1st of every month and main filter ive never changed in over a year. Just clean it on 1st of month and it still looks brand new 14 months later (wow 14 months already)
4. Never used them i buy 6 packs of original ones on ebay or Amazon they are cheap
Thanks for your help! I like the idea of changing the disposable ones on the first of each month.

As for drying them, my house is normal temperature. I do wash them in the morning, but they are still a little wet when I get ready for bed.

As for the original filters, you cannot buy them on Amazon. I checked. But I found the original Philips Respironics disposable ones for $1.29 each elsewhere, and the original non-disposable ones for $3.99 each, with no tax and free shipping, plus a 15% discount and a price match guarantee.
https://www.amazon.com/dp/B01FS0MKC2/re ... LEbRY5W1HR

I checked my buying history $7.30 with prime there should never be tax on cpap supplies. Comes out 1.21 each

_________________
MaskHumidifier

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: Questions about filters for DreamStation Auto CPAP

Post by Chris8243 » Wed Apr 15, 2020 7:23 am

fuchsia wrote:
Mon Apr 13, 2020 5:00 pm
Goofproof wrote:
Mon Apr 13, 2020 3:10 pm
When I wash my washable filter (It's black foam) after the very good rinse, I put it in a cotton towel and squeeze it dry as good as I can get it more than once. Then let it air dry, a fan will speed drying up if needed. Jim
Good advice! A fan is a good idea, but not for the winter. I will try a paper towel in my filter, althrough there really insn't enough space to squeeze, as you can't really fit your fingers in the plastic frame:

Image
You do know they come apart, right? I'm referring to your photo. The filters are for the air going in that will come out of the hose - not for the benefit of the machine ( as you stated "a dirty filter causes the machine to operate hotter."). No, it's for your benefit, not the machines.

The paper filters are cheap, when I get from my DME, almost free after my insurance, but they will only give me two at a time (when I get a new cushion monthly). That's the Medicare schedule, as others have mentioned. Instead I have bought paper filters on Amazon, 6-pack brand-name and just a few dollars. Maybe you can wash them, never tried.

The larger filter I rinse, not a vigorous cleaning, just rinse well when I'm in a cleaning mode, and nothing special for drying. Always dry by bedtime, been a non-issue for me.