BiPAP autoSV Adv 60 Srs filters blowing out

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
den942
Posts: 69
Joined: Thu Oct 02, 2008 10:37 pm

BiPAP autoSV Adv 60 Srs filters blowing out

Post by den942 » Tue Mar 10, 2020 5:45 am

I've been using this machine over a year. I recently switched from the Quattro full mask to a ResMed AirFit F20. Now the filters keep coming out of the machine. They are halfway out to completely missing by morning. Any ideas why this is happening and what to do to fix the problem?

_________________
MaskHumidifier
Additional Comments: Respironics BiPAP autoSV with Encore SmartCard

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: BiPAP autoSV Adv 60 Srs filters blowing out

Post by Pugsy » Tue Mar 10, 2020 5:54 am

Have you been maybe coughing a little more lately?
Coughing when wearing a full face mask seems to create some back pressure that will cause those filters (which are already just barely staying in place) to go flying out.

Take a skinny rubber band and wrap it around the machine and filter to help it stay put.
That's what a lot of people did back when we had the M series and the System One series that had those filters just barely being held in place.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

den942
Posts: 69
Joined: Thu Oct 02, 2008 10:37 pm

Re: BiPAP autoSV Adv 60 Srs filters blowing out

Post by den942 » Tue Mar 10, 2020 1:01 pm

Pugsy wrote:
Tue Mar 10, 2020 5:54 am
Have you been maybe coughing a little more lately?
Coughing when wearing a full face mask seems to create some back pressure that will cause those filters (which are already just barely staying in place) to go flying out.

Take a skinny rubber band and wrap it around the machine and filter to help it stay put.
That's what a lot of people did back when we had the M series and the System One series that had those filters just barely being held in place.
I will try the rubber band. It seems strange to me that the problem was never seen with the ResMed Mirage Quattro™ Full Face CPAP Mask but is with the new one...

_________________
MaskHumidifier
Additional Comments: Respironics BiPAP autoSV with Encore SmartCard

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: BiPAP autoSV Adv 60 Srs filters blowing out

Post by JDS74 » Fri Mar 13, 2020 7:47 am

+1 on the rubber band. That's what I did on my Respironics 960 ASV machine and the problem was solved.
As to the mask switch, looking at pictures of the two masks, the F20 seems to have a smaller cavity than the QUATTRO. Perhaps the smaller volume inside the mask is letting coughing back pressure get back to the filter.
I don't have either, so this is just speculation based on online images.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.