Water leakage from mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Crazytired
Posts: 19
Joined: Tue Jul 30, 2019 11:58 pm

Water leakage from mask

Post by Crazytired » Tue Feb 18, 2020 1:24 pm

I use a DreamWear nasal cushions mask which I love, but lately there have been drops of water that drip down from the mask. I keep humidity and the hose heat both at max of 5, and no heat in the room (around 60 degrees at night). Any thoughts?? Thank you.

_________________
MaskHumidifier

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Water leakage from mask

Post by LSAT » Tue Feb 18, 2020 1:31 pm

Decrease humidity level and increase hose temp.

User avatar
Pugsy
Posts: 65136
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Water leakage from mask

Post by Pugsy » Tue Feb 18, 2020 1:35 pm

Condensation in the mask is often from the moisture in our own exhale breath tipping the scales and causing the air to cool to release the moisture.

Room air cools the air in the hose and mask and when the air cools too much condensation happens.
See here for various options...posts number 2 and 3.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

Are you using the side fleece wraps around the DreamWear frame? If not, that is the first thing I would add.

You have to warm up the air in the mask some how so that condensation doesn't occur.
Warm air will hold on to more moisture than cool air.

An ambient room temp of 60 degrees makes for the conditions being just ripe for condensation to happen.

Reducing the humidity setting may or may not help much. I know people getting condensation (rain out) in the mask and they aren't even using a humidifier with added water.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.