Traveling no distilled water
Traveling no distilled water
If distilled water is not available when traveling usingan ordinary CPAP machine, is it better to use tap water or no water? Or is the machine unusable?
Re: Traveling no distilled water
Use bottled water until you get back. Tap or bottled water will not affect the machine, but it will leave mineral deposits in your humidifier. These deposits can be cleaned with vinegar and water when you return.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
- Dog Slobber
- Posts: 4220
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Traveling no distilled water
Distilled water is a cleaning convienance.
The fact that so many truly believe that distilled water is a medical requirement, to the point that they won't CPAP without it, is a problem.
The fact that so many truly believe that distilled water is a medical requirement, to the point that they won't CPAP without it, is a problem.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
-
- Posts: 616
- Joined: Tue Dec 15, 2015 7:30 am
Re: Traveling no distilled water
Hi pandora,
"Or is the machine unusable?"
The answer is there in the history of these machines, When I bought my first APAP, a PR S1 551, ten years ago, there was no humidifier, and no humidification.
The sensation in the throat was 'cold and dry' – because even if the air in the room wasn't particularly cold or particularly dry, the fact that it was moving meant it was experienced as cold. And the fact that it was moving meant it did some drying as it went along ....
But that air did its job. Airway held open.
Then the manufacturers brought out the add-on humidifier. A separate unit – that clipped on to the main, blower unit. Now what does that tell you?
That it's an option. A design after-thought. A comfort feature. Something which makes the experience better for most people.
But it also was, and also is, more than a comfort feature for some people. The people whose dry throats turned into sore, irritated, infected throats.
Now, if you haven't used your machine without humidification, you might not know if you are one of that small group which gets throat irriration.
Usually, said irritation takes more than a few days to appear.
Meaning: for short trips, doing without humidifcation is do-able.
You could find out which of these two kinds of user you are at home and before you go on your travels by using your machine without humidification for a few nights. Find out what is true for you.
Me, I know from my early days that I'll manage.
Additionally, because my humidifier is an add-on unit, I can unclip it, and so I have less to carry.
"Or is the machine unusable?"
The answer is there in the history of these machines, When I bought my first APAP, a PR S1 551, ten years ago, there was no humidifier, and no humidification.
The sensation in the throat was 'cold and dry' – because even if the air in the room wasn't particularly cold or particularly dry, the fact that it was moving meant it was experienced as cold. And the fact that it was moving meant it did some drying as it went along ....
But that air did its job. Airway held open.
Then the manufacturers brought out the add-on humidifier. A separate unit – that clipped on to the main, blower unit. Now what does that tell you?
That it's an option. A design after-thought. A comfort feature. Something which makes the experience better for most people.
But it also was, and also is, more than a comfort feature for some people. The people whose dry throats turned into sore, irritated, infected throats.
Now, if you haven't used your machine without humidification, you might not know if you are one of that small group which gets throat irriration.
Usually, said irritation takes more than a few days to appear.
Meaning: for short trips, doing without humidifcation is do-able.
You could find out which of these two kinds of user you are at home and before you go on your travels by using your machine without humidification for a few nights. Find out what is true for you.
Me, I know from my early days that I'll manage.

Additionally, because my humidifier is an add-on unit, I can unclip it, and so I have less to carry.
_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Last edited by rick blaine on Sun Feb 02, 2020 11:14 am, edited 7 times in total.
Re: Traveling no distilled water
You can use any potable water (except carbonated, flavored, or colored). I prefer bottled to tap, as tap water sometimes leaves rusty deposits.
Soak reservoir in vinegar when you get back to remove the deposits.
Soak reservoir in vinegar when you get back to remove the deposits.
Re: Traveling no distilled water
Bottled water may have added minerals 'for taste'. Simplest to just use tap water, rinsing the tank out every morning will reduce buildup of anything, since the one nights minerals won't have had time to really crust up.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Traveling no distilled water
Why? explain why "except carbonated, flavored, or colored"
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Traveling no distilled water
We used to carry distilled water when we traveled, but now just use bottled water. We always carry bottled water in the car when traveling because several of the places we have been had horrible tasting tap water.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: Traveling no distilled water
Carbonated will give you CO2 gas. This can cause headaches and other problems.
Flavored is no good, as it might be unpleasant (even if you like the flavor of the water). Also, it's an additive, and it's recommended not to add any additive.
Colored is also no good as it might cause stains and/or discolor the innards of your machine or even your bedding. Also, coloring is an additive.
Re: Traveling no distilled water
Water's ability to contain dissolved CO2 varies inversely with temperature, even room temperature water doesn't hold much CO2, and the warmer it is, the less CO2 it can hold. in a heated humidifier, there won't be much CO2 for any length of time.
"it might", or, it "might not". Stop GUESSING.
And just how is water supposed to "might" cause stains in the "innards" of the machine? or even the bedding?
You're full of "Ohhhh it MIGHT, with no actual facts.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Traveling no distilled water
palerider wrote: ↑Mon Feb 03, 2020 9:30 amWater's ability to contain dissolved CO2 varies inversely with temperature, even room temperature water doesn't hold much CO2, and the warmer it is, the less CO2 it can hold. in a heated humidifier, there won't be much CO2 for any length of time.
"it might", or, it "might not". Stop GUESSING.And just how is water supposed to "might" cause stains in the "innards" of the machine? or even the bedding?
You're full of "Ohhhh it MIGHT, with no actual facts.
Such a wise guy! Put diet coke in your reservoir and let us know what happens!
Re: Traveling no distilled water
The sky might fall during the night so maybe we should all wear hard hats to bed.
Chicken little is alive and well.
Chicken little is alive and well.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Traveling no distilled water
Now THAT sounds uncomfortable:)
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Traveling no distilled water
Just tired of your mindless stupidity.
Nothing will happen, other than it'll help dissolve any mineral deposits, due to the phosphoric acid.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Traveling no distilled water
At 4'10", I cannot afford to have my spine collapse any more.
Shoulda quit Coke at least twenty years earlier.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |