Anyone tried to use AI to detect RERAs?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
tan
Posts: 565
Joined: Sat Aug 30, 2014 11:58 am

Anyone tried to use AI to detect RERAs?

Post by tan » Mon Dec 30, 2019 3:41 pm

Is that a RERA on the screen?

Image

Rather than tediously scrolling through hours of flow rate data, maybe it's now possible to teach a machine to analyze data and flag RERAs? I am sure I am not the first one with such an idea.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Anyone tried to use AI to detect RERAs?

Post by Dog Slobber » Mon Dec 30, 2019 3:56 pm

The Resmed Airsense 10 AutoSet and AutoSet for Her, (perhaps other models) do *attempt* to flag RERAs.

I never found the RERA flagging to be accurate, reliable or beneficial. I've since moved on to an AirCurve, and don't miss not having the RERA flagging.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Anyone tried to use AI to detect RERAs?

Post by palerider » Mon Dec 30, 2019 5:49 pm

tan wrote:
Mon Dec 30, 2019 3:41 pm
Is that a RERA on the screen?

Image

Rather than tediously scrolling through hours of flow rate data, maybe it's now possible to teach a machine to analyze data and flag RERAs? I am sure I am not the first one with such an idea.
Since you stopped breathing, it's an apnea. But it doesn't *count* because it wasn't at least 10 seconds.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Geer1
Posts: 146
Joined: Fri Oct 25, 2019 8:09 pm

Re: Anyone tried to use AI to detect RERAs?

Post by Geer1 » Mon Dec 30, 2019 6:09 pm

Some of the breathes before pause look a little suspect then afterwords they look more sleep like. I am thinking the pause is sleep transition (if over 10 seconds would be called a sleep transition apnea).

The arousal either happened before the data you posted or possibly right before the cessation. Not enough data to know if it was a breathing related arousal or just spontaneous arousal. You can’t really tell without eeg (unless breathing gets obviously flow limited etc for 10 seconds prior to arousal then it is a safer guess of being a rera.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Anyone tried to use AI to detect RERAs?

Post by chunkyfrog » Mon Dec 30, 2019 7:24 pm

Can you hiccup in your sleep?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

tan
Posts: 565
Joined: Sat Aug 30, 2014 11:58 am

Re: Anyone tried to use AI to detect RERAs?

Post by tan » Tue Dec 31, 2019 3:35 pm

chunkyfrog wrote:
Mon Dec 30, 2019 7:24 pm
Can you hiccup in your sleep?
Maybe, question is whether the increased flow, after a short pause, is an indication of sleep-disordered breathing, or is it something to be ignored, or it's still arousal but not breath-related, but I guess it's impossible to tell without EEG.

Thanks, everyone, for chiming in.

User avatar
greatunclebill
Posts: 1503
Joined: Mon Feb 20, 2012 7:48 pm
Location: L.A. (lower alabama)

Re: Anyone tried to use AI to detect RERAs?

Post by greatunclebill » Tue Dec 31, 2019 10:36 pm

Geer1 wrote:
Mon Dec 30, 2019 6:09 pm
sleep transition apnea
LOL... good one.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: myAir, OSCAR. cms-50D+. airsense 10 auto & (2009) remstar plus m series backups
First diagnosed 1990
please don't ask me to try nasal. i'm a full face person.
the avatar is Rocco, my Lhasa Apso. Number one "Bama fan. 18 championships and counting.
Life member VFW Post 4328 Alabama
MSgt USAF (E-7) medic Retired 1968-1990

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Anyone tried to use AI to detect RERAs?

Post by djams » Wed Jan 01, 2020 1:55 pm

Dog Slobber wrote:
Mon Dec 30, 2019 3:56 pm
The Resmed Airsense 10 AutoSet and AutoSet for Her, (perhaps other models) do *attempt* to flag RERAs.

I never found the RERA flagging to be accurate, reliable or beneficial. I've since moved on to an AirCurve, and don't miss not having the RERA flagging.
Interesting. RERA's were a major component of my SDB, and I ignored then for quite a while. If they meant anything, they'd show up on the machine's sleep report, I thought. Then I watched the sleep disordered breathing video that palerider posts pretty frequently and it prompted me to take a look.

Here is an example of what I'd see regularly while getting my settings tweaked. Three minutes of struggling to breathe, with a big arousal at the end. Looks very much like a hypopnea. I speculate that it didn't meet the criteria for flow reduction. But not sure. I'd have 10-20 of these a night. As you point out, not all of them would look as legit as this one. But OA's+H's+RERA's is what made me accept the advice here to keep raising min pressure.

So I wouldn't tell folks to ignore RERA's, were very real in my case.
RERA.png
RERA.png (17.01 KiB) Viewed 19068 times

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Anyone tried to use AI to detect RERAs?

Post by palerider » Wed Jan 01, 2020 5:30 pm

djams wrote:
Wed Jan 01, 2020 1:55 pm
Dog Slobber wrote:
Mon Dec 30, 2019 3:56 pm
The Resmed Airsense 10 AutoSet and AutoSet for Her, (perhaps other models) do *attempt* to flag RERAs.

I never found the RERA flagging to be accurate, reliable or beneficial. I've since moved on to an AirCurve, and don't miss not having the RERA flagging.
Interesting. RERA's were a major component of my SDB, and I ignored then for quite a while. If they meant anything, they'd show up on the machine's sleep report, I thought. Then I watched the sleep disordered breathing video that palerider posts pretty frequently and it prompted me to take a look.

Here is an example of what I'd see regularly while getting my settings tweaked. Three minutes of struggling to breathe, with a big arousal at the end. Looks very much like a hypopnea. I speculate that it didn't meet the criteria for flow reduction. But not sure. I'd have 10-20 of these a night. As you point out, not all of them would look as legit as this one. But OA's+H's+RERA's is what made me accept the advice here to keep raising min pressure.

So I wouldn't tell folks to ignore RERA's, were very real in my case.
Image
What I see there is increasing flow limitation (the chair shapes on the inhalations) then an arousal, recovery breaths, possibly some moving around, and more recovery breaths.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Wed Jan 01, 2020 7:45 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Geer1
Posts: 146
Joined: Fri Oct 25, 2019 8:09 pm

Re: Anyone tried to use AI to detect RERAs?

Post by Geer1 » Wed Jan 01, 2020 6:11 pm

djams wrote:
Wed Jan 01, 2020 1:55 pm
So I wouldn't tell folks to ignore RERA's, were very real in my case.
RERA.png
I don't believe Dog Slobber was saying that you should ignore RERA's, just that you should take the Autoset flagged RERA events with a grain of salt and that you need to review your data to see if what is flagged as a RERA is actually a RERA (like in your case).

I have now had four RERA's flagged by my autoset and three of them are questionable (like the following example). I believe there have been many other more obvious RERA's not flagged by the machine as well, they simply can't flag RERA's accurately as it requires EEG data which is not available on these machines.
RERA.PNG
RERA.PNG (14.02 KiB) Viewed 19053 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Anyone tried to use AI to detect RERAs?

Post by palerider » Wed Jan 01, 2020 7:46 pm

Geer1 wrote:
Wed Jan 01, 2020 6:11 pm
flag RERA's accurately as it requires EEG data which is not available on these machines.
Thank you, captain oblivious.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Geer1
Posts: 146
Joined: Fri Oct 25, 2019 8:09 pm

Re: Anyone tried to use AI to detect RERAs?

Post by Geer1 » Wed Jan 01, 2020 8:45 pm

palerider wrote:
Wed Jan 01, 2020 5:30 pm
What I see there is increasing flow limitation (the chair shapes on the inhalations) then an arousal, recovery breaths, possibly some moving around, and more recovery breaths.
palerider wrote:
Wed Jan 01, 2020 7:46 pm
Thank you, captain oblivious.