Snoring while on CPAP? (newbie)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MarcieCPAPsux
Posts: 10
Joined: Sat Dec 28, 2019 12:22 pm

Snoring while on CPAP? (newbie)

Post by MarcieCPAPsux » Sun Dec 29, 2019 5:46 am

Greetings!
Well, first of all, I DID IT! I made it 5 hours! Muscled through the crippling anxiety that was causing me to hyperventilate each time I put the mask on and obsess relentlessly throughout the day for the past week.

Second, is it possible to still snore while on CPAP? I have the Philips Respionics Dream Station with DreamWear nasal pillows. I messed up the app that tells me the stats because I turned the machine on and off twice before I calmed myself down, so it only registered 100% mask fit and 0 minutes. And then I also use SnoreScore app that showed I was all over the board!!!

I'm kind of thinking it's just picking up the noise from the machine but curious if I might be doing something wrong. Also, it looked like the pressure didn't move off 4 last night (I woke up to 4). Does it ramp down in the mornings?

Thank you so much to anyone who takes the time to read and reply. This board is my lifeline right now.

User avatar
Dog Slobber
Posts: 4316
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Snoring while on CPAP? (newbie)

Post by Dog Slobber » Sun Dec 29, 2019 9:52 am

Yes, you can still snore while on CPAP.

This might be indicative of not enough pressure.
  • Go through the pinned topics
  • Put your equipment in your signature
  • Get an SD card (unless your machine already has one)
  • Download OSCAR
  • Format your graphs as instructed in pinned topics and post graphs
Theres a good chance your therapy can be improved / optimized
Battery Backup: EcoFlow Delta 2

harrywr2
Posts: 72
Joined: Sun Nov 10, 2019 1:01 pm

Re: Snoring while on CPAP? (newbie)

Post by harrywr2 » Sun Dec 29, 2019 10:32 am

Autoset machines ramp up and down in response to flow limits and snores. If the settings haven't been 'fined tuned' then 'all over the place' is pretty standard.
A minimum pressure of 4 is pretty low for an adult.
I personally feel like I am being starved for air at anything less than 6. My better half stops complaining about my snoring at a minimum pressure of 7.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Snoring while on CPAP? (newbie)

Post by Julie » Sun Dec 29, 2019 11:50 am

I would bump the minimum pressure setting to 6 or 7 (the high can stay high) and see how things go.

User avatar
zonker
Posts: 11427
Joined: Fri Jun 19, 2015 4:36 pm

Re: Snoring while on CPAP? (newbie)

Post by zonker » Sun Dec 29, 2019 12:06 pm

MarcieCPAPsux wrote:
Sun Dec 29, 2019 5:46 am
Greetings!
Well, first of all, I DID IT! I made it 5 hours! Muscled through the crippling anxiety that was causing me to hyperventilate each time I put the mask on and obsess relentlessly throughout the day for the past week.

Second, is it possible to still snore while on CPAP? I have the Philips Respionics Dream Station with DreamWear nasal pillows. I messed up the app that tells me the stats because I turned the machine on and off twice before I calmed myself down, so it only registered 100% mask fit and 0 minutes. And then I also use SnoreScore app that showed I was all over the board!!!

I'm kind of thinking it's just picking up the noise from the machine but curious if I might be doing something wrong. Also, it looked like the pressure didn't move off 4 last night (I woke up to 4). Does it ramp down in the mornings?

Thank you so much to anyone who takes the time to read and reply. This board is my lifeline right now.
welcome to the zoo. if you think "cpapsux", you're doing it wrong. :lol: :lol:

you've gotten plenty of good advice. please pay particular attention to the concise list dog slobber posted. it will be invaluable in getting the help you need. the "newbies" link is in my sig below as well as a link for the oscar software.

one of the things you can do to help ease that anxiety about wearing the mask at night is to wear it while not in bed; reading, watching tv or doing chores. this will help your panicked brain get used to it.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Snoring while on CPAP? (newbie)

Post by palerider » Sun Dec 29, 2019 3:34 pm

MarcieCPAPsux wrote:
Sun Dec 29, 2019 5:46 am
Second, is it possible to still snore while on CPAP?
If your pressure is too low, yes, it's possible to snore.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Snoring while on CPAP? (newbie)

Post by palerider » Sun Dec 29, 2019 3:35 pm

harrywr2 wrote:
Sun Dec 29, 2019 10:32 am
Autoset machines ramp up and down in response to flow limits and snores. If the settings haven't been 'fined tuned' then 'all over the place' is pretty standard.
A minimum pressure of 4 is pretty low for an adult.
I personally feel like I am being starved for air at anything less than 6. My better half stops complaining about my snoring at a minimum pressure of 7.
The OP doesn't have an autoset, but AutoSets and the OP's machine all raise pressure in response to snores, flow limitations, hypopneas and obstructive apneas.

OPs machine isn't as responsive as an AutoSet.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: Snoring while on CPAP? (newbie)

Post by D.H. » Sun Dec 29, 2019 6:03 pm

You can snore on CPAP, but it means your treatment is not optimal. If snoring persists, you should call the doctor who prescribed CPAP.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Snoring while on CPAP? (newbie)

Post by palerider » Sun Dec 29, 2019 6:52 pm

D.H. wrote:
Sun Dec 29, 2019 6:03 pm
call the doctor who prescribed CPAP.
Seriously? SERIOUSLY?

This is a SELF HELP FORUM, you dolt.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Stephaniedp3
Posts: 72
Joined: Sat Dec 21, 2019 1:06 pm
Location: Florida

Re: Snoring while on CPAP? (newbie)

Post by Stephaniedp3 » Sun Dec 29, 2019 7:53 pm

Welcome!

How many nights have you used it so far? I know for my first like 5 nights I dreaded going to sleep and put it off as long as I could!

As far as anxiety when starting I found it helpful to breathe slowly and calmly when ramping up. I like the ramp and starting at 4-5, it works for me. But like other said starting at a higher pressure may work better for you.

Also I know I made this mistake so remember it’s constant air IN. You ARE getting air. When I started I thought the machine was pushing air in and sucking it out. Weird right? But no constant air pressure! We aren’t suffocating. We tell the machine we are breathing it just makes sure we have extra air to KEEP us breathing. :D

MarcieCPAPsux
Posts: 10
Joined: Sat Dec 28, 2019 12:22 pm

Re: Snoring while on CPAP? (newbie)

Post by MarcieCPAPsux » Mon Dec 30, 2019 10:56 am

Thank you all! I have an appointment with the DME tomorrow to hopefully up the beginning pressure. Hopefully that will help. Last night anxiety was horrible, almost 2 hours to feel like I wasn't having a heart attack and suffocating...without the mask!!! I'll get there eventually. Thanks for the feedback about the pressure...I feel hopeful about that! :)

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Snoring while on CPAP? (newbie)

Post by Jas_williams » Mon Dec 30, 2019 11:07 am

MarcieCPAPsux wrote:
Mon Dec 30, 2019 10:56 am
Thank you all! I have an appointment with the DME tomorrow to hopefully up the beginning pressure. Hopefully that will help. Last night anxiety was horrible, almost 2 hours to feel like I wasn't having a heart attack and suffocating...without the mask!!! I'll get there eventually. Thanks for the feedback about the pressure...I feel hopeful about that! :)
The DME can do nothing without a dr’s prescription to change your settings. You however can change your own settings get the clinical manual from https://www.apneaboard.com/adjust-cpap- ... tup-manual

Follow our recommendations

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask