Limiting max pressure?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
joeljjk11
Posts: 365
Joined: Wed Apr 22, 2015 6:40 am

Limiting max pressure?

Post by joeljjk11 » Thu Nov 07, 2019 8:09 pm

Does this ever make sense? Will the CPAP machine ever go to pressures that are not needed or is this not likely? Just curious as to why the max pressure would ever be adjusted.

User avatar
LSAT
Posts: 13377
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Limiting max pressure?

Post by LSAT » Thu Nov 07, 2019 8:13 pm

In most cases it makes no sense to limit max pressure.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Limiting max pressure?

Post by jimbud » Thu Nov 07, 2019 8:32 pm

Aerophagia. :cry:
Pressure induced Centrals. :(
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Limiting max pressure?

Post by chunkyfrog » Thu Nov 07, 2019 10:43 pm

I limit my max, because letting it go nuts tends to wake me up--RUDELY.
I know my AHI could be better, but it is not worth the misery.
Or the stress. Stressful sleep is not good sleep.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Limiting max pressure?

Post by Wulfman... » Thu Nov 07, 2019 11:19 pm

chunkyfrog wrote:
Thu Nov 07, 2019 10:43 pm
I limit my max, because letting it go nuts tends to wake me up--RUDELY.
I know my AHI could be better, but it is not worth the misery.
Or the stress. Stressful sleep is not good sleep.
Some people GET IT! :wink:
It's about finding what works best for YOU.

It doesn't take much pressure variation to mess up my sleep.
Consequently, I use (a moderate) straight pressure and sleep like a rock.......and with very low AHI and other numbers.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: Limiting max pressure?

Post by billbolton » Fri Nov 08, 2019 6:19 pm

LSAT wrote:
Thu Nov 07, 2019 8:13 pm
In most cases it makes no sense to limit max pressure.
In fact, in very few cases does it make any sense to do that.

The proposition that anyone should let a machine fully auto range all the time is popular with marketers, but it actually works fairly poorly in practice.

Auto operation over a limited range is generally useful, but autoranging over a large range is generally not. :idea:

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
LSAT
Posts: 13377
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Limiting max pressure?

Post by LSAT » Fri Nov 08, 2019 6:34 pm

billbolton wrote:
Fri Nov 08, 2019 6:19 pm
LSAT wrote:
Thu Nov 07, 2019 8:13 pm
In most cases it makes no sense to limit max pressure.
In fact, in very few cases does it make any sense to do that.

The proposition that anyone should let a machine fully auto range all the time is popular with marketers, but it actually works fairly poorly in practice.

Auto operation over a limited range is generally useful, but autoranging over a large range is generally not. :idea:
I don't think we're talking about full auto range...just maximum.. For all we know the OPs minimum might be 10

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: Limiting max pressure?

Post by D.H. » Fri Nov 08, 2019 6:57 pm

For most, it's probably best to set the max at the machine capacity.

You would want to set it lower if the higher pressure becomes annoying, such as causing sinus pain or waking you up.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Limiting max pressure?

Post by palerider » Fri Nov 08, 2019 11:34 pm

billbolton wrote:
Fri Nov 08, 2019 6:19 pm
LSAT wrote:
Thu Nov 07, 2019 8:13 pm
In most cases it makes no sense to limit max pressure.
In fact, in very few cases does it make any sense to do that.

The proposition that anyone should let a machine fully auto range all the time is popular with marketers, but it actually works fairly poorly in practice.

Auto operation over a limited range is generally useful, but autoranging over a large range is generally not. :idea:
If you're implying that the Max pressure should be limited as a matter of course, you're completely wrong.

The Auto machine will only raise pressure when there is a need to raise it.

The only *valid* reasons to lower the max pressure are if you suffer from aerophagia... then you have to do a delicate balancing act between getting enough pressure to treat your apnea, and having too much pressure that causes stomach pain.

In the rare event you have pressure emergent centrals, then possibly, lowering max *might* be worthwhile, but again, at the cost of more obstructive apnea.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ragtopcircus
Posts: 220
Joined: Mon Sep 09, 2019 11:55 am

Re: Limiting max pressure?

Post by ragtopcircus » Sun Nov 10, 2019 9:02 pm

I have the max (and min) on my wife’s machine set a little lower than what she really needs because of aerophagia. Her AHI and total time in apnea is still too high, but coming down, and it’s still much better than non-compliance. I periodically nudge the max up slightly as she acclimates.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: BiPAP VAuto 10-25 PS 4.4, and wherever my tempestuous turbinates take me ....

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Limiting max pressure?

Post by Wulfman... » Mon Nov 11, 2019 1:14 pm

palerider wrote:
Fri Nov 08, 2019 11:34 pm

The Auto machine will only raise pressure when there is a need to raise it.
Actually, the Auto machines (running in a range of pressures) will raise pressure when they detect a breathing event that it is programmed to respond to. It doesn't know whether it is a "NEED TO" or not. Some people don't have the breathing patterns that match the pre-programmed algorithms of the APAP machines. Some MAY, but others do not.
Primarily, those precursor/triggering events are........Flow Limitations and/or Snoring........and maybe (depending on the make/model and programming) Hypopneas and/or Apneas. If the minimum pressure is set too low, it may not have time to respond.......and then the event it's supposed to prevent happens anyway. The machine doesn't "KNOW" diddly squat. They're just programmed to respond to precursor events that it detects. And, if an Apnea is already in progress, it won't respond at all.......until it's over with (maybe). It still takes about 10 seconds for these machines to determine what event is occurring.

Bottom line is.......YOU need to find out for yourself by studying the reports from the software what events you're having and how YOUR machine may or may not respond to them.

This is also why setting the minimum pressure to eliminate MOST of the events is so important. We don't know how many of the events the minimum pressure prevents from happening, but if you can eliminate the majority of them, then you've gone a long way to improve your therapy.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Limiting max pressure?

Post by palerider » Mon Nov 11, 2019 1:27 pm

Wulfman... wrote:
Mon Nov 11, 2019 1:14 pm
palerider wrote:
Fri Nov 08, 2019 11:34 pm

The Auto machine will only raise pressure when there is a need to raise it.
Actually, the Auto machines (running in a range of pressures) will raise pressure when they detect a breathing event that it is programmed to respond to. It doesn't know whether it is a "NEED TO" or not. Some people don't have the breathing patterns that match the pre-programmed algorithms of the APAP machines. Some MAY, but others do not.
Primarily, those precursor/triggering events are........Flow Limitations and/or Snoring........and maybe (depending on the make/model and programming) Hypopneas and/or Apneas. If the minimum pressure is set too low, it may not have time to respond.......and then the event it's supposed to prevent happens anyway. The machine doesn't "KNOW" diddly squat. They're just programmed to respond to precursor events that it detects. And, if an Apnea is already in progress, it won't respond at all.......until it's over with (maybe). It still takes about 10 seconds for these machines to determine what event is occurring.
Your luddite level distrust of automation is noted, you've been beating that dead horse for ages,

We all know that you think that straight pressure with no pressure relief is the way to go.

yadda yadda yadda.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Limiting max pressure?

Post by Wulfman... » Mon Nov 11, 2019 1:44 pm

palerider wrote:
Mon Nov 11, 2019 1:27 pm
Wulfman... wrote:
Mon Nov 11, 2019 1:14 pm
palerider wrote:
Fri Nov 08, 2019 11:34 pm

The Auto machine will only raise pressure when there is a need to raise it.
Actually, the Auto machines (running in a range of pressures) will raise pressure when they detect a breathing event that it is programmed to respond to. It doesn't know whether it is a "NEED TO" or not. Some people don't have the breathing patterns that match the pre-programmed algorithms of the APAP machines. Some MAY, but others do not.
Primarily, those precursor/triggering events are........Flow Limitations and/or Snoring........and maybe (depending on the make/model and programming) Hypopneas and/or Apneas. If the minimum pressure is set too low, it may not have time to respond.......and then the event it's supposed to prevent happens anyway. The machine doesn't "KNOW" diddly squat. They're just programmed to respond to precursor events that it detects. And, if an Apnea is already in progress, it won't respond at all.......until it's over with (maybe). It still takes about 10 seconds for these machines to determine what event is occurring.
Your luddite level distrust of automation is noted, you've been beating that dead horse for ages,

We all know that you think that straight pressure with no pressure relief is the way to go.

yadda yadda yadda.
And, your idiotic arrogance has been well known on this forum for years.

But, you left something out......the most important.......as usual. My point has ALWAYS been to get the users to find out what works best for each of them. I found out long ago what works best for ME. I used exhale relief for years, then decided I didn't need it. Tried APAP ranges of pressure, too........decided it didn't work for ME.
Bottom line is.......YOU need to find out for yourself by studying the reports from the software what events you're having and how YOUR machine may or may not respond to them.

This is also why setting the minimum pressure to eliminate MOST of the events is so important. We don't know how many of the events the minimum pressure prevents from happening, but if you can eliminate the majority of them, then you've gone a long way to improve your therapy.
Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Limiting max pressure?

Post by palerider » Mon Nov 11, 2019 2:03 pm

Wulfman... wrote:
Mon Nov 11, 2019 1:44 pm
But, you left something out......the most important.......as usual. My point has ALWAYS been to get the users to find out what works best for each of them. I found out long ago what works best for ME. I used exhale relief for years, then decided I didn't need it. Tried APAP ranges of pressure, too........decided it didn't work for ME.
Bottom line is.......YOU need to find out for yourself by studying the reports from the software what events you're having and how YOUR machine may or may not respond to them.

This is also why setting the minimum pressure to eliminate MOST of the events is so important. We don't know how many of the events the minimum pressure prevents from happening, but if you can eliminate the majority of them, then you've gone a long way to improve your therapy.
Any *reasonable*, intelligent person would realize that I 'left that out' because I wasn't disputing the importance of people becoming involved in their treatment and learning how to evaluate their data.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: Limiting max pressure?

Post by zonker » Mon Nov 11, 2019 2:52 pm

Wulfman... wrote:
Mon Nov 11, 2019 1:14 pm
palerider wrote:
Fri Nov 08, 2019 11:34 pm

The Auto machine will only raise pressure when there is a need to raise it.
Actually, the Auto machines (running in a range of pressures) will raise pressure when they detect a breathing event that it is programmed to respond to. It doesn't know whether it is a "NEED TO" or not.