Philips Respironics System One heater not heating up, stone cold

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Philips Respironics System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 9:24 am

Hi all,

I noticed 3 nights ago my water was full in the morning. It's been the same every day since. This morning when I woke up I popped out the humidifier and felt the heater and it was frigid. The humidifier setting is 3 and worked until 3 nights ago. Is there some other troubleshooting I should try before calling Philips Respironics?

Thanks for your help!

_________________
Mask
Last edited by spamgirl on Sat Oct 19, 2019 9:46 am, edited 1 time in total.

User avatar
Dog Slobber
Posts: 4229
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed System One heater not heating up, stone cold

Post by Dog Slobber » Sat Oct 19, 2019 9:33 am

Before calling ResMed, verify what machine you have.

Your profile and device name suggest you have a Philips Respironics, not a ResMed.

spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Re: Resmed System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 9:46 am

Dog Slobber wrote:
Sat Oct 19, 2019 9:33 am
Before calling ResMed, verify what machine you have.

Your profile and device name suggest you have a Philips Respironics, not a ResMed.
You're right :)

_________________
Mask

User avatar
Pugsy
Posts: 65112
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Philips Respironics System One heater not heating up, stone cold

Post by Pugsy » Sat Oct 19, 2019 10:12 am

First thing to do is try the heater check at the beginning of use instead of at the end of the night.
Turn the machine on and set the humidity to as high as it will go and let it blow for about 20 minutes and then check the temperature of the bottom of the water chamber.
We know it should heat up at the beginning of the night. During the night the heater cycles on and off so the fact that it was cold in the morning could just mean it was in the off cycle. Don't know for sure until you verify at the beginning of use.

If it is indeed still stone cold after about 20 minutes or so....don't assume that the problem is in the humidifier section.
It might be a wiring/sensor problem in the blower unit that is causing the humidifier portion to not heat.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Re: Philips Respironics System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 10:25 am

I tried that and it did not heat up at all. Is there anything else I could do to fix it? Thanks for your advice!

_________________
Mask

User avatar
Pugsy
Posts: 65112
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Philips Respironics System One heater not heating up, stone cold

Post by Pugsy » Sat Oct 19, 2019 10:32 am

Bummer.

The is a known problem with the wiring in the System One machines (both 50 and 60 series) and IF (big IF) you are having the same problem there is a fix but it involves opening up the blower and doing some wiring repair and minor welding.

The entire machine can be sent off to acbio.com or similar place for repair but to be quite honest it is going to probably cost more than the machine is worth.

Respironics won't be able to tell you much except to "get with your equipment supplier"....it's going to be out of warranty anyway.

Let me go see if I can find the how to repair thread for when others had this problem.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65112
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Philips Respironics System One heater not heating up, stone cold

Post by Pugsy » Sat Oct 19, 2019 10:38 am

Okay...here's the thread where they talked about fixing the heater problem on the 50 series machine and I believe same steps would be involved with the 60 series.

That is of course assuming that the burned out wire is the problem.

viewtopic/t102275/Fix-Respironics-Syste ... a9#p957492

I do have a PR S1 60 series BiPap Auto model 760 already heated hose capable that I need to find a new home for.
Blower, humdifier, heated hose, water chamber, travel case...around 1500 hours or so.
$200....send me a PM if interested and I will explain the particulars.
It would very likely cost you that much or more to send it to acbio.com for repairs by the time you add in their fees and shipping.
If you already have a 60 series machine the parts should be interchangeable to help you figure out which part is the problem.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Re: Philips Respironics System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 10:44 am

:( Thank you for your help! I'm in Canada so I'm probably going to look for something up here that I can get immediately. I appreciate the offer, though!

_________________
Mask

spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Re: Philips Respironics System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 10:55 am

Pugsy wrote:
Sat Oct 19, 2019 10:38 am
Okay...here's the thread where they talked about fixing the heater problem on the 50 series machine and I believe same steps would be involved with the 60 series.

That is of course assuming that the burned out wire is the problem.

viewtopic/t102275/Fix-Respironics-Syste ... a9#p957492

I do have a PR S1 60 series BiPap Auto model 760 already heated hose capable that I need to find a new home for.
Blower, humdifier, heated hose, water chamber, travel case...around 1500 hours or so.
$200....send me a PM if interested and I will explain the particulars.
It would very likely cost you that much or more to send it to acbio.com for repairs by the time you add in their fees and shipping.
If you already have a 60 series machine the parts should be interchangeable to help you figure out which part is the problem.
I popped it open and the wiring still looks good. Does that mean it's probably the heater instead?

_________________
Mask

User avatar
Pugsy
Posts: 65112
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Philips Respironics System One heater not heating up, stone cold

Post by Pugsy » Sat Oct 19, 2019 11:13 am

If you can find someone who will let you try a different humidifier box to your existing blower you could see if it will heat up or not.
That will tell you if you can get by with just a new humidifier box or not.
It needs to be a 60 series match up though.
Can't try a 60 series blower with a 50 series humidifier.
The 60 series models have been discontinued for some years now and it may be hard to find someone who has some new humidifiers and they will just let you try one.
Hate to buy a new one and find out the problem is indeed in the blower somewhere.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

spamgirl
Posts: 13
Joined: Fri Aug 28, 2009 2:33 pm

Re: Philips Respironics System One heater not heating up, stone cold

Post by spamgirl » Sat Oct 19, 2019 11:21 am

Good news! It seems that popping it open made something work. That means I now have time to go to my dr, get a requisition, go do a sleep test, and then get the government to pay for a new one. :) Long live medicare for all LOL Thank you for your help!!

_________________
Mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Philips Respironics System One heater not heating up, stone cold

Post by palerider » Sat Oct 19, 2019 6:04 pm

spamgirl wrote:
Sat Oct 19, 2019 10:55 am
Pugsy wrote:
Sat Oct 19, 2019 10:38 am
Okay...here's the thread where they talked about fixing the heater problem on the 50 series machine and I believe same steps would be involved with the 60 series.

That is of course assuming that the burned out wire is the problem.

viewtopic/t102275/Fix-Respironics-Syste ... a9#p957492

I do have a PR S1 60 series BiPap Auto model 760 already heated hose capable that I need to find a new home for.
Blower, humdifier, heated hose, water chamber, travel case...around 1500 hours or so.
$200....send me a PM if interested and I will explain the particulars.
It would very likely cost you that much or more to send it to acbio.com for repairs by the time you add in their fees and shipping.
If you already have a 60 series machine the parts should be interchangeable to help you figure out which part is the problem.
I popped it open and the wiring still looks good. Does that mean it's probably the heater instead?
Unplug and replug in that connector a few times to make sure it's good and clean contact wise.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.