Airsense 10 Autoset Time

Post by K5MOW » Fri Oct 11, 2019 7:10 pm

Good evening everyone

Wanted to ask a question about setting the proper time on my older AirSence 10 Autoset. It is exactly 1 hour ahead and when I tried to set The time back one hour. The unit says you cannot set day or time back. Is there anyway to set the time.

Thank you Roger
Moderate to Severe Sleep Apnea

Airsense 10 Autoset AirStart 10 CPAP

Mirage Quattro Full Face Mask

Roger

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 Autoset Time

Post by Pugsy » Fri Oct 11, 2019 7:19 pm

When did you try to set the time?

It won't let a person go backwards when there is data within that reporting sleep session (which starts at noon per the machine clock).
If you tried to reset it this morning....try again now.
If you tried now and it won't let you....did you have any nap time today after Noon 12PM?

Or just leave it and when DayLight Savings Time changes in a few weeks...it will be correct.

I don't mess with the clock myself. Not all that important to me that the clock on the reports be perfect.
8 hours of sleep is 8 hours of sleep no matter what the start and stop times say they are.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Airsense 10 Autoset Time

Post by chunkyfrog » Fri Oct 11, 2019 7:27 pm

My machine is set for an hour and a half ahead.
Then, no matter whether we are afflicted with standard or saving time, it is always ahead.
Then I can sleep a bit past noon in case I am feeling unwell.
Which is not often these days--which is fine with me.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
K5MOW
Posts: 155
Joined: Mon Oct 04, 2010 9:50 am
Location: Friendswood TX
Contact:

Re: Airsense 10 Autoset Time

Post by K5MOW » Fri Oct 11, 2019 7:41 pm

Thank you both for the information. It is not a big deal. It was more of why I couldn’t do it. Now I know why and how to do it if I want. I will probably just wait for daylight savings time to change like suggested here. After all we’re only looking at a few more weeks.

Again thank you very much Roger
Moderate to Severe Sleep Apnea

Airsense 10 Autoset AirStart 10 CPAP

Mirage Quattro Full Face Mask

Roger

User avatar
Dog Slobber
Posts: 4258
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Airsense 10 Autoset Time

Post by Dog Slobber » Sat Oct 12, 2019 6:44 am

When setting time backward:
  • The ResMed day is noon - noon
  • Don't set it backward past 12:00 noon
  • Don't set it backwards into a previous session
  • If the time frame is sufficiently high set it back in smaller increments over a few days span
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 Autoset Time

Post by palerider » Sat Oct 12, 2019 2:29 pm

Dog Slobber wrote:
Sat Oct 12, 2019 6:44 am
When setting time backward:
  • The ResMed day is noon - noon
  • Don't set it backward past 12:00 noon
  • Don't set it backwards into a previous session
  • If the time frame is sufficiently high set it back in smaller increments over a few days span
Also, best time to set the time back is just before you turn the machine on for bedtime.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11341
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 Autoset Time

Post by zonker » Sat Oct 12, 2019 5:21 pm

Dog Slobber wrote:
Sat Oct 12, 2019 6:44 am
When setting time backward:
  • The ResMed day is noon - noon
  • Don't set it backward past 12:00 noon
  • Don't set it backwards into a previous session
  • If the time frame is sufficiently high set it back in smaller increments over a few days span