Flow Limitation

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
elaine10
Posts: 14
Joined: Sun Jun 02, 2019 10:34 am

Flow Limitation

Post by elaine10 » Sat Oct 05, 2019 8:33 pm

Hi is this an ideal Flow Limitation Graph??? Mine has always looked something like this. AHI most of the ime is below 1.5.
Attachments
fl.JPG
fl.JPG (32.63 KiB) Viewed 12855 times

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Flow Limitation

Post by jimbud » Sat Oct 05, 2019 8:53 pm

elaine10 wrote:
Sat Oct 05, 2019 8:33 pm
Hi is this an ideal Flow Limitation Graph??? Mine has always looked something like this. AHI most of the ime is below 1.5.
Ideal? No.

Hard to say much about it.
There is no context with the rest of your chart.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
Pugsy
Posts: 65268
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Flow Limitation

Post by Pugsy » Sat Oct 05, 2019 9:07 pm

Here's an example of what I see most of the time for a flow limitation graph. Pretty much non existent FLs.
zeroleak.JPG

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Flow Limitation

Post by jimbud » Sat Oct 05, 2019 9:25 pm

Hard to say much about it.
There is no context with the rest of your chart.

What I meant by that is:

Are they causing pressure to go up.
Pressure going up can cause leaks.
Leaks can cause you to wake up.
Wake ups usually mean not so good sleep.
Not so good sleep usually mean feeling crappy.

Context. :)

JPB

Low AHi means little if you are not sleeping well. :(

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Flow Limitation

Post by Miss Emerita » Sat Oct 05, 2019 9:46 pm

I wonder whether you could update us on what happened after this thread:

http://cpaptalk.com/viewtopic/t176972/S ... nshot.html

It’d also be helpful to see a recent screenshot. And could you let us know how are you feeling during the night and during the day?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flow Limitation

Post by palerider » Sun Oct 06, 2019 12:23 am

elaine10 wrote:
Sat Oct 05, 2019 8:33 pm
Hi is this an ideal Flow Limitation Graph??? Mine has always looked something like this. AHI most of the ime is below 1.5.
No, it would cause the pressure to go up.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

elaine10
Posts: 14
Joined: Sun Jun 02, 2019 10:34 am

Re: Flow Limitation

Post by elaine10 » Sun Oct 06, 2019 2:40 am

Hi, these are my graphs last night. Am sorry am so quite new to this. If am not mistaken, Flow Limitation could refer to any occurence that could limit the flow of pressure/air. So the goal is, if am correct, to keep leaks low, keep AHI low, keep flow limitation low. I have dental braces also, so am not sure but maybe it also contributes to higher flow limitation? Thanks.
Attachments
oct 5.JPG
oct 5.JPG (173.93 KiB) Viewed 12799 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flow Limitation

Post by palerider » Sun Oct 06, 2019 2:43 am

elaine10 wrote:
Sun Oct 06, 2019 2:40 am
Hi, these are my graphs last night. Am sorry am so quite new to this. If am not mistaken, Flow Limitation could refer to any occurence that could limit the flow of pressure/air.
You're mistaken.
elaine10 wrote:
Sun Oct 06, 2019 2:40 am
I have dental braces also, so am not sure but maybe it also contributes to higher flow limitation? Thanks.
Not unless you're trying to breathe through your teeth.

https://www.youtube.com/watch?v=-gie2dhqP2c


Image

Min15 on pressure would probably be better for the time being.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

elaine10
Posts: 14
Joined: Sun Jun 02, 2019 10:34 am

Re: Flow Limitation

Post by elaine10 » Sun Oct 06, 2019 10:16 am

Pelerider thank you! Will do just that. Hope the flow limitation improves. Everything in the graphs and stats seem ok right except for the flow limitation? Thanks

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Flow Limitation

Post by Miss Emerita » Sun Oct 06, 2019 10:36 am

After you try the new setting for a few days, try posting another chart so the experts can see how things are going. Oh, and see if you can turn off the pie chart before you take the screenshot.

From another site: “Inspiratory Flow Limitation (IFL) is an increasing resistance in inspiratory flow that slows the volume and flow of air as the inhale proceeds. It is visualized in a flow rate graph as a flattened or downward sloping wave-form rather than a smooth peak.”

If you zoom in on the flow rate graph in an area with flow limitations, you will probably see that the shapes above the zero line are not nice smooth curves but are flat, or have a pointy peak, or have a dent or two in them. One reason this could be a negative is that you are having to work harder to draw breaths. Over the course of a night, this can tire you out.

It really would also be helpful to know how you’re feeling, both during the night and throughout the day. Could you say a little about that?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11432
Joined: Fri Jun 19, 2015 4:36 pm

Re: Flow Limitation

Post by zonker » Sun Oct 06, 2019 12:22 pm

elaine10 wrote:
Sun Oct 06, 2019 10:16 am
Pelerider thank you! Will do just that. Hope the flow limitation improves. Everything in the graphs and stats seem ok right except for the flow limitation? Thanks
if i may humbly make a suggestion?

once you've gathered a fresh set of data, please post it here in this thread. it will better help the experts keep track of you and your situation.

thanks and good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flow Limitation

Post by palerider » Sun Oct 06, 2019 2:26 pm

elaine10 wrote:
Sun Oct 06, 2019 10:16 am
Pelerider thank you! Will do just that. Hope the flow limitation improves. Everything in the graphs and stats seem ok right except for the flow limitation? Thanks
A lot of people would say "it's fine" (the graphs and stats), but I look at the pressure line, and I see a lot of pressure jumps, each time the pressure jumps up, it's because something, snores, flow limitations, hypos, obstructive apneas, is disturbing your breathing, and that causes pressure to go up... all those times when pressure goes up indicate *something* is disturbing your breathing, and hence your sleep.

I like to see a smoother pressure line, indicating fewer breathing disturbances.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

elaine10
Posts: 14
Joined: Sun Jun 02, 2019 10:34 am

Re: Flow Limitation

Post by elaine10 » Tue Oct 08, 2019 10:31 am

Miss Emerita, I increased the min pressure to 14 last night while max stays at 20. This morning is the most energetic that I have ever felt--no headaches, just that feeling of happiness for no reason.

This is my graph for tonight. Will try minimum pressure of 15 as suggested by palerider. Have noticed that pressure goes up where there is Flow Limitation. Thanks guys!
Capture.JPG
Capture.JPG (165.41 KiB) Viewed 12660 times

User avatar
Dog Slobber
Posts: 4321
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Flow Limitation

Post by Dog Slobber » Tue Oct 08, 2019 11:51 am

palerider wrote:
Sun Oct 06, 2019 2:26 pm
elaine10 wrote:
Sun Oct 06, 2019 10:16 am
Pelerider thank you! Will do just that. Hope the flow limitation improves. Everything in the graphs and stats seem ok right except for the flow limitation? Thanks
A lot of people would say "it's fine" (the graphs and stats), but I look at the pressure line, and I see a lot of pressure jumps, each time the pressure jumps up, it's because something, snores, flow limitations, hypos, obstructive apneas, is disturbing your breathing, and that causes pressure to go up... all those times when pressure goes up indicate *something* is disturbing your breathing, and hence your sleep.

I like to see a smoother pressure line, indicating fewer breathing disturbances.
It was precisely this advice from palerider that I followed when I was optimizing my therapy. Some observations as I was increasing my minimum:
  • Initially my pressure line had a lot of volatile up and down movement
  • As I was gradually increasing my minimum, the machine did not need to go as high as it used to.
  • The activity on my Flow Limitation reduced dramatically.
I've included two of my charts for comparison. Last nights (top one) and one from about a year ago about when I started to optimize my therapy.
Note:
  • The reduced Flow Limitation activity
  • The lower AHI
  • And most importantly how the machine does not need to use higher pressure. My minimum is 11.2 and there are just two blips where it increases pressure, and never into the twelves. Contrast that to when I had a minimum of 9. My device spent half the night alive 12 and would gone last 13 but I had the maximum capped.
For me (and I'm guessing many others), increasing the minimum, reduces the the need for higher pressures.
Increase_min.png
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flow Limitation

Post by palerider » Tue Oct 08, 2019 12:10 pm


_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.