Help: AHI-6.07 - What is wrong?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ww
Posts: 523
Joined: Sun Mar 16, 2008 11:58 am

Help: AHI-6.07 - What is wrong?

Post by ww » Tue Sep 17, 2019 7:44 pm

My AHI is very high and not helped with pressures up to 17.

The clear airway apneas are almost as high as the obstructive apneas and the hypopneas are over 4/hour. The events are primarily between 6 and 8 am.

Two years ago, AHI was around 1.0. I need to get a handle on why I cannot achiever the previously low AHI numbers.

I tried to schedule a new sleep test (primarily due to the open airway apneas), but the sleep doctor- nurse called and wanted to leave a machine with me at home overnight. I figured that was what they did to evaluate if you had sleep apnea, which I know I have and guessed their test would be a waste of time. They never called back after that.
Daily Sleepyhead 2019-09-17 at 5.10.03 PM.png
Daily Sleepyhead 2019-09-17 at 5.10.03 PM.png (384.4 KiB) Viewed 2050 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included)
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Help: AHI-6.07 - What is wrong?

Post by Julie » Tue Sep 17, 2019 7:53 pm

Hate to ask, but have you gained even a few lbs lately? It's amazing how little can make a difference.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Help: AHI-6.07 - What is wrong?

Post by palerider » Tue Sep 17, 2019 8:07 pm

:!:
ww wrote:
Tue Sep 17, 2019 7:44 pm
My AHI is very high and not helped with pressures up to 17.

The clear airway apneas are almost as high as the obstructive apneas and the hypopneas are over 4/hour. The events are primarily between 6 and 8 am.

Two years ago, AHI was around 1.0. I need to get a handle on why I cannot achiever the previously low AHI numbers.
Min 14, max 20.

See how much that improves things.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Help: AHI-6.07 - What is wrong?

Post by LSAT » Tue Sep 17, 2019 8:08 pm

Increase your Min pressure to 14 and Max pressure to 20...You can see on your charts that the machine wants to go higher than your 17, but it can't. It doesn't look like you are using EPR..Turn it on and set it to 2.

User avatar
ww
Posts: 523
Joined: Sun Mar 16, 2008 11:58 am

Re: Help: AHI-6.07 - What is wrong?

Post by ww » Thu Sep 19, 2019 12:33 pm

:D Thanks Julia, I'm sure you are right, but not enough to have to adjust the suspension seat on my tractor.

The consensus seems to be 14/20 so I set it up to that and things seem to be better.

LSAT: I also set the epr=2.

I will run it at these settings for a few weeks and let you know how these settings are doing.

Thanks to everyone who replied to this post!!!! This was my first post in many years.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear - Fit Pack (All Sizes Included)
Additional Comments: Titrated on Auto CPAP at 7/14 cm: Only licensed medical professionals can give medical advice or write prescriptions