First 30 days - help

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jstapel1
Posts: 1
Joined: Mon Jul 22, 2019 1:53 pm

First 30 days - help

Post by jstapel1 » Mon Jul 22, 2019 7:00 pm

Hi all. New here. My sleep study showed me with an ahi of 32.5. Titrated to a pressure of 10. My apap is set to 4-14. My ahi seems hi and I am still tired all day. Any help would be appreciated
Screenshot_20190722-205343_DuckDuckGo.jpg
Screenshot_20190722-205343_DuckDuckGo.jpg (562.27 KiB) Viewed 916 times

User avatar
zonker
Posts: 11468
Joined: Fri Jun 19, 2015 4:36 pm

Re: First 30 days - help

Post by zonker » Mon Jul 22, 2019 9:01 pm

jstapel1 wrote:
Mon Jul 22, 2019 7:00 pm
Hi all. New here. My sleep study showed me with an ahi of 32.5. Titrated to a pressure of 10. My apap is set to 4-14. My ahi seems hi and I am still tired all day. Any help would be appreciated Screenshot_20190722-205343_DuckDuckGo.jpg
welcome to the zoo!

more than likely, you are going to want to raise your minimum pressure to 6 or even 7. but we will need to see more data from you.

tell you what...follow the links i have at the bottom of my post. the newbie link will talk about sleepyhead software. but for now, substitute the oscar software i have in my other link. at this juncture, they are pretty much the same.

start gathering data and post some oscar charts. then the experts here (i ain't one!) can offer helpful suggestions.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: First 30 days - help

Post by ajack » Tue Jul 23, 2019 12:24 am

If it was my chart I'd have min 9 max 20 and review. You may need a touch more min

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV